A Null-Balance Potentiometer

02/10/2012 5:19 PM

I am taking a Penn Foster instrumentation course and would like someone to check this question for me. I have studied the book for sometime and would have to say that this would be the correct answer. Question : A null-balance potentiometer indicates the temperature of a thermocouple by,,, A: measuring the current with a galvanometer B: measuring the voltage across the thermocouple reference junction C: making the slide wire voltage equal to the voltage from the thermocouple. After reading about this several times i would have to say the correct answer would be C. Could i be wrong?

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
3
Power-User
Engineering Fields - Instrumentation Engineering - New Member Engineering Fields - Control Engineering - New Member Technical Fields - Technical Writing - New Member Technical Fields - Education - New Member Fans of Old Computers - Apple II -

Join Date: Sep 2011
Location: Bellingham, WA
Posts: 191
Good Answers: 46
#1

Re: A Null-Balance Potentiometer

02/10/2012 7:07 PM

------------------------------

From the FAQ page:

CR4 is not a homework cheat site. While some here might relish the opportunity to sharpen up old rusty skills by working the homework problem, consider the following and consider it well. If you cheat on your homework by using someone else's answers, you are only cheating yourself, because the purpose of any homework or other academic assignments is to help you learn - by practice, repetition, and self-discovery.

------------------------------

However, as an educator I am curious about the Penn-Foster course you are taking, and I appreciate the effort you have made to answer the question, rather than just ask for others to answer it as so many other posters do.

First, null-balance potentiometers are museum pieces. There are some applications in metrology which still use a manual null-balance technique, but for industrial use this is woefully obsolete. Answer "C" looks the best to me, but answer "A" contains an element of truth as well: you manually adjust the slide wire until the voltage equals that of the thermocouple, but the key to knowing when you have achieved that balance is that the galvanometer will register zero current.

Out of curiosity, what kind of instructor feedback do you get with the Penn-Foster curriculum? I'm guessing there must be some interaction and help you receive for your tuition. Is there an instructor from Penn-Foster you can contact for more information when you need clarification or additional help?

Also on this website are worksheets filled with practice problems. The courses themselves are "inverted" in the sense that students are required to research new material on their own, then valuable classroom time is spent solving problems rather than dispensing information (i.e. lecture). The worksheets give students realistic problems to solve, and so you may find them useful in your studies:

__________________
They call me "lightning" when wielding a hammer, because I never strike twice in the same place
Register to Reply Good Answer (Score 3)
Active Contributor

Join Date: Feb 2012
Posts: 13
#2

Re: A Null-Balance Potentiometer

02/10/2012 7:22 PM

Thank You for the reply, I have talked to one of the Penn- Foster teacher and he tried to help me but all he more less said was it's in the book. I"ve read and re-read the book on this question and even had the thought as you have said, that " A " may be a possible answer to. I'am a little confused the more i read about this. Thank You for the website links, i will look at them.

Register to Reply
Power-User
Engineering Fields - Instrumentation Engineering - New Member Engineering Fields - Control Engineering - New Member Technical Fields - Technical Writing - New Member Technical Fields - Education - New Member Fans of Old Computers - Apple II -

Join Date: Sep 2011
Location: Bellingham, WA
Posts: 191
Good Answers: 46
#3
In reply to #2

Re: A Null-Balance Potentiometer

02/10/2012 8:27 PM

Here is another website you may find helpful, as it described the operation of potentiometric voltage measurement circuits like the one in your question:

http://www.allaboutcircuits.com/vol_1/chpt_8/3.html

__________________
They call me "lightning" when wielding a hammer, because I never strike twice in the same place
Register to Reply Score 1 for Good Answer
Power-User

Join Date: Dec 2006
Posts: 442
Good Answers: 32
#7
In reply to #2

Re: A Null-Balance Potentiometer

02/12/2012 8:36 PM

It is not unusual that courses teach information that is outdated, and of little practical use today.

