Issues in Measuring Power Factor in Energy Meters

02/18/2016 10:49 AM

Dear Community Members.

Myself Ankit Jain.

I am heading a Finance Department of a Testing/Calibration Laboratory in GUjarat, India.

Over here we are facing a very typical issue with regards to the measurment of Power Factor in the Energy Meter.

When we are connecting DOL type motor in our setup the energy meter shows the current & Voltage accurately but the power Factor displayed is inaccurately displayed at two different meters. From Last 1 month we have tried everything from changing the wiring to changing the CT Coils and measuring the pf using a independent power factor meter as well. The energy meter we are using are of standard company i.e. Schnieder.

Unable to identify the problem. Can anyone help us on this issue as due to this issue our centre is stuck in a dead lock situation.

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#1

Re: Issues in measuring POwer Factor in Energy Meters

02/18/2016 11:19 AM

The word <...inaccurately...> requires more meaning, and some numbers would help.

Please describe the outcome of the telephone discussions that have taken place with the equipment manufacturer <...i.e. Schneider...> on this topic.

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
Guru
Engineering Fields - Power Engineering - New Member

Join Date: May 2007
Location: NYC metropolitan area.
Posts: 3230
Good Answers: 444
#2

Re: Issues in measuring POwer Factor in Energy Meters

02/18/2016 12:25 PM

I assume that your metering and the utility billing meter (Schneider 8650) don't agree during your motor starting tests. This normal for two reasons; first, DOL starting is a highly transient situation, second, unless you are using exactly the same meter as the utility, the sampling rate differences will affect the real-time calculation of the PF. This will be especially noticeable if one meter is digital and the other is analog.

Hire an experienced electrical engineer to install a modern digital Power Quality Meter to record what's really happening during these transients.

__________________
“Tell me and I forget. Teach me and I remember. Involve me and I learn.” Ben Franklin.
Register to Reply
Guru

Join Date: Jan 2006
Location: Tamworth, UK.
Posts: 1782
Good Answers: 45
#12
In reply to #2

Re: Issues in measuring POwer Factor in Energy Meters

02/19/2016 7:24 PM

As a novice in the subject of analog and digital meter design I am curious how the PF is actually measured or calculated or displayed. I can guess but I don't know.

But I would have thought the results would be the same if using meters made by reputable manufacturers.

Or put another way, if I bought a PF meter I would expect it to give the correct reading.....whatever make or design concept used.

.....but judging from the posts it seems not.....

Knowing the correct PF is not on my list of things that matter to me at the moment.... Just curious.

__________________
When arguing, remember mud-slinging = lost ground.
Register to Reply
Guru
Engineering Fields - Power Engineering - New Member

Join Date: May 2007
Location: NYC metropolitan area.
Posts: 3230
Good Answers: 444
#13
In reply to #12

Re: Issues in measuring POwer Factor in Energy Meters

02/20/2016 1:25 AM

Don't judge by the posts, just get a copy of the ANSI C12 series or IEC 62503-21 standards, they describe all the testing and accuracy criteria for revenue metering accuracy.

__________________
“Tell me and I forget. Teach me and I remember. Involve me and I learn.” Ben Franklin.
Register to Reply
Guru

Join Date: Jan 2006
Location: Tamworth, UK.
Posts: 1782
Good Answers: 45
#16
In reply to #13

Re: Issues in measuring POwer Factor in Energy Meters

02/21/2016 6:43 AM

" Don't judge by the posts, just get a copy of the ANSI C12 series or IEC 62503-21 standards, they describe all the testing and accuracy criteria for revenue metering accuracy."

Thank you RAM for this, but I doubt if a copy of these standards will answer the question about the different results the OP is faced with.

If these standards have any force, then manufacturers would be obliged to comply with them. In which case it would be reasonable for the user to expect the same results. Which is not the case here it seems.

If meters are the 'same' then it must come down to methodology - ie, how and where is the meter connected to measure the PF.

eg. if the meter is connected to measure the PF in the final leg of the motor circuit, and then compared to a reading from a meter connected at the incoming mains, probable with a time delay between readings, it is perhaps not surprising there is a difference.

Just a thought.

