Calibrating a Level Switch

11/25/2008 10:53 PM

dear all

i have a level switch of E&H model FMI51 capacitive type. it is intended to be used on ammonium nitrate solution which is viscous liquid. My problem is that i tried to caliberate with water but no response was there. Is something wrong with the Switch or i am using wrong solution. please reply

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.
Active Contributor

Join Date: Nov 2008
Posts: 10
#1

Re: caliberation of level switch

11/26/2008 5:40 AM

hi dear,

Viscosity is not the problem fluid must be conductive,you check the directly means short the common and sensing lead and observe the level switch it working or not ? if no then problem with level switch side. try this.

Register to Reply
Guru
United States - Member - Charter Member Engineering Fields - Instrumentation Engineering - Charter Member

Join Date: Mar 2008
Location: The People's Republic of Massachusetts
Posts: 1946
Good Answers: 73
#2
In reply to #1

Re: caliberation of level switch

11/26/2008 8:23 AM

Yes, it should work, and that's a good way the check for operation. My experience with these is they must be properly calibrated, to work at all. Check you manual, if you don't have one, I think E&H has the info on their website. But it's usually no big deal, hit the reset, however, the probe must not be in the solution when you do this.

The unit should work just fine in water.

__________________
I go into every human encounter expecting to be framed for a crime I didn't commit. Dilbert, 2013
Register to Reply