Can i use other materials like SS-316, Al instead of 4140 as target material for eddy current probes. What will be the effect on performance of probes by using different material?

Register to Reply
Pathfinder Tags: target materail
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.
Guru

Join Date: Aug 2007
Posts: 1790
Good Answers: 87
#1

Re: Eddy current probes performance

02/23/2010 7:06 AM

It can be done but you will have to run a new calibration curve with the new material. With stainless it will change the scale factor (slope of the curve) and it will shorten the linear range. Probably better to buy a probe and proximitor system that is calibrated for your material.