__________________
When arguing, remember mud-slinging = lost ground.
Register to Reply
Guru
Engineering Fields - Power Engineering - New Member

Join Date: May 2007
Location: NYC metropolitan area.
Posts: 3230
Good Answers: 444
#17
In reply to #16

Re: Issues in measuring POwer Factor in Energy Meters

02/21/2016 10:01 AM

Suggest you read post 15 below. Until OP has proved that both the utility and his metering read exactly the same under identical conditions (which is the test methodology in the standards), then he is doomed to failure. Steady state measurements will usually be close, but transients and nonlinear loads (both of which occur during motor starting) can cause great differences in readings.

When I started doing power testing everything was analog. Along came digital metering and my the testing personnel started reporting results that differed from the previous tests. All it took was some concurrent testing (current inputs in series, voltage inputs in parallel), coupled with oscilloscopes to monitor the various waveforms and a whole bunch of metering theory research, to convince them why the readings varied so drastically. nb- all our instrumentation was traceable to the National Bureau of Standards and calibrated periodically.

__________________
“Tell me and I forget. Teach me and I remember. Involve me and I learn.” Ben Franklin.
Register to Reply
Guru

Join Date: Oct 2008
Posts: 42355
Good Answers: 1693
#3

Re: Issues in Measuring Power Factor in Energy Meters

02/18/2016 12:37 PM

Get one of these and find someone who can use it.

Fluke 1730 Three Phase Electrical Energy Logger Measures voltage, current, power and power factor, comprehensive logging, optimized user interface, complete in-the-field setup

Part NumberFLUKE-1730/US
Base Price$2,429.99
Current ProbesWith 1500A Flexi Clamp 12in current probes (+$270.00)
Configured Price $2,999.99 $2,699.99
Starting at $2,699.99
Register to Reply
Power-User

Join Date: Nov 2011
Posts: 258
Good Answers: 2
#4

Re: Issues in Measuring Power Factor in Energy Meters

02/18/2016 10:54 PM

Mr Ankit Jain,

DO NOT ASSUME that standard company "SCHNIEDER" does not mean the algorithm used to compute power factor is standard. There are different definitions of power factor and different methods of measurement. So you may find differences between meters as each may be using internally different techniques. It is particularly so , if the wave forms are distorted. Don't ask engineers and college students- they only know textbook definitions, but not real life definitions..

Register to Reply Score 2 for Off Topic
Guru
Engineering Fields - Power Engineering - New Member

Join Date: May 2007
Location: NYC metropolitan area.
Posts: 3230
Good Answers: 444
#5
In reply to #4

Re: Issues in Measuring Power Factor in Energy Meters

02/18/2016 11:35 PM

Quite the opposite, utilities purchase meters from various sources and test them to international standards to ensure that they all read the same statistically, despite internal differences in circuitry, algorithms, etc.

Consumers aren't held to the same standard so that it is more probable that the error is on their side. The only way that the OP should have an issue is if he is using exactly the meter type, CTs and PTs, as the utility does and the readings are different then the standards would allow.

__________________
“Tell me and I forget. Teach me and I remember. Involve me and I learn.” Ben Franklin.
Register to Reply
Guru
Canada - Member - Specialized in power electronics

Join Date: Feb 2008
Location: Montreal, Canada.
Posts: 1372
Good Answers: 80
#6

Re: Issues in Measuring Power Factor in Energy Meters

02/19/2016 5:47 AM

Could it be that one meter gives the displacement PF while tho other gives the overall PF (displacement and distortion)? With a perfectly sinusoidal current and voltage, they are both the same but they start diverging when distortion is present.

__________________
Experienced is earned, common sense is taught, both are rare essentials of life.
Register to Reply
Participant

Join Date: Feb 2016
Posts: 4
#7

Re: Issues in Measuring Power Factor in Energy Meters

02/19/2016 6:56 AM

Thanks for all the responses but let me first explain the situation in detail.

As per the diagram when i am supplying the power to submersible pumpset the Energy Meter installed at the power panel for Single phase is showing correct current values and Voltage values as verified with the Clamp on meter. But the power factor sensed is 0.99 due to which the efficiency of the Pump Set decreased like anything. Aain a pf of .99 is highly impossible scenario as the Pump Manufacturer himself claims pf of 0.8 as the max. This one issue similar issues are observed at other power supplies as well but this one needs to sorted out first for us.

Register to Reply
Guru
Safety - Hazmat - New Member United States - US - Statue of Liberty - New Member Engineering Fields - Chemical Engineering - Old Hand

Join Date: Mar 2011
Location: Lubbock, Texas
Posts: 14331
Good Answers: 162
#8
In reply to #7

Re: Issues in Measuring Power Factor in Energy Meters

02/19/2016 8:52 AM

"As per the diagram when i am supplying the power to submersible pumpset the Energy Meter installed at the power panel for Single phase is showing correct current values and Voltage values as verified with the Clamp on meter. But the power factor sensed is 0.99 due to which the efficiency of the Pump Set decreased like anything." --But are you testing a single phase pump motor when looking at the single phase reading?