Sometimes the specific examples are not useful today, but are the best, clearest way of getting across a vital, fundamental concept.

Sometimes text books fail on both of these counts. Miserably.

Sometimes you will feel a conflict in giving the text-book correct answer, vs. what you really have researched to be the correct answer.

When you find yourself in this situation, congratulations! You have really learned something, whatever score you got on the test.

Register to Reply
Guru

Join Date: Jun 2011
Location: Phnom Penh
Posts: 4019
Good Answers: 102
#4

Re: A Null-Balance Potentiometer

02/12/2012 6:58 AM

Now, if all the students hitting CR4 were more like theo2 they might get some friendly help.

Tony's reference to voltmeters was pretty informative and shows some null voltage applications. Note that the applications involve measuring voltage not current. There may be a clue in there.

My first thought was a bridge circuit. Routinely used where small ohmic values need to be indirectly measured eg ground resistance measurements.

The answer may be hiding in your text book, you just need to look at the problem out of the box. Clever question if that is the case.

Is this application discussed at all in the text? Does a drawing accompany the question? Have you seen this demonstrated? Are there laboratory exercises in the text?

What is the Thevenin equivalent circuit of a thermocouple? Another clue for you there.....

It's an adventure.

__________________
Difficulty is not an obstacle it is merely an attribute.
Register to Reply
Guru
Hobbies - DIY Welding - New Member

Join Date: Sep 2007
Posts: 1601
Good Answers: 58
#5

Re: A Null-Balance Potentiometer

02/12/2012 9:14 AM

Don't over think this problem. It can't be A) because galvanometers measure current and thermocouple output is calibrated in volts. Galvanometers only measure volts when they are part of a potentiometric measurement. The answer can't be B) because the voltage across the reference junction is equal to the reference junction voltage at reference temperature (normally 0 deg. C). This leaves only C) which is the correct answer and also describes the slide wire mechanism used to convert the galvanometer to a potentiometric indicator.

Register to Reply
Associate

Join Date: Jun 2011
Posts: 39
Good Answers: 1
#6

Re: A Null-Balance Potentiometer

02/12/2012 5:29 PM

I always thought that a cold junction was required when making temperature measurements with a thermocouple since the two connection clips from any measuring instrument will be in contact with the thermocouple leads and will create a potential at their junction of dissimilar metals that will either add or subtract from the reading.

Certainly a null-balance potentiometer will not draw any current from the circuit and will read the true voltage of the source but in this case the cold junctions in the circuit due to the connection will give a false reading.

A better example would have been a 1.5 volt battery in series with a 100 meg resistor then this instrument would read the voltage across the resistor which in this case would be the open circuit voltage of the battery.

Marv

Register to Reply
Guru
Hobbies - DIY Welding - New Member

Join Date: Sep 2007
Posts: 1601
Good Answers: 58
#10
In reply to #6

Re: A Null-Balance Potentiometer

02/13/2012 8:48 AM

There is some truth to what you say but the original problem is correctly worded. In modern instruments, a cold junction is not required because it has been replaced by an electronic equivalent. The terminals of modern instruments are made out of the correct thermocouple metals for the intended thermocouple. If you add to the length of a thermocouple with another alloy wire, you will create new junctions that will add and subtract to the reading. Even this will not affect accuracy if both the new junctions are at the same temperature. None-the-less, it is not considered good practice.

Register to Reply
Guru

Join Date: Jul 2010
Posts: 669
Good Answers: 176
#11
In reply to #10

Re: A Null-Balance Potentiometer

02/13/2012 1:57 PM

> In modern instruments, a cold junction is not required because it has been replaced by an electronic equivalent.

I agree. A physical ice bath and separate thermocouple is not required because ice point or cold junction compensation is now done electronically.

>The terminals of modern instruments are made out of the correct thermocouple metals for the intended thermocouple.

