I am trying to check whether the 2 pieces of EPCOS 1-line filter feed-through capacitors are still in good condition.
Description: EPCOS 1-line filter capacitor.
Part Number: B85121-A-B6
Value: 0.05 mfd (Y) 25A, 250V AC 60 Hz, 440V DC
Type: Feed-Through
Usage: The two filter capacitors were connected as series capacitor to the input winding a Siemens Transducer Module, Model 6KA9901 EL/605. The other ends of the capacitor leads are connected to a sensor module.
I am confused and do not know whether the two 1-line filter capacitors are defective. When I measured them with a SANWA CD771 Digital Multi-meter on the ohm scale, the reading is 0.000 ohms but it changes to 000.1 and stayed at 000.2 ohms. I switched the measuring to the capacitor mode and observed the reading starts from 00.40 nf and changes to OL mf (over range reading) and stayed there.
I checked the capacitors on the QMAX 200, an ICFT tester and also found that the V-I trace indicated a short but its trace shifted very slightly later.
I guess these filter capacitors are functioning fine but I could not know a sure and concrete method to test them. Presently I am going to buy a shunt resistor and put it across the across the transducer input and pump in the AC signal (0-150mV) from a voltage injector and measure the waveform before and after the filter capacitors that were connected in series to the winding of the transducer.
Can anyone please help or comment? Am I doing the next correct thing? I would greatly appreciate your help to provide the answer to me. And I buy the beer for you. Cheers to all.
Thank you,all.
Rockybeer