Usually the fight in the marketplace takes place over apples being compared to oranges. Although another usually results when one person's apple is another persons applesauce.

by the way, labels are for soup cans, placards are for motors.

__________________
If it ain't broke, don't fix it. Just build a better one.
Register to Reply
Participant

Join Date: Feb 2016
Posts: 4
#9
In reply to #8

Re: Issues in Measuring Power Factor in Energy Meters

02/19/2016 10:12 AM

Yes Sir. We are very looking at the single phase reading while testing a single phase motor. Anyways thanks a lot for your valuable inputs.

Register to Reply
Guru
Safety - Hazmat - New Member United States - US - Statue of Liberty - New Member Engineering Fields - Chemical Engineering - Old Hand

Join Date: Mar 2011
Location: Lubbock, Texas
Posts: 14331
Good Answers: 162
#10
In reply to #7

Re: Issues in Measuring Power Factor in Energy Meters

02/19/2016 11:01 AM

Are you sure this is the maximum power factor, because 0.8 pf sounds more like a power factor at maximum LOAD? When a motor (including single phase) is unloaded and just spinning, I think you can rest assured the power factor is much nearer 1.0 than it is 0.9.

__________________
If it ain't broke, don't fix it. Just build a better one.
Register to Reply
Guru

Join Date: Jan 2006
Location: Tamworth, UK.
Posts: 1782
Good Answers: 45
#11
In reply to #10

Re: Issues in Measuring Power Factor in Energy Meters

02/19/2016 6:52 PM

"... a motor (including single phase) is unloaded........ the power factor is much nearer 1.0 than it is 0.9...."

Are you sure about this ? I think you will find the PF is very low for motors running off-load.

__________________
When arguing, remember mud-slinging = lost ground.
Register to Reply
Participant

Join Date: Feb 2016
Posts: 4
#14
In reply to #11

Re: Issues in Measuring Power Factor in Energy Meters

02/20/2016 2:06 AM

Yes the pf for motor at No Load should be quite low. But herein the motor is working at rated load as the submersible pump set is submerged in the water tank and throwing pumping the water. Thus it is very much loaded. But still a pf of 0.99 is not possible.

Can improper grounding be a reason for the same or something be acting as capacitive load which leads to high power factor.

Can anybody please highlight on the same.

Register to Reply
Guru
Engineering Fields - Power Engineering - New Member

Join Date: May 2007
Location: NYC metropolitan area.
Posts: 3230
Good Answers: 444
#15
In reply to #14

Re: Issues in Measuring Power Factor in Energy Meters

02/21/2016 12:56 AM

Until you verify that your metering setup is identical to the utility's you're wasting your time looking elsewhere.

__________________
“Tell me and I forget. Teach me and I remember. Involve me and I learn.” Ben Franklin.
Register to Reply
Guru
Safety - Hazmat - New Member United States - US - Statue of Liberty - New Member Engineering Fields - Chemical Engineering - Old Hand

Join Date: Mar 2011
Location: Lubbock, Texas
Posts: 14331
Good Answers: 162
#18
In reply to #11

Re: Issues in Measuring Power Factor in Energy Meters

02/22/2016 9:51 AM

OK, I had it all backwards: PF = Re(EI)/Total (EI)

Apparent power = Total power (both real and imaginary components). At no load, there is apparent power without much real power at all, and PF could approach 0, (not necessarily = 0).

At full load, the PF is higher by quite a lot, but not as high as 0.99, unless over compensated by some capacitance, i.e. another no load motor/generator with over excitation, or some other stray capacitance.

__________________
If it ain't broke, don't fix it. Just build a better one.
Register to Reply
Guru

Join Date: Jan 2006
Location: Tamworth, UK.
Posts: 1782
Good Answers: 45
#19
In reply to #18

Re: Issues in Measuring Power Factor in Energy Meters

02/22/2016 10:37 AM

Fine. To my mind's eye I 'see' power factor as vectors of the phase angle between peak current and peak voltage. With a motor being highly magnetic reactive the current always lagging the voltage - more on light load than on high load.

__________________
When arguing, remember mud-slinging = lost ground.