Thermocouple Type 'K'

03/06/2012 10:40 AM

I used 'K' type thermocouple to measure the oven temperature.

When i measured temperature it reads good value upto 500 degree celcius.

The thermocouple connected via analog card and displays in the panel view.

after 500 dec its reads abnormal value.

What could be the reason

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Power-User

Join Date: Feb 2007
Location: Aberdeen, Scotland
Posts: 393
Good Answers: 21
#1

Re: Thermocouple Type 'K'

03/06/2012 10:41 AM

Out of linear range. I'm a mechanical engineer so that is a guess.

Register to Reply
Guru
Hobbies - Musician - Engineering Fields - Chemical Engineering - New Member Engineering Fields - Control Engineering - New Member Engineering Fields - Instrumentation Engineering - New Member

Join Date: Jan 2007
Location: Moses Lake, WA, USA, Thulcandra - The Silent Planet (C.S. Lewis)
Posts: 4216
Good Answers: 194
#2

Re: Thermocouple Type 'K'

03/06/2012 10:52 AM

That is well within the type K temp range. Does it have a sheath, or is it a bare TC? If it is 304 or 316 stainless steel, it won't stand up to those temperatures. Inconel is the MOC (material of choice) for that temp range.

__________________
"Reason is not automatic. Those who deny it cannot be conquered by it. Do not count on them. Leave them alone." - Ayn Rand
Register to Reply
Guru

Join Date: Jul 2007
Posts: 4448
Good Answers: 143
#3

Re: Thermocouple Type 'K'

03/06/2012 11:29 AM

Just for fun, I'd stick the TC onto a multimeter to see what you get there. 500° C is getting up around 10 mV if I remember right; you may be hitting some kind of limit in a defective analog card.

__________________
"Well, I've wrestled with reality for 35 years, Doctor, and I'm happy to state I finally won out over it." Elwood P. Dowd
Register to Reply
Guru
Hobbies - Fishing - Old Salt Hobbies - CNC - New Member United States - US - Statue of Liberty - New Member

Join Date: Mar 2007
Location: Rosedale, Maryland USA
Posts: 5197
Good Answers: 266
#4
In reply to #3

Re: Thermocouple Type 'K'

03/06/2012 3:32 PM
__________________
Life is not a journey to the grave with the intention of arriving in a pretty, pristine body but rather to come sliding in sideways, all used up and exclaiming, "Wow, what a ride!"
Register to Reply
Guru

Join Date: Jul 2007
Posts: 4448
Good Answers: 143
#5
In reply to #4

Re: Thermocouple Type 'K'

03/06/2012 6:05 PM

So much for memory. And that was the last organ still working (sigh).

__________________
"Well, I've wrestled with reality for 35 years, Doctor, and I'm happy to state I finally won out over it." Elwood P. Dowd
Register to Reply
Associate

Join Date: Jun 2010
Location: India
Posts: 25
#7
In reply to #5

Re: Thermocouple Type 'K'

03/06/2012 11:38 PM

I checked the Tc with temperature controller its measures correctly upto 900 dec

When i connect with the analog card it could not read correctly

Is there any memory problem

Register to Reply
Associate

Join Date: Aug 2008
Location: Sligo, Ireland
Posts: 33
#9
In reply to #7

Re: Thermocouple Type 'K'

03/07/2012 8:20 AM

Has this problem just arisen on a system that was working ok, or has something been changed? Before getting in too deep check the readout instrument.It is possible that the readout instrument is not scaled to read above 500DegC or the scaling has been changed (many of these instruments can be programmed). If it is just a simple LCD / LED readout and the only thing between the Tc and the readout is the card, then it looks like the card, it may be the scaling on that and there may be some means of trimming the card. You really need a Tc calibrator, or a mV source with an equivalents chart to check the system properly.

__________________
Tom Sweeney
Register to Reply
Guru

Join Date: Jul 2007
Posts: 4448
Good Answers: 143
#11
In reply to #7

Re: Thermocouple Type 'K'

03/07/2012 4:24 PM

Only my memory. I remembered the tc voltage incorrectly. Check the ps voltages on the analog card.

__________________
"Well, I've wrestled with reality for 35 years, Doctor, and I'm happy to state I finally won out over it." Elwood P. Dowd
Register to Reply
Guru

Join Date: Jul 2010
Posts: 669
Good Answers: 176
#6

Re: Thermocouple Type 'K'

03/06/2012 8:00 PM

Answer: drift

Any thermocouple is a junction of two dissimilar metals. The interpretation of a temperature value from a millivolt signal comes from millivolt versus temperature reference tables. (yes, it can be a polynomial function or a look-up, but the basis is the standardized reference tables)

Those reference tables are valid only when the junction is made from 'pure' alloys, in the case of type K, chromel and alumel alloys.

