Isolating Transformer Alternative

02/24/2013 3:57 AM

In one of our projects today, we were fitting an isolating transformer at the input terminals of a control panel for a Laboratory. I understand the need for the isolation of the supply source from the panel circuit. However, the specs sheet for the project states the transformer configuration to be 240V in/240V out.

My question is, would there be an alternative to providing the same level of isolation as the isolating transformer with out using a transformer? What I'm opting for is a more economic and room efficient alternative for the isolating transformer, if one exists.

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

Good Answers:

These comments received enough positive votes to make them "good answers".
3
Guru

Join Date: Sep 2012
Location: Iowa, USA
Posts: 577
Good Answers: 50
#1

Re: Isolating transformer Alternative.

02/24/2013 4:44 AM

How much power do you need to move through this isolation?

If it's only a discrete control signal, then opto isolators can be used. But I'm guessing it's power supply isolation so your stuck, unless you want to convert to DC, invert switch through a high frequency isolation transformer (switching power supply with opto isolated feedback), then convert back to your frequency the system needs via an inverter. Lot's more complex.

You need to state what signals are, and how much power is needed through this isolation.

__________________
ignator -
Register to Reply Good Answer (Score 3)
Associate

Join Date: Jul 2012
Posts: 39
#2
In reply to #1

Re: Isolating transformer Alternative.

02/24/2013 5:32 AM

The signal is AC power, not a control signal(discrete). And our company only deals with LV products, but my question really is more general than relates to the particular application I mentioned earlier.

I appreciated your participation in this thread.

Register to Reply
Commentator

Join Date: Nov 2012
Location: Sunnyvale, CA
Posts: 62
Good Answers: 10
#8
In reply to #1

Re: Isolating transformer Alternative.

02/24/2013 11:52 PM

Good answer, but I'll add one note. The isolation transformer is 1:1 primary to secondary winding ratio, and its sizing is determined by the power it needs to deliver, not by the voltage. It needs to be large enough to deliver the required power without saturating. The fact that it's spec'd as 220 in / 220 out is irrelevant per se -- except that the insulation on the windings must be adequate for 220. That doesn't affect the size of the core to any extent.

The alternative that ignator describes -- AC to DC to HF AC and back -- actually can be used to deliver isolated power in a smaller and lighter package. There are situations it may be justified, but it won't be on the basis of cost.

Register to Reply
4
Guru
United Kingdom - Member - New Member

Join Date: Aug 2010
Location: Resting under the Major Oak
Posts: 4347
Good Answers: 181
#3

Re: Isolating transformer Alternative.

02/24/2013 6:00 AM

You're not going to get a cheaper alternative if that's what you're looking for.

__________________
The reports of my death are greatly exaggerated.
Register to Reply Good Answer (Score 4)
Associate

Join Date: Jul 2012
Posts: 39
#4
In reply to #3

Re: Isolating transformer Alternative.

02/24/2013 7:39 AM

Thank you, this pretty much answers it.

Register to Reply
Guru
Engineering Fields - Electrical Engineering - Been there, done that. Engineering Fields - Control Engineering - New Member

Join Date: Dec 2008
Location: Long Island NY
Posts: 15600
Good Answers: 981
#5

Re: Isolating Transformer Alternative

02/24/2013 11:04 AM

Since this is a laboratory there can be many different reasons for the use of a specific isolation transformer that will not be obvious to many. During testing an instrument and/or device under test (DUT) may be pulsed or sustained at a voltage different than ground potential. Ground loops can make reading tiny signals impractical if not impossible. Having a power feed path with a core at or near saturation can reduce non-linear switching noises from the grid to enter sensitive circuits. Isolating power can be a safer usage for electrocution can only happen with two errant circuit paths.

This is not an exhaustive list of reasons an isolation transformer may be used. Without knowing why a transformer was inserted in the power feed, an alternative cannot be offered. Very often in a legacy design, an isolation transformer may have been needed at one time but with optical and wireless coupling capabilities of sensors and the real possibility of a test no longer being performed the need can be no longer present for an isolation transformer.

__________________
"Don't disturb my circles." translation of Archimedes last words
Register to Reply
Guru

Join Date: Oct 2006
Location: Surrey BC Canada
Posts: 1571
Good Answers: 42
#6

Re: Isolating Transformer Alternative

02/24/2013 7:04 PM

There have been many reasons to isolate and not to isolate.

Current North American codes would require the grounding of the secondary of your transformer - unless it is a special application, then Ground Fault Indication/Protection would be required.

You did not state 1 or 3 phase, KW / Kvar, or the reason it needed to be isolated.

With legacy 3 phase to DC power conversion equipment isolation transformers were sometimes used to provide the necessary impedance for the converters. Three phase reactors can often be used just as effectively.

More description and application details would be welcome.

Register to Reply
Power-User

Join Date: Nov 2012
Posts: 142
Good Answers: 14
#7

Re: Isolating Transformer Alternative

02/24/2013 10:25 PM

Be very careful when using an isolated supply to power more than one class 1 device as the isolated supply will not detect a first fault to frame, and any subsequent fault to frame via the other active conductor, and a person coming between the two could prove fatal.

Register to Reply
Guru
Australia - Member - New Member Engineering Fields - Mechanical Engineering - New Member

Join Date: Sep 2006
Location: NSW Australia
Posts: 1101
Good Answers: 23
#9
In reply to #7

Re: Isolating Transformer Alternative

02/25/2013 12:00 AM

Ground fault indication can be built in to the secondary side of the supply, and is recommended.

Also this can be used to trip the supply either on the secondary or primary depending on requirements of the distribution system

__________________
Dont get on to the roundabout if you dont know how to get off
Register to Reply
Power-User

Join Date: Oct 2006
Location: Kolkata, India
Posts: 447
#10

Re: Isolating Transformer Alternative

02/25/2013 9:37 AM

For what's is the purpose you want to use isolation Trafo? If it is to limit fault, you can use Short ckt current protector.

__________________
BNDas
Register to Reply
Guru

Join Date: Feb 2006
Posts: 1758
Good Answers: 6
#11

Re: Isolating Transformer Alternative

02/25/2013 10:13 AM

Isolation is at Life-saving level for men and equipment.

Its power-rating depends on the load put on this transformer.

Input voltage may vary from 240V but output must be 240V as it is specified for the equipment put on its load.

NB: The voltage indicated generally means the normal supply voltage of the country / area and not a tight binding as you will notice in different countries on 110 V(normal) quite a variety is:

110, 115, ..

and for 220V (Normal) 220, 230, 240.

In star that is 3+N for output of 2 phases 380, 400, ...

So look for the the tolerance for input and output voltage requirement and you will be relaxed.

Have a good day.

Register to Reply
Register to Reply 11 comments

Good Answers:

These comments received enough positive votes to make them "good answers".
Copy to Clipboard

Users who posted comments:

AW (2); biswanath.das (1); bwilko (1); garth (1); GW (1); Haajee (1); ignator (1); redfred (1); Silverthorn (1); TonyS (1)

Previous in Forum: DC 110V 10A Battery Charger   Next in Forum: Use Aluminum Wire or Tubular Busbar

Advertisement