Isolated Power System in Operation Rooms

05/29/2014 10:51 AM

Hello, what if a transformer used for isolation in operation rooms fail? Since we make transfer for only ups or cable failures there occurs a critical situation in case a transformer fails. What will a technician do if it fails suddenly and there goes on an operation at the same time?

__________________
jinxnao
Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Guru

Join Date: Mar 2007
Location: by the beach in Florida
Posts: 33392
Good Answers: 1817
#1

Re: Isolated power system in operation rooms

05/29/2014 10:55 AM

It depends on the critical nature of the application....If it is fail-safe design you have a redundant system backup....

__________________
All living things seek to control their own destiny....this is the purpose of life
Register to Reply
Guru

Join Date: Jun 2013
Location: Central Canada
Posts: 677
Good Answers: 28
#2

Re: Isolated power system in operation rooms

05/29/2014 1:05 PM

Do you wish to design a safety protocol to follow in this case ?

One of the primary functions would be to remove the patient immediately or ASAP.

Hopefully there is more than one OR and it has a separate XFMR because

another primary function would be to CLOSE the OR until the fault is repaired.

Safety of patient and staff first; repair second.

__________________
Smart as a post and twice as fast.
Register to Reply Score 1 for Good Answer
Guru
New Zealand - Member - Kiwi Popular Science - Weaponology - New Member Engineering Fields - Power Engineering - New Member Engineering Fields - Electrical Engineering - New Member

Join Date: Sep 2006
Location: Auckland, New Zealand
Posts: 8777
Good Answers: 376
#3

Re: Isolated Power System in Operation Rooms

05/29/2014 3:48 PM

If the transformer fails then you have no power.

What are the chances the transformer will fail compared to how critical keeping the power on is. You need to do an analysis and if necessary say add an automatic transfer switch or some other means of keeping the power on or getting it back on quickly. This could be as simple as having a spare transformer handy.

__________________
jack of all trades
Register to Reply
Guru

Join Date: Jun 2013
Location: Central Canada
Posts: 677
Good Answers: 28
#4
In reply to #3

Re: Isolated Power System in Operation Rooms

05/29/2014 3:53 PM

I took it as isolation fail, as opposed to total failure.

__________________
Smart as a post and twice as fast.
Register to Reply Off Topic (Score 5)
Guru
New Zealand - Member - Kiwi Popular Science - Weaponology - New Member Engineering Fields - Power Engineering - New Member Engineering Fields - Electrical Engineering - New Member

Join Date: Sep 2006
Location: Auckland, New Zealand
Posts: 8777
Good Answers: 376
#5
In reply to #4

Re: Isolated Power System in Operation Rooms

05/29/2014 4:06 PM

I took it as the transformer is being used as an isolation transformer, and the operation room is actually a site control room.

Your guess is as good as mine until more information is offered!

__________________
jack of all trades
Register to Reply
Guru

Join Date: Jun 2010
Location: srilanka
Posts: 2725
Good Answers: 5
#6

Re: Isolated Power System in Operation Rooms

05/29/2014 10:57 PM

They have earth monitoring system too.

__________________
pnaban
Register to Reply
Anonymous Poster #1
#7
In reply to #6

Re: Isolated Power System in Operation Rooms

05/29/2014 11:08 PM
Register to Reply
Guru
Engineering Fields - Power Engineering - New Member

Join Date: May 2007
Location: NYC metropolitan area.
Posts: 3230
Good Answers: 444
#8

Re: Isolated Power System in Operation Rooms

05/30/2014 12:19 AM

Assuming it's a medical operatory the consequence is no worse than the failure to operate of any piece of electrically powered medical equipment. Everyone in that room is trained to respond quickly and either move the equipment or move the patient, and continue the procedure.

Surprisingly OP considered only one failure mode, fail to provide power, when the more life threatening situation is loss of isolation between patient circuits and power/grounding circuits; that situation could prove to be instantly deadly.

__________________
“Tell me and I forget. Teach me and I remember. Involve me and I learn.” Ben Franklin.
Register to Reply Score 1 for Good Answer
Guru
United States - Member - Born, raised halfway 'round .....