High temperatures promote ion transfer between the junction and the protection tube, protection sheath or even the atmosphere. The hotter the junction, the more actively ion transfer takes place. The results of ion transfer is chemical pollution; the junction is no longer made up of its original 'pure' alloys, and most importantly, the standard reference tables do not apply.

The thermocouple in question has a polluted junction and hence it is no longer a type K thermocouple, it's a thermocouple of some unique chemical composition with its own unique and unpublished millivolt versus temperature table. The millivolt signal does not correspond to the published standard thermocouple reference tables, so the temperature value is incorrect.

Drift can be in either direction from the true value (up or down) and of varying magnitude.

There is no cure for a drifting thermocouple, except replacement with a new thermocouple. There is no correction factor for a drifting thermocouple because the chemical make-up is an unknown, hence there's no basis for making a correction.

Industries with critical temperature measurement requirements, like heat treating of aircraft components, require periodic replacement of thermocouples to avoid the worst cases of thermocouple drift.

Register to Reply
Guru

Join Date: Jun 2006
Posts: 501
Good Answers: 8
#8

Re: Thermocouple Type 'K'

03/07/2012 4:09 AM

Since the thermocuple reads correctly on a temperature control you must look at your analog board. Thermocouples must have a cold junction either real or a compensating electronic artifical junction. Even though the thermocouple is rated for a highter temperature, I have cheap multimeter that reads type K TCs but only to 1000 degrees F. So the analog board may not have the correct range. Or there may be a bad connection on the board - when you are dealing with millivolts it does not take much to give a false reading.

Register to Reply
Guru
United States - Member - Born, raised halfway 'round .....

Join Date: May 2010
Location: Metro.Manila, Philippines.
Posts: 1269
Good Answers: 27
#10

Re: Thermocouple Type 'K'

03/07/2012 9:23 AM

You may have to re-calibrate your analog circuit's sensitivity, range and linearity so that it will track and match linearly the thermocouple's temperature linear range.

__________________
vsar
Register to Reply
Guru

Join Date: Jun 2006
Posts: 501
Good Answers: 8
#12
In reply to #10

Re: Thermocouple Type 'K'

03/07/2012 5:59 PM

Thermocouples are linear! One of the problems when designing a calibrator or recorder.

Register to Reply
Member

Join Date: Dec 2010
Location: Los Angeles
Posts: 9
#13
In reply to #12

Re: Thermocouple Type 'K'

03/07/2012 7:23 PM

Thermocouples are not linear. There are devices designed to linearize the signal though. Also, these devices have a cold junction reference to account for varying temperature of the instrument itsself. Please see http://www.icpdas-usa.com/products.php?PID=3246 as an example of device of converting TC readings to on demand computer readable digital signals. It's type K range is up to 1372C.

Register to Reply Score 1 for Good Answer
Guru

Join Date: Jun 2006
Posts: 501
Good Answers: 8
#15
In reply to #13

Re: Thermocouple Type 'K'

03/19/2012 7:15 PM

Drink an extra glass of wine and you leave out the "NOT"!

Register to Reply
Member

Join Date: Dec 2010
Location: Los Angeles
Posts: 9
#16
In reply to #15

Re: Thermocouple Type 'K'

03/20/2012 2:57 PM

Yes, I wondered if it might be a typo.

Register to Reply
Guru
Hobbies - Musician - Engineering Fields - Chemical Engineering - New Member Engineering Fields - Control Engineering - New Member Engineering Fields - Instrumentation Engineering - New Member

Join Date: Jan 2007
Location: Moses Lake, WA, USA, Thulcandra - The Silent Planet (C.S. Lewis)
Posts: 4216
Good Answers: 194
#17
In reply to #13

Re: Thermocouple Type 'K'

03/20/2012 3:46 PM

I'm giving you a GA for that. TCs are indeed not linear:

__________________
"Reason is not automatic. Those who deny it cannot be conquered by it. Do not count on them. Leave them alone." - Ayn Rand
Register to Reply
Power-User
India - Member - Justin Anto Engineering Fields - Electrical Engineering - Marks never made me laugh, but good memories did

Join Date: Sep 2010
Location: Thrissur, Kerala, India
Posts: 177
Good Answers: 3
#14

Re: Thermocouple Type 'K'

03/18/2012 5:26 AM

Calibrate the temperature measurement system in a calibration laboratory.

__________________
If you can't find a way, make one........
Register to Reply
Register to Reply 17 comments

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

Iris (1); jmart23 (3); justinanto1987 (1); MACA (1); Mikerho (2); ozzb (1); senthilkodisundaram (1); TechEric (2); Tom Sweeney (1); TVP45 (3); vsar (1)