Not true. All instrumentation I know of uses nickel plated copper terminals. The temperature effects of a junction created by a 3rd material, like nickel, is cancelled by the "law of intermediate metals". Omega's diagram explains it well.

Base metal thermocouple connectors (plugs and jacks) are made from respective thermocouple alloys.

All manufacturers assume their field terminal connections are isothermal; both terminals at the same temperature, whether that's the case or not.

>If you add to the length of a thermocouple with another alloy wire, you will create new junctions that will add and subtract to the reading.

That leads to really bad temperature errors, that is, using copper wire as extension wire for thermocouples leads to really bogus temperature values. The practice is more common than I'd like to think.

Register to Reply Score 1 for Good Answer
Guru
Hobbies - DIY Welding - New Member

Join Date: Sep 2007
Posts: 1601
Good Answers: 58
#12
In reply to #11

Re: A Null-Balance Potentiometer

02/13/2012 4:27 PM

I agree 100%. You rephrased what I said. When I used the word "terminals" I meant plugs, jacks. etc. We try to NEVER use non-thermocouple extension wires, but sometimes it is necessary in field conditions. When this happens, we maintain the 2 dissimilar junctions at the same temperature.

Register to Reply
2
Guru

Join Date: Jul 2010
Posts: 669
Good Answers: 176
#13
In reply to #6

Re: A Null-Balance Potentiometer

02/13/2012 8:19 PM

You are correct that the absence of cold junction compensation would give false readings, but the necessity of cold junction compensation (CJC) or ice point compensation (two different terms for the same thing) has been recognized since Mr. Seebeck's era.

All commercial potentiometric thermocouple recorder/indicator/controllers definitely used some form of CJC.

The diagrams below show either

- manual temperature compensation (required use of a thermometer to measure the reference junction temperature, after which an adjustment was made that offset the hot junction reading or

- automatic cold junction compensation mode which used nickel wire located where it reflected the temperature of the field terminals,

Diagrams are from the Instrumentation Training Course, Electronic Instruments Vol 2. (1968).

Register to Reply Good Answer (Score 2)
Power-User

Join Date: Aug 2010
Posts: 209
Good Answers: 8
#8

Re: A Null-Balance Potentiometer

02/13/2012 12:48 AM

This is an interesting thread because (to me) it seems to indicate the perils of using other peoples' language. There is something puzzling about the question. So what is it? I think the problem is in the question, namely that the null-balance potentiometer does not indicate the temperature of the thermocouple, it measures the potential of the thermocouple. This creates a situation where there is not an observable chain of logic that can be rationally examined on its face. Of course, when you are doing your own problem solving, you can phrase the questions your want to answer however you want. You can also examine the experimentally found data and check it with a separate method if actually finding the temperature is the desired goal. Can you think of a different way to phrase the question so that this "problem" would not be present? (remember, this stuff can be fun)

Register to Reply
Guru
Hobbies - DIY Welding - New Member

Join Date: Sep 2007
Posts: 1601
Good Answers: 58
#9
In reply to #8

Re: A Null-Balance Potentiometer

02/13/2012 8:29 AM

You are correct, the output of a thermocouple is measured is in volts. One must then go to a table to convert the measured volts to temperature. I think the problem was worded precisely as intended. Answer (A) is wrong because a galvanometer does not measure volts. It would not have been fair to provide more than one correct answer to a multiple choice question.

Register to Reply
Active Contributor

Join Date: Feb 2012
Posts: 13
#14

Re: A Null-Balance Potentiometer

02/14/2012 7:51 PM

Hello All, Took the test and the correct answer was " C " I want to Thank You all for your input and guidance,, really made me think about and mainly Learn ! Thank You all again!! I'll probably be asking more questions down the road.

Register to Reply
Register to Reply 14 comments

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

Iris (2); Marv Trott (1); Tad (1); theo2 (2); tonykuphaldt (2); Wal (1); welderman (4); woodpower (1)

Previous in Forum: PLC Course   Next in Forum: Electrostatic Interference

Advertisement