Active Power Measurement at Very Low pf (0.003)

06/23/2014 8:13 AM

Hi,

We want to measure active power consumed by highly inductive air core coil - full load current around 177 A, 50Hz, inductance 205 mH.

Since the power factor is very low, different power meters show different readings.

Which Active power meter would you recommend for this application & why ?

Kindly share your experience.

Thanks !

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

Good Answers:

These comments received enough positive votes to make them "good answers".
Guru

Join Date: Jul 2007
Location: Commissariat de Police, Nouvions, occupied France, 1942.
Posts: 2599
Good Answers: 77
#1

Re: active power measurement at very low pf (0.003)

06/23/2014 8:21 AM

I wouldn't. I would use those numbers and to hell with it.

__________________
Good moaning!
Register to Reply
Guru
Engineering Fields - Electrical Engineering - Been there, done that. Engineering Fields - Control Engineering - New Member

Join Date: Dec 2008
Location: Long Island NY
Posts: 15600
Good Answers: 981
#2

Re: active power measurement at very low pf (0.003)

06/23/2014 9:32 AM

The real power consumed by the coil will be turned into heat. I recommend you measure the heat produced by measuring the temperature gradient through a single, dominant thermal conductive path that the thermal conductivity is well known. Make sure to take enough time to see thermal equilibrium.

The best way to validate any measurement technique is to compare with another measurement technique.

I suspect that if you can do an error analysis of your two power meters that you will find your calorimeter measurement will reside in an overlapping uncertainty region of the two power meters.

__________________
"Don't disturb my circles." translation of Archimedes last words
Register to Reply
2
Guru

Join Date: Sep 2009
Posts: 1056
Good Answers: 88
#3

Re: Active Power Measurement at Very Low pf (0.003)

06/23/2014 9:59 AM

A good active power approximation to compare your meters against, is the actual current squared, times the coil's (small) ohmic resistance. Since the coil is coreless and the frequency is low, the only thing that significantly adds to active consumed power, is that. Probably your meters have different internal impedances, that affect the actual current, hence the different readings. I'd go with the meter that shows higher value, but I'd still compare it with the theoretical one, above. S.M.

__________________
Life is complex. It has a real part and an imaginary part.
Register to Reply Good Answer (Score 2)
Guru

Join Date: Apr 2010
Location: About 4000 miles from the center of the earth (+/-100 mi)
Posts: 9910
Good Answers: 1141
#5
In reply to #3

Re: Active Power Measurement at Very Low pf (0.003)

06/23/2014 1:59 PM

As SimpleMind has said, measure the resistance and calculate I squared R. To measure a small resistance accurately, you need to use the 4 wire method where a test current is passed through the resistance and the dropped voltage is measured directly across the unknown resistance.

http://en.wikipedia.org/wiki/Four-terminal_sensing

Register to Reply
2
Power-User

Join Date: Mar 2010
Location: BHOPAL, INDIA
Posts: 201
Good Answers: 20
#6
In reply to #5

Re: Active Power Measurement at Very Low pf (0.003)

06/23/2014 11:22 PM

For more accuracy, measure the resistance at stabilized temperature. This can be done by switching off the power circuit after the unit has attained its stabilized temperature at full load and then measuring the resistance by suitable method depending on the expected value.

Register to Reply Good Answer (Score 2)
Guru
Engineering Fields - Electrical Engineering - New Member

Join Date: Oct 2007
Location: Sour Lake, TX 30°08'59.68"N 94°19'42.81"W
Posts: 675
Good Answers: 13
#4

Re: Active Power Measurement at Very Low pf (0.003)

06/23/2014 1:59 PM

Try the theory:

A 205 mH, at 60Hz, has an inductive impedance of 77.24 ohm

Assuming a voltage of 15000Vac (just a number) the current that you have measured in a load Z= √(R2+XL2) is 177A. I would find Z=84.75ohm and R = 34.9ohm.

The angle between U and I should be

φ = arctan(XL/ R) = 66°

Now you can measure the resistance and calculate directly the angle.

