Weigher

08/07/2011 10:21 PM

If you want to do a zero and span calibration for a weigher, and it is not practical to use a known weight for the span.Is it possible to use a precession instrument to induce the mVolts that you will get from the span weight to complete this task.

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Guru
Technical Fields - Technical Writing - New Member Engineering Fields - Piping Design Engineering - New Member

Join Date: May 2009
Location: Richland, WA, USA
Posts: 21017
Good Answers: 795
#1

Re: Weigher

08/07/2011 11:17 PM

No.

Btw, "precession" is the wandering of the earth's or other gyroscopic axis, and does not really apply here.

__________________
In vino veritas; in cervisia carmen; in aqua E. coli.
Register to Reply Score 1 for Good Answer
Power-User

Join Date: Sep 2006
Location: Houston Texas
Posts: 282
Good Answers: 16
#7
In reply to #1

Re: Weigher

08/08/2011 11:34 PM

Not necessarily true, if the precession altered the gravitational pull it would also alter the scale calibration and weight readings.

__________________
Specializing in Dynamic Weighing Systems for Powder and Bulk Solids Handling
Register to Reply Off Topic (Score 5)
Power-User

Join Date: Mar 2011
Location: oz
Posts: 252
Good Answers: 10
#2

Re: Weigher

08/07/2011 11:44 PM

sorry, meant precision

Register to Reply
Guru
United States - Member - New Member Engineering Fields - Control Engineering - New Member

Join Date: Mar 2006
Location: Washington USA
Posts: 566
Good Answers: 53
#3

Re: Weigher

08/08/2011 1:46 AM

I concur with Tornado.

Depending on the acceptable margin of uncertainty, you may be able to check the span of the weigher with a known weight that is a fraction of the full span.

Register to Reply
Guru
Technical Fields - Technical Writing - New Member Engineering Fields - Piping Design Engineering - New Member

Join Date: May 2009
Location: Richland, WA, USA
Posts: 21017
Good Answers: 795
#4

Re: Weigher

08/08/2011 2:00 AM

The mV generator can test whether the downstream electronics are working, but it cannot tell whether the load cell(s) are sending an accurate signal. Thus, it fails to calibrate the system as a whole.

__________________
In vino veritas; in cervisia carmen; in aqua E. coli.
Register to Reply
Guru
Hobbies - Fishing - Old Salt Hobbies - CNC - New Member United States - US - Statue of Liberty - New Member

Join Date: Mar 2007
Location: Rosedale, Maryland USA
Posts: 5197
Good Answers: 266
#5

Re: Weigher

08/08/2011 8:01 AM

No it over looks the mechanics of the scale. Which may not apply the full weight of what ever is on the pan on the load cell. A lot of scales still use lever to isolate the load from the load cell.

__________________
Life is not a journey to the grave with the intention of arriving in a pretty, pristine body but rather to come sliding in sideways, all used up and exclaiming, "Wow, what a ride!"
Register to Reply
Power-User

Join Date: Sep 2006
Location: Houston Texas
Posts: 282
Good Answers: 16
#6

Re: Weigher

08/08/2011 11:26 PM

There are a few loadcell manufacturers that either document or adjust the mV output to allow calibration without test weights.

While the best practice is to calibrate the scale using test weights equaling 50% of the measuring span, there are situations where it is simply impractical to add & properly distribute the required calibration weights. Due to the linearity of the loadcell installation, I will trust a full-span electronic calibration over a poorly performed test weight calibration with insufficient span.

We specify Hardy Instruments C2 loadcells when calibration with test weights is not practical. Click the link for additional information:
http://hardyinstruments.com/process_weighing/c2%AE+electronic+calibration/

__________________
Specializing in Dynamic Weighing Systems for Powder and Bulk Solids Handling
Register to Reply Score 1 for Good Answer
Power-User

Join Date: Mar 2011
Location: oz
Posts: 252
Good Answers: 10
#8
In reply to #6

Re: Weigher

08/09/2011 1:00 AM

Thank you for a good answer, An electronic calibration is the only practical choice we have for our current application.

