Electromagnetic Flow Meter Calibration

02/29/2008 1:53 AM

I have ABB KENT-Taylor Magmaster Electromagnetic flow meter,

I need to calibrate it, For calibaration we need to use the KENT Magmaster software,where can I find the Magmaster software and Instruction manual?.

Can you plese help me for above calibation software tool?

Rds,

Karthik

__________________
Karthik_PE
Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.
Guru

Join Date: Jul 2007
Location: Commissariat de Police, Nouvions, occupied France, 1942.
Posts: 2599
Good Answers: 77
#1

Re: Electromagnetic flow meter calibration

02/29/2008 6:18 AM

Calibrate it? Why? Refer to the manufacturer's type test certificate; that is all you should need.

Unless it needs recalibration on a fluid the conductivity of which falls outside the manufacturer's range. In which case the best thing to do is to use a different measuring technique, and that means a different instrument.

Have you been on to ABB Kent for advice, or is this just an academic exercise?

__________________
Good moaning!
Register to Reply
Active Contributor

Join Date: Oct 2007
Location: Philippines
Posts: 17
#2
In reply to #1

Re: Electromagnetic flow meter calibration

02/29/2008 11:09 PM

Electro Magnetic flow meter are configured at the factory itself. No further calibration is needed on site.

Register to Reply
Guru

Join Date: Nov 2007
Location: Eastern Arizona mountains on Route 666 about a mile from God's country
Posts: 1676
Good Answers: 122
#3

Re: Electromagnetic Flow Meter Calibration

03/03/2008 10:54 AM

Are you actually wanting to calibrate the flow meter? Or are you just wanting to set the operating range?

__________________
They said; "Brain size?" I heard; "Train size?" so I said: "I'll take a small one, thank you."
Register to Reply
Participant

Join Date: Jan 2008
Posts: 4
#4
In reply to #3

Re: Electromagnetic Flow Meter Calibration

03/04/2008 2:49 AM

Actually we need to carryout every year calibration check as a routine maitenance.

for that we have to simulate the signals by ABB-Kent magmaster software.By SIMULATION, uner "TEST mode" we can test the Flow range, by entering Flow value in the converter and then verify corresponding 4-20mA output current siganl in the converter output.

Finally I got from our customer itself. thanks

__________________
Karthik_PE
Register to Reply
Member

Join Date: Jul 2008
Location: N.Ireland
Posts: 9
#5
In reply to #4

Re: Electromagnetic Flow Meter Calibration

07/30/2008 3:21 PM

generally you would setup a flow rig, 2 tanks, a pump and sometimes a certified flowmeter - pump a known volume of water from one tank to the other and check if the magflow reads correctly, if your rig has a flowmeter check the flow rate's match, if not adjust the k-factor slightly and repeat

Register to Reply
Anonymous Poster
#6

Re: Electromagnetic Flow Meter Calibration

01/28/2009 5:05 AM

Well it would appear that you just wish to verify the flowmeter and associated circuitry is working ok.. You don't actually need any fancy software to do this.

All you need is a null modem cable, a pc with a terminal emulator program running such as procomm or Hyperterm. set the comm port to 1, set the program to 4800 N 8 1 and you can talk directly to the flowmeter and control it by generating test signals ie, set the flow range, accum, etc etc... one final thing, sign in as "engineer" at level 1 to gain full access rights... Magmaster software allows you to extract (and upload txt files to the magmaster meter but I don't use it on a daily basis as it has lots of issues.. Stick to hyperterm and you wont go wrong.. The null modem cable has to be used regardless of the software you use.. John (AKA as Takorika) Blenheim New Zealand.

Register to Reply
Anonymous Poster
#7
In reply to #6

Re: Electromagnetic Flow Meter Calibration

01/11/2010 5:42 PM

the easiest way is to log into a hyperterminal and log in as '31*engineer' although this is not the full access password it will allow access to simulate a flow. firstly check menu 41 to see flow span ( ie. the flow that will exact 20 ma) then press escape c1*1 then enter.press escape and type c2*(whatever value is in menu 41)then enter. this will generate 20 ma across terminals IC- and IC+. then press escape and type c2*0 then press enter. this will generate 4ma across IC- and IC+ . To escape test mode press escape then type c1*0 press return and you will be back on live flow....hope this helps....

Register to Reply
Anonymous Poster
#8

Electromagnetic Flow Meter Calibration

05/11/2010 7:37 AM

Dear all I am facing problem with magmaster flow transmitter

when connecting to the transmitter by hyperterminal the first menus ( Read ) will continue scrolling and will never stop and it will not allow me yo login the instrument or wtite any thing just it will continue scrolling

Read flow

Resd net

----

Please help

Reply to me by mail at ahmedlotfy@hotmail.com

thanks

Register to Reply
Participant

Join Date: Apr 2011
Posts: 4
#9

Re: Electromagnetic Flow Meter Calibration

05/24/2011 4:05 AM

U providing a useful information!!!, it is really appreciative. I'm also dealing same products like Electromagnetic Flow Meters, Differential Pressure Transmitters, Temperature Controllers and all other industrial automation products. Please see "us.azbil.com".

Register to Reply
Register to Reply 9 comments
Copy to Clipboard

Users who posted comments:

alindsay81 (1); Anonymous Poster (3); azbil123 (1); Crabtree (1); karthik_PE (1); popoy12345 (1); SHOCKHISCAN (1)

Previous in Forum: Disadvantages of Customized Software   Next in Forum: Message Boards / Forums

Advertisement