gents,

Is there anyone who has experince with ABB pressduct loadcell. I am experiencing problem with a weighing system using ABB press duct. I am experiencing some spike in weight reading. When I trended excitation voltage, current and loadcell output voltage, I get a spike in excitation voltage and output voltage(corresponding to weight spikes), but current remains stable.

Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.
Commentator

Join Date: Sep 2007
Location: Here, there and everywhere...
Posts: 61
#1

Re: ABB Loadcell

01/09/2008 6:32 PM

'Been a while since I was involved in load cells of the strain guage Wheatstone bridge type (dunno about the ABB ones), and I think the most modern ones have their own signal conditioning in the cell. However my suggestion (for the old type) is that is your current and ammeter of the same digital sampling rate? You may not be seeing an intermittant spike in your current measurement, and I would tend to use an anologue ammeter for that - in case your not.

Sorry I can't be more helpful.

__________________
An EXPERT is a 'has been', that is a drip under pressure - My Boss
Reply
Guru
Philippines - Member - New Member Engineering Fields - Instrumentation Engineering - New Member Engineering Fields - Control Engineering - Who am I?

Join Date: Oct 2006
Location: Northern Mindanao, Philippines
Posts: 2147
Good Answers: 53
#2

Re: ABB Loadcell

01/10/2008 5:28 AM

I haven't used ABB loadcells but they should operate under the same principle. I use Sartorius (which used to be Global Weighing which used to be Philips).

That spike in the excitation shouldn't be there. You didn't say how big it was though. I've got two guesses:

  • You've got a loose connection somewhere between the excitation and sense terminals. If the sense lines detect a drop in the excitation, it will tell the circuit to adjust by increasing the voltage. If it's intermittent, you'll get a spike.
  • Your shielding in the cables, junction boxes, transmitters, or loadcells themselves, is disconnected and allowing noise from switches, contactors, etc. to punch through. This can also happen if you've routed your loadcell cables next to high voltage lines (>220V).

You can also try replacing your transmitter with a new one just to check if the transmitter is at fault.

__________________
Miscommunication: when what people heard you say differs from what you said. Make yourself understood.
Reply
Active Contributor

Join Date: Jul 2007
Posts: 17
#3

Re: ABB Loadcell

01/13/2008 8:30 AM

It is not very clear from the question how many loadcells are there in your system. If it is a single loadcell unit, check the exiter power supply, the regulator, current limitter cct and ofcource any loose contact or drysolder points in loadcell circuit. And lastly the Loadcell itself.

If it is a multi loadcell unit, straight away interchange the connections of the troublesome loadcell with another one, if the problem tranfers with the loadcell replace it, otherwise check thaat line of wirings.

PS; check the mechanical mounting also.