__________________
Bridge rule #1: Nobody is as good as he thinks about himself nor as dumb, as his partner thinks...
Register to Reply
Power-User

Join Date: Jun 2010
Location: Canada
Posts: 139
Good Answers: 6
#7

Re: Active Power Measurement at Very Low pf (0.003)

06/24/2014 3:36 AM

What you are saying is that you have a 2MVA shunt reactor rated at 11kV, 50Hz, and want to measure its losses.

Yes, you would NOT measure the losses in this reactor with any sort of ordinary wattmeter as ordinary wattmeters are very inaccurate near zero power factor. You would need a "low power factor wattmeter" that exhibits very low phase angle errors and is intended for such low power factor measurements. In addition to this, you would need a voltage transformer to step down the voltage - 11kV to 110V, and a current transformer to step the 177A to something the wattmeter can handle, . . like a 200/5 CT. Both the VT & CT need to be very accurate - the ordinary commercial variety used for metering will just not do.

Once you assemble this SPECIAL TEST EQUIPMENT, you are likely to obtain an accuracy of +/- 9% on the power reading. These errors would be attributed to phase errors of +/-1 minute in each of the components - the wattmeter, the CT, and the VT. This one can compare to an accuracy of about +/-300% in the power measurement when using ordinary CT, VT and wattmeter. Yes, in fact the wattmeter may well read negative power.

A more accurate method of measuring the losses of the reactor is to use a high-voltage bridge with a loss-free standard capacitor to scale down the voltage and a current comparator to scale down the current. You may well receive an accuracy of +/-1% in the power measurement at your power factor of 0.003.

Register to Reply
Power-User

Join Date: May 2010
Location: Cambridge, MA
Posts: 390
Good Answers: 82
#8

Re: Active Power Measurement at Very Low pf (0.003)

06/24/2014 5:35 AM

I see several suggestions that you make assumptions about calculations using the inductor's ac resistance. The ac resistance will no doubt be considerably higher than the dc resistance. I'd approach this as a bench measurement, and resonate the coil at about 60Hz with a paralleled 33uF film capacitor. You can drive this from an ordinary bench function generator, e.g., through a 1k or 10k resistor, and monitor with a 'scope. Vary the frequency and measure the Q, to directly determine the AC resistance. Then you can accurately calculate the losses in service at 177A.

__________________
Thanks, Win
Register to Reply
Power-User

Join Date: May 2013
Posts: 351
Good Answers: 22
#9

Re: Active Power Measurement at Very Low pf (0.003)

06/24/2014 7:06 AM

Yes, the ability to measure the real in power in these situations is difficult with normal meters. I was in a factory contact engineering position in a ftransformer factory that built some HV air-core reactors for utilities. (Cast in cement) While your question is for MV application the principal is the same. The resistance is driven as low as possible to try to eliminate in phase losses. Yes, heat loss is one way to cope with measuring losses in these situations. Instruments designed to measure vars only can be used and the in-phase losses calculated by comparing the Volts X Amps to the Vars. We did have instrumentation to read the losses directly, and very accurately, but we had to verify accuracy frequently. Some customers (utilities) wanted to make their own checks, so I remember various excersizes in reading osilloscopes, etc. I was only out of college about 5 years, so my math was still solid, thankfully. Instrumentation to measure this accurately is very expensive, and utility grade or better. It is of a nature that it is permanently installed as a custom installation. Usually it takes the form of a relay actually, which trips if the coil senses a shorted turn.

Register to Reply
Power-User

Join Date: Jun 2010
Location: Canada
Posts: 139
Good Answers: 6
#10
In reply to #9

Re: Active Power Measurement at Very Low pf (0.003)

06/24/2014 7:51 AM

TRENCH Electric is the leader in manufacturing LARGE, air-core reactors. They us a transformer-arm capacitance bridge to measure losses. They have to mount the test reactor in the middle of the parking lot for the measurement, . . . as the losses in the steel components of the test station increase the losses substantially.

Register to Reply
Power-User

Join Date: May 2013
Posts: 351
Good Answers: 22
#14
In reply to #10

Re: Active Power Measurement at Very Low pf (0.003)

06/24/2014 10:19 AM

Yes, Trench is a leader now, but not then. Any steel or conductive loop in the area surrounding the reactor has an effect on it, but there are other ways of shielding them without resorting to large distances.

