Previous in Forum: Servo Motors   Next in Forum: Measurement Dropped During Lightening
6 comments
Silica Testing Meter
10/29/2012 3:11 AM
We purchased new Hanna silica photometer HI 96705 for silica testing but meter shows the high reading than actual. When we test from photometer (Hanna) 1.5 to 1.9 mg/L
Reading from Merx Silica Kit ( Visual testing) 0.64 to 1.0 mg/L WE also contact Hanna Inst Technical Dept they say Meter is Accurate & showing the correct reading.
Please describe why the reading is unmatched in both.
Re: Silica Testing Meter
10/29/2012 4:16 AM
OK. Please state the temperatures at which both measurements are being taken.
10/31/2012 3:15 AM
sir
Both testing are done at same temp. between 25 to 30 Deg C
10/29/2012 7:50 AM
This will be another exercise in futility.
10/29/2012 10:54 PM
As the old saying says: the man who has one clock knows what time it is. The one who has two clocks is never sure.
I suggest preparing a set of standard samples of known silica concentrations determined by analytical methods. Then read the samples with both instruments to see which one is right.
If a normalized calibration protocol exists, follow it. The instruments shall be calibrated periodically, especially if transactions based on silica contents are performed.
10/31/2012 3:19 AM
sir
the instrument is new & also we caliberate it with the standard sample which are supplied by Hanna.
10/31/2012 4:22 AM
So what?