Test Modes for Induction Measurements

12/22/2008 3:32 AM

I am working in an electronic Company .

We are manufacturing of Electronic components like toroidal high frequency transformers & chokes.

We are newly purchased a measuring instrument -Wayne Kerr Precision Magnetic Analyzer-3260B

In this instrument there are two types of mode of testing for inductance measurements- one is series circute mode & another one is paralell circute mode

What is the difference between these two modes & which mode to be selected while testing an inductance?

Please explain who one know about this issue.

Regards,

Sampathkumar

Register to Reply
Pathfinder Tags: Testing mode
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Power-User

Join Date: Dec 2008
Location: Western Colorado, USA
Posts: 202
Good Answers: 16
#1

Re: Test Modes for Induction Measurements

12/23/2008 2:44 PM

Dear Sampathkumar:

Okay, here's the low down on inductors. The equivalent series circuit is used for low 'Q' (or lossy) inductors, i.e. the inductor has a fairly high DC resistance value. This type of inductor can be difficult to read and get accurate results. The equivalent parallel circuit is used for high 'Q' inductors, those with very low losses (resistance). This type is generally the easiest to measure and with very good accuracy.

Usually toroidal inductors have Qs above 10 and would qualify as high Q so use the parallel equivalent circuit for best results. 'Q's below 10 are generally considered high loss and an inductor with a 'Q' below 1 or 2 is very difficult to measure.

There are several different bridge configurations used for inductance, each having their own pros and cons but I have not seen any bridges that can measure inductors with very low 'Q's <1 with any success directly.

For values of 'Q' less than 100, the series and parallel inductance differ measurably by more than .01% The 'Q' of the equivalent parallel circuit always equals the 'Q' of the equivalent series circuit. The formulae are usually given in the bridge's manual but I can give them is you don't have them. The phase and magnitude of the resulting impedance are identical for both circuits.

Hope this helps.

Regards,

Ed