Need to select a air compressor for my application

07/17/2008 5:08 PM

I need to figure out the size of compressor required for my application. The compressed air will be used as carrier gas in the GC at 15 psi and it will also be used to clear the gas line (1/4" diameter tube, 200 feet length) at 40 psi pressure every 1 hour. The compressor output will be appx 40 psi, we will reduce it down to 15 psi for GC. The 40 psi will go to a valve which will be open for say, 15 secs to clear the line then close - repeat this every one hour.

for first part uses appx 0.07 SCFM.

for second part, I think to clear gas line i will need (tube ID) * pi * length of tube = 13.08 cubic feet of air within 15 secs at 40 psi pressure. If we dont take pressure ino account i would need 13.08/.25 = 52.32 CFM. how to incooperate pressure into the calculations - may be just divide by pressure.

This application would be running 24/7 for weeks to months and i need to use oil-free air compressor. I dnt want it to be running (ON) all the time. I can add large capacity tank to the compressor. how do i calculate tank size, duty cycle of compressor, ON time of compressor with tank etc? Thanks

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Power-User

Join Date: Mar 2007
Location: Newburgh, IN
Posts: 283
Good Answers: 10
#1

Re: Need to select a air compressor for my application

07/17/2008 8:39 PM

You should be able to find all the formula you need here:

http://www.compressedairchallenge.org/

__________________
Bud Trinkel
Register to Reply
Anonymous Poster
#2

Re: Need to select a air compressor for my application

07/19/2008 1:35 AM

in sizing of the compressor this are information you will need

1. Application

2 Maximum flowrate

3. Maximum working pressure

4. Ambient conditions

5. Required dewpoint

The above information is very important in selecting your compressor. Most of the time engineer neglect this information. They will remember this only when they have already encountered problem in the following:

1. condensation

2. pressure drop

3. energy consumption

This two problem mostly the head ache of enduser.

In calculating your compressor capacity

1 hp = 4 cfm

if you have 52.32cfm total capacity

you can calculate the HP : 52.32/4 = 13.08hp or equivalent to 15hp

with regards to the pressure and time to stop the compressor. its already in compressor features. can be adjusted by the service engineer the settings of this requirements.

the tanks size can be calculated base on compressor capacity and pressure. please see your table on thank calculation.

I hope this idea/information can help you insizing your compressor

Register to Reply
Active Contributor

Join Date: Jul 2008
Posts: 21
#6
In reply to #2

Re: Need to select a air compressor for my application

07/22/2008 11:56 PM

Thanks for taking time to reply. tis information is definitely helpful.

Register to Reply
Guru

Join Date: Jan 2006
Location: Tamworth, UK.
Posts: 1782
Good Answers: 45
#3

Re: Need to select a air compressor for my application

07/19/2008 1:00 PM

Don't forget you are working in cu.ft. A round pipe has an area (Π*r2).

A 1/4" dia pipe has a radius of 1/8" giving an area of 0.049 sq.ins. As you are working in units of ft you need to convert the area to sq.ft by dividing by 144. This is a very small number = 0.000341 sq.ft.

Now multiply by length to get the total internal volume. 200 x 0.00034 = 0.068 cu.ft.

If compressed to 40 psig the pressure ratio is (40+14.7)/14.7 = 3.72. Multiply the Vol by this ratio to get the actual free air volume of air held in the pipe at 40 psig = 0.00126 cu.ft. If this is consumed in 15 secs, then you 4 times this amount in one minute = 0.0005 cfm.

Check my figures for accuracy on your own calculator (I have got big fingers and small buttons)

Once you have got the basic units on a common base you can equate this to your pattern of demand.

GC machines work on a puff of air so no way for the moment do you need a 15hp compressor.

__________________
When arguing, remember mud-slinging = lost ground.
Register to Reply Score 1 for Good Answer
Active Contributor

Join Date: Jul 2008
Posts: 21
#5
In reply to #3

Re: Need to select a air compressor for my application

07/22/2008 11:48 PM

thanks a lot for replying. what is 14.7 value in pressure ratio.. is it a standard formula? Also, u have multiplied the pressure ratio with 0.00034 - but total internal volume is 0.068. again, thanks for ur response its v helpful.

