Separate Earthing Requirements for Instruments

11/06/2007 11:47 PM

Some Instrument manufacturers calls for saparate earthing for all instruments.

They don't want the Electrical system earthing connected to their instruments.

Is there really any harm to the Instrument if the instrument earthing is taken from the Electrical System Earthing?????

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

Good Answers:

These comments received enough positive votes to make them "good answers".
Associate

Join Date: Sep 2007
Posts: 49
Good Answers: 1
#1

Re: Separate Earthing Requirement for Instruments???

11/07/2007 12:21 AM

Isolated or separate earthing system for instrumentation is not recommended in many codes and standards. Electrical Earthing System and Instrumentation Earthing System finally should be connected to each other. You can find more about this matter in IEC 61000-5-2 1997 and also IEEE Std 142 1991.

__________________
A thousand-mile journey begins with the first step.
Register to Reply
3
Power-User

Join Date: Nov 2007
Location: A secret. Bats are a Protected Species.
Posts: 186
Good Answers: 4
#2

Re: Saparate Earthing Requirement for Instruments???

11/07/2007 8:52 AM

It won't 'harm' the instrument though it may affect the readings you get from it at times, depending on what it is. Depending on the electromagnetism in your area, it is possible to set up earth loop currents and eddy currents (look these up - Google? Wiki?) in the earth conductors and these can affect instrument readings. Remember the old analogue gramaphone pickups that play vinyl and bakelite records? Disconnect an earth screen from an electromagnetic pickup and you may find that the frequency of the mains supply gets amplified and comes out of your speakers. Horrible! The same can happen to instruments at times, and you don't want plant control systems to be swamped by AC mains noise if it can be avoided.

So the best practice is to bond all the power-consuming cables' earth conductors to one earth bar, all the instrument earths to a separate earth bar, and connect the two together as close to the building earth mat as is reasonably practicable. Don't mix 'em up.

__________________
Innuendo is not Italian for a suppository.
Register to Reply Good Answer (Score 3)
Guru

Join Date: Aug 2007
Location: Indiana, USA
Posts: 579
Good Answers: 61
#3

Re: Saparate Earthing Requirement for Instruments???

11/07/2007 8:53 AM

In many installations, the service power ground is not well maintained. I have found this to be especially true in manufacturing facilities, where the process machinery is the focus of all attention from the maintenance staff. Another potential problem is that, in large facilities, the service ground may be a hundred or more meters from the instrument, and the ground wire from the instrument (which is usually fairly small) will have significant resistance due to the length.

To ensure their instrumentation is not affected, some instrument manufacturers want a separate, local, earth connection for their devices. However, if you have a good grounding system throughout the entire installation, you won't have any problems. The standards mentioned in the earlier post both include information on proper design and maintenance of the overall grounding system.

__________________
Experience: The knowledge you gain just AFTER you needed it.
Register to Reply
Guru

Join Date: Sep 2007
Location: Wrong end of the yellow brick road in Oz
Posts: 930
Good Answers: 15
#4
In reply to #3

Re: Saparate Earthing Requirement for Instruments???

11/07/2007 11:51 PM

I have lit a incandescent light bulb across Neutral and Earth at 1 installation that was having problems with the stability of one of the motor drives my last company sold them..

__________________
Qn, Whats the differance between a Snake and a Onion? Ans, No one cries when you chop up a Snake
Register to Reply Off Topic (Score 5)
Active Contributor

Join Date: Aug 2007
Location: Oakville, Ontario Canada
Posts: 14
#6
In reply to #3

Re: Saparate Earthing Requirement for Instruments???

11/08/2007 7:48 AM

Amen to that. In our facility we had a terrible time with RTD's - the readings would shift 5-6 degrees when equipment was turned on and off. Ultimately, the culprit was was poor grounding systems. There was as much as 23V difference in potential between a wind tunnel running 460V and the ground of my temperature sensors - connected to a 110V separate source (we have 3 primary power feeds into the building). Installing a separate grounding rod to bring the two sources in question to the same point didn't help. In the end, running a new control transformer from the wind tunnel power feed brought them to the same level and eliminated the ground loop errors. I'm sure that the system can be improved, but the point is be wary of tying all ground systems together when you have poor ground quality to begin with. Good Luck, AndyC.

