Previous in Forum: Transformer full load   Next in Forum: Motor over current
How to test contact vibration sensors
12/08/2021 2:10 PM
Hi all.
I wonder if anybody can help me figure out a system to test the frequency response of an electro-dynamic contact vibration sensor.
Shape:cylindrical
Diameter 15 mm - Height 13 mm
Impedance: 500 ohm.
Mass; 6 grams
Frequency range: 60 Hz to 12 kHz.
Relative small amplitude of vibrations. I would like to avoid a commercial shaker.
So far I have been using a speaker playing a sine sweep with the sensor glued to the cone. I get a frequency response but how do I calibrate the rig?
Thank you all in advance!
Pathfinder Tags: Sensors calibration response
Good Answers: 88
Re: How to test contact vibration sensors
12/08/2021 5:41 PM
Use a flat response mike in a sound pressure feedback servo loop to stabilize cone sound volume output and make your sweeps slow. Speaker should NOT be on an enclosure but on a relatively heavy flat panel and arrange for as anechoick conditions as possible on both sides. You can then make comparative measurements of sensor output. S.M.
12/08/2021 9:45 PM
Ok thanks. Yes, the speaker in not in an enclosure just like you said and I am using a flat diaphragm speaker from Tectonic.
https://www.parts-express.com/Tectonic-TEBM46C20N-4B-BMR-3-Full-Range-Speaker-4-297-2157
I have few mics around but I think I should buy a tiny condenser capsule from Primo.
Few questions:
Should I get a directional capsule? How far from the diaphragm would you place it?
Could you explain the servo loop? Can I just build one that can controls the amplitude of the (slow) sweep going to the amplifier?
Thank you so much.