Register to Reply
Power-User

Join Date: Sep 2006
Location: Houston Texas
Posts: 282
Good Answers: 16
#11

Re: Active Power Measurement at Very Low pf (0.003)

06/24/2014 8:51 AM

Have you tested the Power Cell by Load Controls Inc.?http://www.loadcontrols.com/products/products.html

__________________
Specializing in Dynamic Weighing Systems for Powder and Bulk Solids Handling
Register to Reply
Power-User

Join Date: Nov 2011
Posts: 258
Good Answers: 2
#12

Re: Active Power Measurement at Very Low pf (0.003)

06/24/2014 9:29 AM

At what AC voltage or AC current level are you interested in making power measurement? Check AC rms curernt and pass same rms DC current through the coil and you can use DC power measurement- why use AC at all? You will get (I^2)*R same whether it is AC or DC.

Register to Reply
Guru
Engineering Fields - Electrical Engineering - Been there, done that. Engineering Fields - Control Engineering - New Member

Join Date: Dec 2008
Location: Long Island NY
Posts: 15600
Good Answers: 981
#13
In reply to #12

Re: Active Power Measurement at Very Low pf (0.003)

06/24/2014 9:52 AM

No, the current will be significantly lower with an AC voltage than an identical DC voltage. By definition a very low power factor has significantly more reactive impedance than real impedance.

With over 100 amperes of current I expect that the wire thickness may be large enough for skin effects to change the resistance, too. So a DC resistance measurement may not be an accurate enough resistance measurement at 50HZ.

__________________
"Don't disturb my circles." translation of Archimedes last words
Register to Reply
Power-User

Join Date: Nov 2011
Posts: 258
Good Answers: 2
#16
In reply to #13

Re: Active Power Measurement at Very Low pf (0.003)

06/25/2014 8:57 AM

There are no skin effects at 50 Hz. Hence DC measurements and AC power measurements will not be different. Since he has high inductive load and low resistance and PF is low- surely he is facing probelm with measurements.So I feel DC measurement will be the best.

Register to Reply
Guru
Engineering Fields - Electrical Engineering - Been there, done that. Engineering Fields - Control Engineering - New Member

Join Date: Dec 2008
Location: Long Island NY
Posts: 15600
Good Answers: 981
#17
In reply to #16

Re: Active Power Measurement at Very Low pf (0.003)

06/25/2014 9:16 AM

Really, so you know with absolute certainty that the copper conductor used to handle 177 amperes of current is smaller than the 9 mm skin effect depth for 50 Hz.

Go to my link and do the math. The low resistance and high current implies to me a fairly thick conductor.

__________________
"Don't disturb my circles." translation of Archimedes last words
Register to Reply
Power-User

Join Date: May 2013
Posts: 351
Good Answers: 22
#18
In reply to #16

Re: Active Power Measurement at Very Low pf (0.003)

06/25/2014 3:06 PM

Quote: "There are no skin effects at 50 Hz"

You are wrong. There is definitely skin effect at 50 and 60Hz. Look at the following for values. marc's pages --Skin Effect

The last time I saw any thing printed doubting the Skin effect of 50-60 Hz was published in the 30's. Are all the Utility Companies wrong? I don't think so.

Register to Reply
Power-User

Join Date: Nov 2011
Posts: 258
Good Answers: 2
#19
In reply to #18

Re: Active Power Measurement at Very Low pf (0.003)

06/26/2014 5:00 AM

Thanks & noted. At very small currents it is negligible. But at 177 A that has been mentioned skin effect is significant even at as low as 50 or 60 Hz.

Register to Reply
Power-User

Join Date: Dec 2008
Location: Anthem, AZ
Posts: 392
Good Answers: 8
#15

Re: Active Power Measurement at Very Low pf (0.003)

06/24/2014 10:30 AM

This can be measured at full power by using a modern digital scope with a current probe and a voltage probe.

1. First de-skew the two probes.

2. Calibrate the two probes and acquisition circuits to assure true V and I.

3. Trigger the scope from the zero crossing of the incident voltage.

4. Calculate the power waveform using scope's math.

5. Compute the FFT real and imaginary components.

6. Use a cursor to read off the REAL power (the in-phase component), and the IMAGINARY power (the out of phase component).

Register to Reply
Register to Reply 19 comments