Register to Reply
Guru

Join Date: Jan 2006
Location: Tamworth, UK.
Posts: 1782
Good Answers: 45
#7
In reply to #5

Re: Need to select a air compressor for my application

07/23/2008 7:12 AM

Pressure gauges show the actual pressure above atmospheric pressure. But all the calculations with air use 'absolute' pressures and temperatures. Therefore you have to allow for the atmospheric pressure where you are working - thus when using imperial units (feet and inches and pounds) - you have to add 14.7 psi to all gauge pressures to start with - then deduct 14.7 from the result to get back to gauge pressures when you finish.

You 40 psi is 'gauge' pressure which is in fact 54.7 psi 'absolute'. The pressure ratio is the actual change in 'absolute' pressures. Thus 54.7/14.7 = 3.72.

What this means is; your 200 ft of 1/4" pipe has an actual volume of 0.068 cu.ft - and if it held air at a gauge pressure of 40 psi - this air, if let out and measured - you would find it would occupy a volume of 0.068 x 3.72 = 0.253 cu.ft. (sorry my previous figure was a typo - clumsy fingers).

This matters because compressors are rated in 'free' air volumes. That is, the volume of air at atmospheric pressure.

So if your pipe has to be cleared every 15 secs, or 4 times a minute. Your air flow is 0.253 x 4 = 1.012 cfm.

However, this is not necessarily the actual size of the compressor, because it depends on the frequency of the purging exercise. If only carried out once per hour, the flow rate would 0.253 cf/hr.

Like others have said, you need a small receiver with a regulator and valve.

__________________
When arguing, remember mud-slinging = lost ground.
Register to Reply
Active Contributor

Join Date: Jul 2008
Posts: 21
#8
In reply to #7

Re: Need to select a air compressor for my application

07/24/2008 2:45 AM

thanks for the explanation. I need the compressor to turn on once every 5-6 hrs. so if i use a 20 gallon tank, 0.133680556*20 = 2.67361112 cu ft. (assuming Compressor output =100 psi, I will reduce it down to 40 psig for my application) With pressure = 100 psig, pressure ratio = 7.802, actual volume stored in tank will be 20.86 cu ft... plz correct me if i m wrong.

can u suggest any links with basic formulas related to compressed air calculations/compressor req.

Register to Reply
Guru

Join Date: Jan 2006
Location: Tamworth, UK.
Posts: 1782
Good Answers: 45
#9
In reply to #8

Re: Need to select a air compressor for my application

07/24/2008 5:32 AM

You seem to have got it right, but I take your word for the conversion from galls to cf.

Your compressor size will need to supply 20.86 cf in 5-6 hours.

Air quality will be your real problem.

There are dozens of sites with compressed air formulas - I can't say which is best.

__________________
When arguing, remember mud-slinging = lost ground.
Register to Reply
Guru
United States - US - Statue of Liberty - New Member Hobbies - Fishing - New Member

Join Date: Nov 2007
Location: Gone to Alabama with my banjo on my knee...
Posts: 5595
Good Answers: 20
#4

Re: Need to select a air compressor for my application

07/22/2008 5:04 PM

I never worked with a GC that used air as the carrier, we always used helium or nitrogen, gas cylinders with a regulator to set the working flow. I think if you can use air, even with potential impurities, the compressor should fill a receiver tank with a pressure regulator on the outflow line. That way, the compressor will keep the tank full and the tank will run your flow. I'd still wonder about those impurities, though...

__________________
Veni, vidi, video - I came, I saw, I got it on film.
Register to Reply
Associate

Join Date: Feb 2012
Posts: 37
#10

Re: Need to select a air compressor for my application

05/28/2012 5:45 AM

Hi....I am looking to buy an air compressor to aid in a project basement finishing of a 12' x 12' room. One tool I would like to use is a framing nailer with an air compressor to frame the partition walls. I really don't want to spend a lot of money on a compressor. Can anyone help me by suggesting an appropriate compressor ?

Register to Reply
Register to Reply 10 comments

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

Anonymous Poster (1); BudT (1); EnviroMan (1); horace40 (3); MarkhamCornoit (1); mrunal (3)