Calibration

07/12/2005 8:21 AM

When calibrating equipment, how should I specify the acceptance criteria (AC)? In case of an AC of 10% how do you do this with the 0 (zero) reading?

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.
Participant

Join Date: Jun 2005
Posts: 2
#1

Calibration Spec

07/13/2005 3:28 AM

You would normally specify calibration limits in terms of +- X%. The problem with a zero point is overcome by starting your calibration at 10%. A general rule of thumb is to calibrate at 10, 25, 50, 75, 90 and 100%. Remember also that as well as the actual tolerance discussed above, you must also allow for uncertainty in measurement, i.e. the accuracy and least significant bit of the calibrating instrument.

Register to Reply