Using an Optical Bench to Measure Focal Length

11/28/2006 12:23 PM

Dear members of CR4,

I'm an engineering student working as a trainee. I'm working on a project to design and construct an optical bench for measuring different properties of ccd- objectives:

- degree and kind of distortion

- resolution of the objectives for different magnification

- brightness declension of the objective from centre to the edge of the lenses

- Measuring the focal length of the objective….

- Etc.

So far I almost can measure all the properties, using my Optical bench in a good accuracy, beside the focal length.

How can I measure a focal length of an objective (with mount), with out knowing the position of the main Plane?

I have tried to use the Bessel method it doesn't work for objectives with focal length smaller than 20 mm (Its good for singlets) because of the mount. My objectives have a focal length (fixed) of 4,6,12 and 16mm. And I also tried to calculate the focal-length form the object –image distance and the magnification. The result is not good (5-15% inaccuracy). "Here may be I need a complicated correction factor!"

Could you please give me a tip how to solve this problem without using complicated methods?

Register to Reply
Pathfinder Tags: focal length optical bench
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.
Participant

Join Date: Nov 2006
Location: Gent - BELGIUM
Posts: 2
#1

Re: Using an Optical Bench to Measure Focal Length

11/29/2006 4:34 AM

Hi,

In a former life ('70 to '80) I have been using a collimating bench from ASKANIA (Germany) to measure many lens parameters like EFL - FFL - BFL and the like.

Their principle is simple:

- use a long reference collimator objective lens with good characteristics in a small NA and on a small FOV, equipped with a reference target in its focal plane, and a light source for illumination (generally filtered in the green, say 540 nm, for better resolution)

- positionning at infinity of the system by self focussing of this collimator is simple by reflection on a good mirror and looking at the reflected image on the test target

- calibration of the target (with engravings in um or mm) can be done by distance measurement techniques

- calibration of the collimator EFL can be done by theodolite or an autocillimator with angular measurement properties, or simply by measuring its long distance, or combined

- measuring any unknown lens EFL is done by imaging the target from the collimator through this lens, and reading the size of the obtained image with a calibrated microscope eyepiece

- the advantage of this large collimator is that you can change the working peak wavelength even with narrow filters, so measure chromatic aberration also (when the collimator lens characteristics are known, say by optical calculus)

- all of the required components can be purchased of the shelf from sources like Edmund

If this looks to long and to much overhead, any illuminated and calibrated (micronmeter reference) target imaged through your unknown lens by viewing through a theodolite (focussed at infinity) will do the job also; just read the angles D of a small distance h and use h= EFL * tan D; keep distances or angles small to avoid distortion and aberrations.