Fluke 435 II Power Quality Analyzer

11/07/2018 11:00 AM

Good day

Could any one tell for how long (maximum time) FLUKE 435 with 8GB SD card can record data before downloading.

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

Good Answers:

These comments received enough positive votes to make them "good answers".
Guru
Engineering Fields - Electrical Engineering - Been there, done that. Engineering Fields - Control Engineering - New Member

Join Date: Dec 2008
Location: Long Island NY
Posts: 15602
Good Answers: 982
#1

Re: Fluke 435 II Power Quality Analyzer

11/07/2018 12:07 PM

I suspect this all depends on how one configures the instrument to capture data. Configuring it to capture a brief time interval (30 sec?) of "good" power quality once but then the same time interval before and after a preset anomaly trigger level will allow this to be monitoring until its batteries die.

__________________
"Don't disturb my circles." translation of Archimedes last words
Register to Reply
Guru

Join Date: Oct 2008
Posts: 42355
Good Answers: 1693
#2

Re: Fluke 435 II Power Quality Analyzer

11/07/2018 12:13 PM

What redfred said. Sample rate must be considered, size of data file, etc, etc.

Too many vairables to say.

Get a 32 GB sd card.

Register to Reply
2
Guru
Hobbies - DIY Welding - Wannabeabettawelda

Join Date: May 2007
Location: Annapolis, Maryland
Posts: 7940
Good Answers: 458
#3

Re: Fluke 435 II Power Quality Analyzer

11/07/2018 1:18 PM

Depends on several things. Are you planning on capturing wave events/transients that exceed certain limits? Those capture events are at 200 kS/s. You are able to capture up to 9,999 waveform events. Depending on how you set your limits, you may get a lot or very few waveform captures.

Logging data of course uses a lot less memory. And that will depend on how you set your averaging time. The default is a 1 second averaging time. The range of logging intervals is 0.25s to 2 hours. If you log data every 2 hours, you will be dead long before you fill up an 8 gig card.

The best thing to do, is to set up the instrument with your configuration and let it run for an hour. That will give you a yardstick to estimate the duration that an 8G card will collect data.

Register to Reply Good Answer (Score 2)
2
Guru

Join Date: Mar 2007
Location: by the beach in Florida
Posts: 33392
Good Answers: 1817
#4

Re: Fluke 435 II Power Quality Analyzer

11/07/2018 3:29 PM

It tells you on the start page menu how much time you have according to the sample rate you've chosen....

__________________
All living things seek to control their own destiny....this is the purpose of life
Register to Reply Good Answer (Score 2)
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#5

Re: Fluke 435 II Power Quality Analyzer

11/12/2018 12:27 PM

Yes. The equipment manufacturer will be happy to help over one of those under-utilised facilities these days - the telephone call.

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
Register to Reply 5 comments

Good Answers:

These comments received enough positive votes to make them "good answers".
Copy to Clipboard

Users who posted comments:

Brave Sir Robin (1); lyn (1); PWSlack (1); redfred (1); SolarEagle (1)

Previous in Forum: 4- 20 mA Transmitter Signal Greater Than 20 mA   Next in Forum: A Problem Related to WinCC

Advertisement