Join Date: May 2010
Location: Metro.Manila, Philippines.
Posts: 1269
Good Answers: 27
#12
In reply to #8

Re: Isolated Power System in Operation Rooms

05/30/2014 2:35 PM

The only reason why an isolation transformer were installed in the O.R. is to ensure that patient were safe from all electrically operated devices. Safe from High leakage current problem generated by every medical equipment used and connected to the patient in the olden days. Minimizing patient's exposure to what is known as micro shock hazards that normally can trigger heart fibrillations.

In the early 90's isolation transformers are no longer a requirement due to changes on modern equipment designs which solved majority of these electric shock hazards. Changes which in turn trigger the corresponding changes or revisions in the applicable codified NFPA 70 standards. Presently all hospitals / healthcare providers are standardized to use "Hospital Grade" U-gnd outlets on all critical or not areas of such facilities.

__________________
vsar
Register to Reply
Guru
United Kingdom - Member - New Member

Join Date: Aug 2010
Location: Resting under the Major Oak
Posts: 4347
Good Answers: 181
#9

Re: Isolated Power System in Operation Rooms

05/30/2014 5:23 AM

Although not applicable to you're location BS7671 corrigendum 2 section 710 is dedicated to medical locations.

It's well worth a read.

__________________
The reports of my death are greatly exaggerated.
Register to Reply
Guru
United States - Member - Born, raised halfway 'round .....

Join Date: May 2010
Location: Metro.Manila, Philippines.
Posts: 1269
Good Answers: 27
#10

Re: Isolated Power System in Operation Rooms

05/30/2014 8:17 AM

This old scenario existed in the 70's - 80's, where operating suites in the hospital are designed to use isolated power supplies. This was necessary to enhance patient safety in the O.R. During those days when situations like this occurs and utility power were interrupted for whatever reasons, the on-going operation procedure is not and cannot be interrupted even to the point of just using flashlights to light the surgical site.

If the medical procedure requires high powered, advanced medical instrumentation, ie., (open heart bypass or transplant), another room will be prepared for case transfer, but the ongoing procedure continues until completed!

The technicians will just have to wait until that room is vacated or until the clinical procedure is completed!

Isolated power supplies in operating rooms are no longer necessary nor required since the year 2000 in the states.

__________________
vsar
Register to Reply
Guru
Safety - Hazmat - New Member Engineering Fields - Retired Engineers / Mentors - New Member Engineering Fields - Piping Design Engineering - New Member Hobbies - DIY Welding - New Member Fans of Old Computers - PDP 11 - New Member

Join Date: Feb 2007
Location: Stronger Than The Storm
Posts: 2394
Good Answers: 203
#11

Re: Isolated Power System in Operation Rooms

05/30/2014 9:00 AM

Many of the modern OR equipment in use today has internal battery backup systems. A well known unit similar to these is the blood pressure monitor used for patients in rooms. Usually the more critical the unit the better the back-up. Other alternatives, especially for ability to see are battery powered back-up lights (similar to emergency lighting), flashlights, larger fire fighters hand lights, and various other methods. some OR's have back-up generators solely supplying them.

During/after Hurricane Sandy many hospitals declared states of emergency so they could readily obtain generators, over several thousands of them. Since no or very little preplanning had been done they were located near the end-user. This then created another problem- Carbon Monoxide levels in excess of the permitted levels in the vicinity of a high generator concentration. Some unique temporary solutions were used but mostly relocation was used.

Good Luck, Old Salt

__________________
Any day on the green side of the grass is a GREAT DAY!, --- me +++++++++. I believe creativity is an inherent part of everyone. --- Kermit T. Frog
Register to Reply
Register to Reply 12 comments

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

Anonymous Poster (1); jack of all trades (2); LongintheTooth (2); old salt (1); pnaban (1); RAMConsult (1); SolarEagle (1); TonyS (1); vsar (2)

Previous in Forum: New Battery for Laptop, Commissioning   Next in Forum: Protection of Transmission Lines by Over Current Relay

Advertisement