Can H2 be used as carrier gas with TCD detector if H2 is also one of the components to be measured?
If yes how H2 in carries gas doesn't affect H2 in process gas measurement?
I already tried to google the answer and didn't get it.
Be happy :)
Re: Carrier Gas
10/28/2014 10:46 AM
If the gasses are know and only their proportions are being measured, then you should be fine.
If the gasses are unknown, a mass spectrometer in lieu of a katharometer may be indicated.
Re: Carrier Gas
10/28/2014 12:13 PM
The h2 is run absent the process gases as calibration....it is then the comparison that determines specific gas volumes of known process gases....
"It functions by having two parallel tubes both containing gas and heating coils. The gases are examined by comparing the rate of loss of heat from the heating coils into the gas. The coils are arranged in a bridge circuit so that resistance changes due to unequal cooling can be measured. One channel normally holds a reference gas and the mixture to be tested is passed through the other channel."
http://en.wikipedia.org/wiki/Thermal_conductivity_detector
Re: Carrier Gas
10/29/2014 12:06 AM
Hydrogen from your sample can overwhelm your thermister when pure hydrogen of your sample crosses over the thermister. To avoid this use a carrier gas that is approximately 7 % Hydrogen in Helium. You will effectively measure the difference between your carrier gas background hydrogen and the samples' hydrogen as the base line for your chromatograph will include the 7% H2 in the carrier.