I have over 40 years experience in metrology, so asking for help doesn't come easy. Nevertheless, this "spook" is driving me crazy, so I need professional help!
Here's the situation.
I wrote a program to calibrate a secondary standard Holt 11 thermal voltage converter (TVC) set from a primary standard Holt 11 set (1V to 1000V at 10Hz to 1MHz) for AC-DC differences. All seemed good with repeatability of better than 3 parts per million (ppm) for most points. Then I added an option to calibrate a 0.5V unit (Fluke A55) from the 1V Holt 11 (with 0.5V input). The A55 has been calibrated on another system for many years with 1kHz as the reference instead of DC. The new program has better repeatability and should have agreed within about 10ppm at each frequency. There was a obvious problem at 500kHz. It was about 300ppm too large and was larger than the 1MHz value. Lower frequencies were much better.
The two units are connected with a tee (Ballantine 16051A) which has GR-874 connectors to accept the TVCs. It has 4 banana jacks where a voltage source (Fluke 5720A) is connected in 4-wire mode to drive the 2 units. The output of each unit (7mVDC) is connected to separate nano-voltmeters (Agilent 34420A). I tried a Datron 4700 in place of the Fluke 5720A, and the results were much better. After much investigation, A large ground braid tied between the tee case and the Fluke 5720A ground post solved the problem, even though the tee case was already grounded by the low lead.
How did this ground braid fix the problem?