Please read dilemma #1 (posted in instrumentation) before responding to this thread.
After the 0.5V problem was "solved", I moved on to higher voltages, comparing the results using the Fluke 5720A to the Datron 4700. There were problems at 300V and above, with 1000V being the worst. This time the good results were using the Fluke. With the Datron 4700 (or a 4708) the results showed errors of the device under test (DUT), another Holt 11, to be about +80ppm at low frequencies such as 100Hz. The true error is less than +5ppm at 100Hz. All frequencies were bad (30kHz max at 1kV), with 30kHz being the worst.
Putting a clamp-on toroid on the GPIB cable where it comes out of the computer reduced the problem to about 40ppm at 100Hz. I had the Buss cables in a "star" configuration with the junction tied to the bench frame. Going to a daisy-chain hookup (resulting in fewer and shorter buss cables) reduced the problem to about 25ppm. I had 2 power strips for the instruments on the bench with the computer stuff separate. They were replaced with a 10 outlet strip for the instruments with computer stuff still separate. At that point the 100Hz point read -10ppm (too far the other way).
One day I grounded the bench frame and the computer case to the wall, not expecting any change. The 100Hz point went up to +20ppm. It turned out that the bench frame was grounded through a power strip for customer units that was plugged into another breaker. Unplugging that strip made no difference. I then ungrounded the computer and the bench frame, but 100Hz stayed the same!
The cables to the nano-volt meters had a banana plug connection, so a few inches were unshielded. I ordered new cables and connectors and made new cables that go directly from the Holt 11s to the nano-volt meters with no connection in-between. Using them the problem went back up to +80ppm at 100Hz.
At one point unhooking the LAN cable from the computer helped. I changed all 7 LAN cables in the room to shielded cables and grounded their shields to the wall. It may have helped slightly, but the problem has been getting worse. For the last 3 days I can't get a 100Hz reading better that +60ppm with the Datron 4700 regardless of GPIB cable connections, cables used, or toroid usage. The Fluke 5720A is still good. I have checked the 4700 output (using a divider) with an oscilloscope and with a spectrum analyzer (under load) up to 13.2GHz, and it looks clean. Any ideas?