Silica Testing Meter

10/29/2012 3:11 AM

We purchased new Hanna silica photometer HI 96705 for silica testing but meter shows the high reading than actual. When we test from photometer (Hanna) 1.5 to 1.9 mg/L

Reading from Merx Silica Kit ( Visual testing) 0.64 to 1.0 mg/L WE also contact Hanna Inst Technical Dept they say Meter is Accurate & showing the correct reading.

Please describe why the reading is unmatched in both.

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#1

Re: Silica Testing Meter

10/29/2012 4:16 AM

OK. Please state the temperatures at which both measurements are being taken.

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
Associate

Join Date: May 2012
Posts: 26
#4
In reply to #1

Re: Silica Testing Meter

10/31/2012 3:15 AM

sir

Both testing are done at same temp. between 25 to 30 Deg C

Register to Reply
Guru

Join Date: Oct 2008
Posts: 42355
Good Answers: 1693
#2

Re: Silica Testing Meter

10/29/2012 7:50 AM

This will be another exercise in futility.

Register to Reply
Active Contributor

Join Date: Aug 2012
Posts: 24
Good Answers: 2
#3
In reply to #2

Re: Silica Testing Meter

10/29/2012 10:54 PM

As the old saying says: the man who has one clock knows what time it is. The one who has two clocks is never sure.

I suggest preparing a set of standard samples of known silica concentrations determined by analytical methods. Then read the samples with both instruments to see which one is right.

If a normalized calibration protocol exists, follow it. The instruments shall be calibrated periodically, especially if transactions based on silica contents are performed.

Register to Reply Score 1 for Good Answer
Associate

Join Date: May 2012
Posts: 26
#5
In reply to #3

Re: Silica Testing Meter

10/31/2012 3:19 AM

sir

the instrument is new & also we caliberate it with the standard sample which are supplied by Hanna.

Register to Reply
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#6
In reply to #5

Re: Silica Testing Meter

10/31/2012 4:22 AM

So what?

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
Register to Reply 6 comments

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

Lalit Shukla (2); lyn (1); OSCAR FRIAS (1); PWSlack (2)

Previous in Forum: Servo Motors   Next in Forum: Measurement Dropped During Lightening

Advertisement