Calibration Standards

10/21/2011 2:25 PM

As new transmitters get more accurate it seem's I need to constantly upgrade my test equipment. Here is my quick question:

I have a transmitter with a stated accuracy of .075% and I want to calibrate it .

What accuracy should the test equipment I use be?

Register to Reply
Pathfinder Tags: calibration
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Power-User

Join Date: Mar 2011
Posts: 238
Good Answers: 14
#1

Re: Calibration Standards

10/21/2011 3:27 PM

The simple answer is that you want to keep a 4:1 ratio. So a +/-0.01875% of reading instrument would be great.

Register to Reply Score 1 for Good Answer
Associate

Join Date: Sep 2006
Posts: 47
Good Answers: 1
#2
In reply to #1

Re: Calibration Standards

10/21/2011 3:50 PM

This 4:1 ratio is that a rule of thumb,a good practice or is it a standard.

Register to Reply
Guru

Join Date: Jan 2008
Posts: 1753
Good Answers: 59
#3

Re: Calibration Standards

10/22/2011 12:07 AM

When you talk Calibration and Standards, you really raising NIST Traceable Standards. The Boulder Colorado outfit provides Standards and Services to calibrate agains national standards. Then there are outfits who provide calibration services against their second tier standards, and so on. If you want it certifiable, or to ISO standards, that is the way to go.

You may decide to have your equipment recalibrated periodically by such a service, or have a better standard at your facility, where you yourself can perform periodic recalibration. Neither is inexpensive.

Get well acquainted with it, before making decisions.

Register to Reply
Power-User

Join Date: Dec 2010
Location: Tulsa, OK
Posts: 184
Good Answers: 2
#4

Re: Calibration Standards

10/22/2011 8:24 AM

NIST suggest 4:1 but as a minimum your equipment should be 0.375%

Register to Reply
Guru

Join Date: Jan 2008
Posts: 1753
Good Answers: 59
#5
In reply to #4

Re: Calibration Standards

10/22/2011 9:34 AM

It appears to be a misprint. A test equipment normally needs be more precise, than the stated limits of the equipment to be tested. In this case 4x more.

The other point is that, if you want the results certifiable, it needs to be traceable to NIST standards. Hence, the multitier approach, also the need of periodic recertifications.

Register to Reply
Power-User

Join Date: Oct 2010
Location: OZ, otherwise known as Oklahoma were the wind comes sweeping down the plains.
Posts: 159
Good Answers: 4
#6

Re: Calibration Standards

10/24/2011 11:26 AM

Do you need the .075% accuracy, if not you can calibrate the transmitter to the accuracy your standards will allow.

Register to Reply
Register to Reply 6 comments

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

altennant (1); Grand Poobah (1); harley (1); leveles (2); vargaalex (1)

Previous in Forum: Range of LT   Next in Forum: Analog Output Different with Local Indicator

Advertisement