If a Radiation Survey meter is callibrated with a low activity (Gamma Source ), can it be extrapolated to measure higher Dose Rate also or does it necessarily needs to be callibrated with Higher Dose Rates (say starting mR/H and extends up to Rad/H.).
How much error can creep in, if done so ?
Good Answers: 918
Re: Radiation Survey Monitors
09/02/2015 8:10 AM
Why haven't you consulted the manufacture?
Good Answers: 26
09/02/2015 8:29 AM
No. If you know your curie strength of the calibration source a very close match can be established as to mr/hr. Higher dose rates will be proportional to the calibration source curies. We use Cs-137 as a cal source and it covers the emission of all our suspect contamination's.
X-ray use Co-60, radioactive sources use Cs-137.
Good Answers: 918
09/02/2015 11:10 AM
True, but will the sensor respond that way? It might saturate the sensor. You don't know.
Contact the manufacture, but why would he listen to them when you get a whole spectrum of conflicting opinions from the internet?
Good Answers: 26
09/02/2015 2:53 PM
I established the procedure that if saturation is questioned then back away from the suspect slab , slowly, until the meter starts to read (inverse square law) then evaluate from a distance. This has not happened , yet, but we did find some old cooler units that were hot. Still don't know what the radioactive element was.
Good Answers: 918
09/02/2015 3:48 PM
That doesn't tell you what the sensor's linear range is.
A telephone call probably would.
However, if the device is specified to a limit there just may be a reason for that published limit.
Good Answers: 26
09/03/2015 7:18 AM
We establish the linear range from the Cs-137 by movement of the GM tube using a 24" caliper. This is performed monthly and records kept. Really don't need to make a call, i have been doing this for over 40 years and my procedures work.
When i was younger it was part of my job to retrieve disconnected sources, mainly Ir 192 and Co 60. I have yet to have or had an overexposure. My current life exposure is slightly over 23 Rems, which considering the above is a good record.
09/03/2015 9:48 AM
If the monitor is saturated, then all you know is 'off the scale,' but not how FAR off the scale.
If I had a radiation monitor that was saturated, I would be backing away QUICKLY, if not turning and running until I could be sure I was out of any potential 'death zones.'
09/03/2015 9:57 AM
At least here you demonstrate exceptionally good common sense!
09/03/2015 10:43 AM
Despite our political differences, I'll take that as a compliment.
I am also looking forward, as I suspect you are, to mid-November 2016, when all the political posturing will be over, so the news isn't filled with all the 'hot issue' topics we tend to disagree on. It's nice to relax and just 'talk shop' without politics or religion getting in the way.
(With luck they won't start campaigning for the 2020 elections until at least after the 2018 mid-terms, but with the way things seem to be going, we might be looking at the start of constant 'election fever' in this country. the 2016 elections are still over a year away, these are the primaries. Nobody's supposed to CARE about the primaries, it's just a bunch of yobbos prattling on about stuff they'll forget all about once they get the party nomination, *IF* they get the party nomination. If they don't win the primary, they'll just slink back into the shadows and not be heard from again.)
09/03/2015 10:59 AM
Ah contraire, mon ami! It never actually ends any more. This "stuff" (insert your own adjective here, expletives included) goes on ad nauseum until everyone is sick of it. I am ignoring spell checker, since it ignores me. (All of my current rants will be about spell checker on CR4 until they fix it) Does spell check not work on Windoze 8? According to spell checker, CR4 is a misspelled word.