Register to Reply
Power-User
Engineering Fields - Control Engineering - Newby Member Long Time User Engineering Fields - Instrumentation Engineering - New Member

Join Date: Jun 2006
Posts: 224
Good Answers: 9
#9

Re: Weigher

08/09/2011 2:00 AM

When a load cell is calibrated (in the lab, or by a vendor) a calibration curve of mV output versus pounds input (at a specific excitation voltage) is produced. The calibration should also include a resistance calibration in it. Within this calibration, precision resistors are placed across the active leg of the load cells wheatstone bridge and the mV outputs are measured. This tells you what the pounds input would then be for each resistor used.

Precision resistors are then applied to the load cell in the field as a means for spanning the amplifier. A two point adjustment can be performed by first applying no weight (to adjust the amplifier zero) and then by applying the resistor across the bridge to produce a known output (and to adjust the amplifier gain). Repeat these two steps until zero and gain are stable. Apply a different resistor across the load cell to verify the gain and zero values.

Register to Reply
Guru

Join Date: Dec 2010
Location: North West England
Posts: 1170
Good Answers: 153
#10

Re: Weigher

08/09/2011 3:36 AM

While it is not possible to do by inducing mVolts, it is possible to calibrate without a load.

You may need some modification to you load cell mountings to provide anchorage points. (Jack up your mountings and take out your load cells before you weld anything. A welding earth passing across a load cell will kill it.)

You can simulate the load by pulling down an empty vessel or pushing up a full one by a known amount. You need jacks with integral load cells for this purpose. In the UK these can be hired, but it is usual to sub-contract the procedure to a specialist supplier.

Accuracy of the result equals directly loading with known weights if the hired equipment comes with calibration certificates, but you have to adjust the accuracy tolerance to take into account that this is a secondary method of calibration. If you are calibrating to comply with a regulatory standard, you need to check with the regulating authority that this method is acceptable.

Register to Reply
Power-User

Join Date: Jun 2009
Location: New Zealand
Posts: 403
Good Answers: 5
#11

Re: Weigher

08/09/2011 4:45 AM

Hi Eltech

The answer is yes and no, that is depending on the weigh controller you have, I would probably have to say no, Unless you have a weigh controller from New Zealand manufactured by EMC, their 1400 series of controllers, I would have to say are the best small money can by, for instance if you have a silo that weighs 100 ton how do you put test weights on it, rather difficult isn't it, but with these controllers you can programme in the MV ratio and then Tare or Zero the silo.

Of course there are other details that need to be programmed but test weights are not necessary, nor any calibration device.

When ever i get the chance i use these units, I believe that they are the best on the market, they are easy to programme and reliable.

Hope this of some assistance

Best Regards

Joe

Register to Reply Score 1 for Good Answer
Power-User

Join Date: Mar 2011
Location: oz
Posts: 252
Good Answers: 10
#12
In reply to #11

Re: Weigher

08/09/2011 5:03 AM

exactly that, its for a 200 ton silo.I will have to check if the controller you mention meets the hazardous requirements for that zone.

Register to Reply
Power-User

Join Date: Jun 2009
Location: New Zealand
Posts: 403
Good Answers: 5
#13
In reply to #12

Re: Weigher

08/09/2011 6:02 AM

Eltech, I do not think that you have mentioned this in your post.

If you have a load cell in an hazard area, you need to have an intrinsic barrier between the hazard zone and the safe area.

From now on please be more descriptive with your application.

Cheers

Joe

Register to Reply
Power-User

Join Date: Apr 2005
Location: Scotland
Posts: 164
Good Answers: 8
#14

Re: Weigher

08/09/2011 6:02 AM

The problem with applying simulated inputs to the measuring circuits is you still don't know if the load cells are giving accurate outputs so you will always have a degree of uncertainty about the indicated weights. If the silo contents are being used in a manufacturing process the measured quantities way be incorrect and may spoil the end product. If the contents are being purchased I doubt if you will get Trading Standards to apply their seal to the measuring system. Being able to 'tweak' the output readings on any measuring system will surely ring alarm bells in the Quality Control office!!

If the silo is being refilled from road tankers/lorries I would take the weighbridge measurements for the vehicle in and out and cross check the readings you get on the silo. Not perfect, but it will at least give some confidence about the cell outputs.

If you can't use test weights to calibrate the system you could also hold a couple of spare calibrated load cells and set up a program to send them away for refurbishing and recalibration.

Register to Reply
Register to Reply 14 comments

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

Ace Boeringa (1); BDThompson (1); Delmar (2); eltech (3); jhhassociates (1); Joe Sparky (2); ozzb (1); tom (1); Tornado (2)