Instrument Type Single wavelength, direct reading colorimeter
Wavelength range: 400 to 700 nm
Measurement Wavelengths: Interchangeable glass filter
420 nm, 460 nm, 510 nm, 530 nm, 570 nm or 605 nm
Readable Resolution: Determined by test factor
Photometric Precision: ±0.001 Absorbance Unit
Range: 0 - 2.00 absorbance units, Transmittance: 0 ~ 100% T,
Absorbance0 ~ 1.999 Abs, Concentration 0 ~ 1999 Conc.,Display: 3 ½ digits,
Response Time: 2 seconds, Warm-up Time: Not required, Lamp: LED
• Operating Temperature: should be between 10° to 45°C
• Operating Humidity: should be between 10% - 90%
• Should have large area illumination (> 1 inch) for sample measurement with Variable Sample Illumination, Detector: Silicon Photodiode
Cuvettes: Round 10 mm ID, 12 mm OD 105mm L Square 10 mm x 10 mm x 45 mm, Sample: 10 mL in capped tube
Sample Chamber: Accepts 25mm diameter flat-bottomed tubes(capped)
Power Source: Battery Operation: 9 Volt Alkaline
Line Operation: 110V/60Hz, 220V/50Hz, with available adapter
Size (L XWX H): 8.5 x 16.2 x 6.7 cm, 3.4 X 6.4 X 2.6 inches
Shipping Weight: 11 oz (312 g) meter only
Serial Interface: RS232, 8 pin m DIN, 9600b, 8, 1, n