Natural Gas Flow Measurement

10/20/2010 7:59 PM

Once upon a time, the natural gas supply to our facility had an extremely stable composition. In the last few months, we have seen swings from the typical 36.6 MJ/M3 to as high as 41.46 MJ/M3. This change in gas composition imparts upward of 20% change in gas density and affects our boiler gas volumetric metering. We utilize Rosemount 3095MV transmitters for our gas metering, but their accuracy depends on the accuracy of the gas table programmed into the transmitter. Does anyone know of an easy way to have this table updated in realtime, possibly from an external density meter? The preference is to perform this within the transmitter instead of via DCS functions. We can't rely on periodic updates as the swings in heating value and density can transpire in a matter of minutes.

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Guru
Technical Fields - Technical Writing - New Member Engineering Fields - Piping Design Engineering - New Member

Join Date: May 2009
Location: Richland, WA, USA
Posts: 21017
Good Answers: 795
#1

Re: Natural Gas Flow Measurement

10/21/2010 12:27 PM

Put a pressure gauge on it and adjust the density accordingly? I thought heating value was per weight rather than per volume.

__________________
In vino veritas; in cervisia carmen; in aqua E. coli.
Register to Reply
Guru

Join Date: Oct 2010
Posts: 1294
Good Answers: 35
#2
In reply to #1

Re: Natural Gas Flow Measurement

10/21/2010 12:50 PM

When I used to work on coke ovens, we used to install something called, iirc, a Wobbe meter. I think it measures the heating value of the gas. It may be rather expensive for your operation (since I don't know how big your operation is), but perhaps it could be useful.

Register to Reply Score 1 for Good Answer
Guru
Engineering Fields - Instrumentation Engineering - New Member

Join Date: Dec 2009
Location: San Antonio, TX USA
Posts: 844
Good Answers: 29
#5
In reply to #2

Re: Natural Gas Flow Measurement

10/21/2010 3:12 PM

Here you are. Learn about the Wobbe Index.

__________________
"Do not worry about your difficulties in Mathematics. I can assure you mine are still greater". - Albert Einstein
Register to Reply
Associate

Join Date: May 2008
Posts: 30
#7
In reply to #1

Re: Natural Gas Flow Measurement

10/21/2010 4:03 PM

Pressure and temperature are already compensated for in the Multivariable transmitter. It is the gas density that changes over time that we are unable to account for right now.

Register to Reply
Guru
Technical Fields - Technical Writing - New Member Engineering Fields - Piping Design Engineering - New Member

Join Date: May 2009
Location: Richland, WA, USA
Posts: 21017
Good Answers: 795
#8
In reply to #7

Re: Natural Gas Flow Measurement

10/21/2010 5:22 PM

Other than pressure and temperature, what else would change the density nearly as much? Is the composition (e.g., what fractions of various hydrocarbons) really all over the map? Or lots of water vapor or other contamination?

__________________
In vino veritas; in cervisia carmen; in aqua E. coli.
Register to Reply
Associate

Join Date: May 2008
Posts: 30
#9
In reply to #8

Re: Natural Gas Flow Measurement

10/21/2010 5:38 PM

Density varies as much as 20% with the change in gas constituency. Gas composition varies from 95% methane and negligible ethane/propane/pentane/butane to 74% methane and 15% ethane/4% propane. When trying to run at 1.2% excess O2 in flue gas, these swings can really mess with combustion control.

Register to Reply
Guru
Technical Fields - Technical Writing - New Member Engineering Fields - Piping Design Engineering - New Member

Join Date: May 2009
Location: Richland, WA, USA
Posts: 21017
Good Answers: 795
#10
In reply to #9

Re: Natural Gas Flow Measurement

10/21/2010 6:16 PM

Bot, that is a wide variation. Is this typical of gas supplier QC?

__________________
In vino veritas; in cervisia carmen; in aqua E. coli.
Register to Reply
Guru

Join Date: Jul 2010
Posts: 669
Good Answers: 176
#3

Re: Natural Gas Flow Measurement

10/21/2010 2:12 PM

Cosa supplies a variety of calorimeters to determine the BTU content

http://www.cosa-instrument.com/products/calorimeters.html

Register to Reply
Guru
Engineering Fields - Instrumentation Engineering - New Member

Join Date: Dec 2009
Location: San Antonio, TX USA
Posts: 844
Good Answers: 29
#4

Re: Natural Gas Flow Measurement

10/21/2010 3:00 PM

First, that Rosemount of yours can be integrated into a HART loop to make program changes more accessible. Emerson has those guys now if you want to read up on it or look at some of their other products.