It would be nice if the candidates had to prove themselves in a sort of "iron-man" competition beginning with serious tongue lashings from the public, then proceeding with a 20 mile marathon (where they actually have to "run" for office), not saying how fast they have to run. Next in the line-up would be hand-to-hand combat using the Obamacare documents for a weapon. Obviously, here, the "gentleman" with the higher ground will indeed have the higher ground, based on the weightiness of the matter at hand. Finally, the survivors (if there are any) will proceed to the convention of their party and be voted up or down, or off the island. After that, they will dress up as Lucille Ball in the cupcake factory, and see what good they really are. Then comes the general election. Candidates to be placed on adjacent (hopefully) nearby rock spires (natural formations of course). They have to be able to listen, but yell to be heard. Which ever one yells the best is allowed the long rope to repel down to the bottom. They guy with the short rope, oh well. If neither candidate survives the ordeal, then the runners-up (if they are living) would be allowed an alternate quest to achieve greatness, such as allowing the populace to actually vote.
09/03/2015 11:58 AM
That's actually more entertaining than my usual idea, a breath-holding competition at the bottom of lake Michigan, with all contestants wearing weight belts padlocked around their waists, to prevent the bodies from floating back up.
Aside from His Regal Hairpieceness(1), they're all 'career politicians,' which means they started out as lawyers. So yes, I *AM* saying they can ALL (both parties) go jump in the lake for all I care.
Notes:
  1. Yes, I know that's The Donald's real hair, I just need to pretend it's a bad toupee to protect my sanity. That hair just looks wrong, WRONG, like staring-at-a-five-sided-triangle wrong. Every fiber of my being screams that in a universe with sane, consistant laws, such a thing CAN NOT BE! That hair is the C'thulu of hairstyles.
09/03/2015 4:28 PM
The best answer I ever heard from The Donald about his hair went something like this...
"Of course this is my real hair. With my money do you think I'd pay for a toupee that looked this bad?"
09/04/2015 10:12 AM
Okay, the man an make self-effacing comments, he's actually GOT a sense of humor. I'm a little more impressed with him. Don't find much to like about him, but I can still afford him the same basic respect due all humans.
Although, with his money, don't you think he could afford a better barber? The only heads of state that can get away with bad haircuts are generally the really old ones, who usually go for the Einstein hair look, and the ones from regions where hair is generally wilder and out of control. It would not be amiss for an Arabic leader to have totally INSANE hair under his turban, if he is from one of the sects that do not believe in cutting the hair off the head And the Jamacan president would not be considered unusual for having dreadlocks that reached his knees, for the same reasons, Rastafarians also consider it inappropriate to cut the hair off the head.
For white, male, Catholic or Prodistant world leaders who are younger than 80, conservative, fashionable haircuts are the order of the day, the JFK hairstyle if you can manage it (that man had Perfect Hair), and if you're starting to go bald, the Patrick Stewart close-cropped. (Hair? Who needs hair when you're the best diplomat to ever command a starship. Leave the hair to the XO, in fact, let him have a double dose, he can use it for a beard or something.)
09/04/2015 2:45 PM
You do bring up an interesting point. Would any barber or hair stylist freely admit to cutting that hair?
09/04/2015 3:36 PM
"Prodistant"? Really? Does that mean "like you from a distance"?
Did you mean Protestant? As in Martin Luther? Not Martin Luther King, although, he too was a Protestant...
09/04/2015 4:19 PM
If Martin Luther was Prodistant, would he have used a crossbow at 100 yards to put his 95 treatises on that door?
09/10/2015 3:59 PM
Something like that, but the if Pope back then had only been a bit more agreeable, there might not be any Lutherans, or Episcopalians.
09/10/2015 9:01 PM
An interesting premise but I think it is more likely the Reformation would've just happened at a different time. It is just as likely an earlier or later Pope with the same pious, "holier than thou", attitude could just as easily trigger the same result just at a different time.
09/11/2015 9:40 AM
Agreed. Done. History.
Good Answers: 26
09/03/2015 10:55 AM
There is a safety issue involved. As the radiation monitoring is being done in a steel slab yard by moving slower and going behind a stack of slabs the tech is shielded(time,distance shielding). Running in a slab yard is not safe and the tiny amount of radiation that would be received is allowable rather than slip, trip,fall. These techs are radiation trained and allowed up to 5R/yr. But still the minimum is the goal.