Register to Reply
Guru

Join Date: Aug 2005
Location: Hemel Hempstead, UK
Posts: 5826
Good Answers: 322
#5

Re: Separate Earthing Requirements for Instruments

11/08/2007 3:45 AM

It depends what you're doing. If the equipment you're trying to make the measurements on is earthed to "dirty" mains you may well be better off earthing your instrument to the same point. I'm not familiar with other "instrumentation" but if you try to earth an oscilloscope in a "bench" socket whilst the equipment your monitoring is powered from elsewhere your readings can be swamped with noise (even with a short earth connection on the probe).

__________________
If you spend all your time looking for people and things to complain about: trust me, you will find plenty to complain about.
Register to Reply
Active Contributor

Join Date: Mar 2007
Location: Shawnee, Kansas, USA
Posts: 20
Good Answers: 1
#7

Re: Separate Earthing Requirements for Instruments

11/08/2007 8:25 AM

We just this past month moved our operations to a new plant. We have a Mitutoyo CMM that was among the items moved. When we had our CMM calibrated this week, the Mitutoyo Calibration Tech recommended that we establish a separate ground to "earth" to the controller. There is a stud provided for that purpose. This is a protection to the controller in case of lightning strike, which can fry the controller. If the controller is damaged during a lightning strike, and the controller is not grounded to "earth" Mitutoyo will not warrant the device. The controller is worth mucho $$$!

We are going to "earth" this puppy!

Register to Reply
Participant

Join Date: Nov 2007
Posts: 2
#8
In reply to #7

Re: Separate Earthing Requirements for Instruments

11/12/2007 11:15 PM

I also feel that separate earthing is required for instruments. Generally all the AC/DC machines or heating appliances will generate the noise in the form of emmitted radiations (electromagnetic) and conducting radiations (noise on the power lines). The emmitted radiations captured by the body of machines/appliances and inducts the colleted noise (electrical form) on the common bus bar ground. This reverse ground noise affects the readings of the ground sensitive measurement devices.

Bye

Shivashankar Deginal

Register to Reply
Participant

Join Date: Nov 2007
Posts: 2
#9

Re: Separate Earthing Requirements for Instruments

11/12/2007 11:18 PM

I also feel that separate earthing is required for instruments. Generally all the AC/DC machines or heating appliances will generate the noise in the form of emmitted radiations (electromagnetic) and conducting radiations (noise on the power lines). The emmitted radiations captured by the body of machines/appliances and inducts the colleted noise (electrical form) on the common bus bar ground. This reverse ground noise affects the readings of the ground sensitive measurement devices.
Bye
Shivashankar Deginal

Register to Reply
Anonymous Poster
#10

Re: Separate Earthing Requirements for Instruments

12/13/2009 2:15 AM

plz calll me

Register to Reply Off Topic (Score 10)
Guru
Engineering Fields - Mechanical Engineering - New Member

Join Date: May 2008
Location: CHENNAI, TAMIL NADU, INDIA.
Posts: 1851
Good Answers: 65
#11

Re: Separate Earthing Requirements for Instruments

01/29/2017 3:08 AM

Dear Mr.Prasanna1283,

It is correct to have a seperate/dedicated earthing for instruments.

In the power circuit earthing there will be always some unbalanced current flow which will be significant and will affect the Instruments.

Inductance effect will be felt in the power circuit earthing which will have impact on the instrument earthing.

DHAYANANDHAN.S

Register to Reply
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#13
In reply to #11

Re: Separate Earthing Requirements for Instruments

04/09/2025 5:04 AM

<...In the power circuit earthing there will be always some unbalanced current flow which will be significant and will affect the Instruments...>

The purpose of the earth conductor is to operate the circuit protective device(s) in the event of a fault, thereby preserving life and equipment. So, that sentence doesn't make sense: ordinarily no current flows in a properly-installed earth conductor.

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#12

Re: Separate Earthing Requirements for Instruments

07/24/2024 9:36 AM

It is done to preserve the <... Instrument manufacturers ...>' reputations as most instruments cannot survive a fault current of a substantial number of amperes flowing into them.

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
Register to Reply 13 comments

Good Answers:

These comments received enough positive votes to make them "good answers".
Copy to Clipboard

Users who posted comments:

Abbas (1); AndyC (1); Anonymous Poster (1); dhayanandhan (1); drdave (1); Horrible Old Bat (1); pwr2thepeople (1); PWSlack (2); Randall (1); shivashankar (2); Snaketails (1)

Previous in Forum: Diesel Generator Problem   Next in Forum: Bus Coupler Scheme in 6.6 KV Breakers

Advertisement