Second, there are other flow measurement options to consider. Here's one from Sick that is primarily for custody transfer but it's a good place for you to start. Hang on though, they're proud of their stuff. From what you've said, you'll probably have to go ultrasonic to get the composition information you're looking for. Good luck!

__________________
"Do not worry about your difficulties in Mathematics. I can assure you mine are still greater". - Albert Einstein
Register to Reply
Associate

Join Date: May 2008
Posts: 30
#6
In reply to #4

Re: Natural Gas Flow Measurement

10/21/2010 3:58 PM

Thanx for the comments so far guys. We have a direct reading of HV telemetered to our site via the gas supplier, so we have the information on HV (not absolute constituency), just not a clean way to get it into the gas table built into the Rosemount transmitters. We will entertain changing the gas meters on all of our boilers and dryers to Coriolis meters, but in the interim, I was hoping for a technically feasible and safe method of utilizing existing transmitters until we can arrange delivery of new metering and outage time to install (24/7/350 operation). We want to avoid adding a function within the DCS as the burner control programming is "locked down" by the combustion train provider and they are reluctant to stray from their template.

Register to Reply
Commentator

Join Date: Jan 2006
Location: Western flank of the Pennines in Lancashire England
Posts: 93
Good Answers: 8
#11
In reply to #6

Re: Natural Gas Flow Measurement

10/22/2010 2:15 AM

You have not told us what air/gas ratio equipment you have, but this sounds like it's crying out for Oxygen trim ratio control.

Regards,

Gasman

Register to Reply
Associate

Join Date: May 2008
Posts: 30
#12
In reply to #11

Re: Natural Gas Flow Measurement

10/22/2010 10:59 AM

O2 trim is implemented on the boilers as well as CO feedback. As mentioned, we target 1.2% O2 in the flue, very near stoichiometric. If we see this swing in heat value, it can occur in a matter of 10 seconds sending O2 down to 0.8% and CO climbing. This also causes HP steam header to spike suddenly (more energy in). We can live with the process upsets and are working with the fuel supplier to iron out these swings, but we can't deal with the offsets and errors in our metering as people are making decisions based on bad information.

Register to Reply
2
Commentator

Join Date: Jan 2006
Location: Western flank of the Pennines in Lancashire England
Posts: 93
Good Answers: 8
#13
In reply to #12

Re: Natural Gas Flow Measurement

10/23/2010 3:19 AM

Sorry, I did not pick up the use of O2 trim in your earlier post. Metering a Wobbe change taking place in 10 seconds or less is very difficult. Emerson do a meter with a sample time of 2 minutes using chromatoraphy, Elster do one with a 10 second sample time using ultrasonics, and Zentaur do on with a 5 second to 90% response time. They will all be expensive.

Register to Reply Good Answer (Score 2)
Associate

Join Date: May 2008
Posts: 30
#14
In reply to #13

Re: Natural Gas Flow Measurement

10/23/2010 1:50 PM

Excellent. The Xentaur option appears a bit clumsy, but the Elster option shows definite promise if we can convince the supplier to install and maintain to give us the 10 seconds of resonance time required for realtime display. Thanx much Gasman.

Register to Reply
Participant

Join Date: Nov 2010
Posts: 1
#15
In reply to #6

Re: Natural Gas Flow Measurement

11/02/2010 12:11 PM

You can hook up a gas chromatograph to the flow computer. The GC can analyze the gas and dump components, gravity and heating value into the flow computer. Most gas measurement service companies can help you set this up. Emerson will have contacts for contractors who can do this.

Sounds like you need data right away to adjust plant processes. If not, another less expensive alternative is to procure periodic samples of the gas and have them analyzed. Manually adjust the constituent mol percents in the flow computer as you get the results back from the analysis. You can develop more accurate average readings of these mol percents with an automatic sampling system which accumulates small volumes of a sample into the same cylinder based on time or volume.

Keith Fry

Senior Product Specialist

Quorum Business Solutions

Register to Reply
Register to Reply 15 comments

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

Gasman (2); Iris (1); keithfry (1); RDGRNR (2); rhkramer (1); tersla (5); Tornado (3)

Previous in Forum: Thermocouple Cold Junction Compensation?   Next in Forum: Vibration Measuring

Advertisement