09/03/2015 11:10 AM
Do the radiation techs carry scissors while running? Walk, don't run. Hey good on you for the work you do, stay safe out there.
09/03/2015 12:06 PM
The setting was not stated, so I took the most logical result to finding out mid-job that I am in an area with more radiation than I can measure: "Screw this, I'm not paid to die, I'm outa here."
With steel slabs to move behind for protection, my response would be much different:
  1. Confirm from radiation monitor than where I am is safe.
  2. Extend sensor end of radiation monitor past side of shield, confirm path to next shield is safe.
    1. If safe, move to safe zone behind next slab.
    2. If reading are not safe, or off the chart, do NOT enter area and report findings to supervisor.
  3. Repeat until all measurements are taken, or until monitor cannot provide accurate readings.
09/02/2015 9:17 AM
Yes, one can extrapolate higher dose rates if one knows a lot more information. Depending on the nature of the unknown radiation sensor the reference rate and the higher rate may both reside in a linear range of the detector. In this case a simple linear extrapolation can be performed. If the higher rate causes the sensor or the electronics reading the sensor to non-linearly saturate then one cannot extrapolate an accurate number.
Then there's the different measurements of ionizing radiation that gets very confusing to many laymen.
Good Answers: 122
09/02/2015 10:23 AM
Meter accuracy is optimized if the readings taken lie within 80% of maximum scale. Based on this I would not use the meter to measure anything out of it's calibrated range.
09/11/2015 9:46 AM
Generally that is a strongly recommended practice for nearly any measurement.
09/02/2015 2:41 PM
This I don't think can be generally stated as it would depend on the unit itself....but generally speaking, no....
http://122.physics.ucdavis.edu/sites/default/files/files/Gamma%20Spec/Guide.pdf
https://en.wikipedia.org/wiki/Survey_meter
Good Answers: 153
09/03/2015 4:42 AM
A general rule for any calibrated sensor/instrument is that you calibrate as near to the full scale reading as possible. There are two aspects to calibration, zero and span. For the zero reading you need to know your local background radiation figure. To get a true zero, cancel out the background radiation by setting the zero on your instrument at minus the background radiation figure. Assuming the scale is linear, a 1% calibration error when setting the span at 5% of full scale will produce progressively higher errors further up the scale reaching approximately a 20% error at full scale. But a 1% calibration error when setting the span at 95% of full scale produces progressively lower errors as you move down the scale reaching approximately 0.02% at zero reading. Your span calibration error, as described in detail above has to be added to your zero calibration error to arrive at your total calibration error. This may also need to be amended for temperature variations.
09/03/2015 9:39 AM
Thank you and the responding bloggers for this thread. I found the units link most helpful, along with the discussion. There are indeed certain principles of measuring any radiation (including light), that depend on known facts about the source of the radiation.
(1) Is the source considered a point source, distributed source, or coherent source? It makes a world of difference...
(2) What is the known linear response range of the detection instrument for the source at hand? The source must not be far different in energy (wavelength) from any calibration source. The detector if used in particle (or photon) counting mode must truly be set up for that (obviously), and has to have a fairly sophisticated discrimination circuit built-in as part of detection of bona-fide events.
(3) How does the intensity of the detected source change with distance upon approach?
(4) If the source is radioactivity, can discrimination between alpha, beta, and gamma radiation be readily made by the operator? Hint: if the detector could possibly detect all three, there are simple tests with attenuating materials, such as paper for beta, aluminum foil for alpha, and lead for gamma. What to do if the source is not ionized as such or ionizing, but represents perhaps neutrons, or other particles (will leave this to someone's imagination)? Obviously then one must begin to delve into the black arts of neutron detection, etc.
Good Answers: 26
09/11/2015 5:37 AM
In answer to #4.
http://www.goldmine-elec-products.com/prodinfo.asp?number=G19153