I work in a test lab and I have been ask to test the air velocity of a hand held vacuum. I decided to test the air speed using two different methods in order to validate my results. I can measure the inches of water it produces and calculate the air speed or I can use a gauge and measure the air flow directly. I am performing both tests on the same vacuum.
V = 1096.7 SQR (P/D)
V=velocity (ft/min)
Constant=1096.7
P=pressure in inches of water
D=density of air lbs/ft³ (@30.3"Hg and 535°R)
The problem is I get two very different results. For example, 18 inches of water give me V=16,800ft/min. But when I measure the velocity with the air velocity gauge my results show a maximum air speed of 5400ft/min. I have taken every precaution to take the readings at the impellar where air speed is maximum. Both the manometer and air velocity gauge are calibrated.
What could be the reason for the descrepency in the results?