Using FLUKE1403

06/03/2008 9:57 PM

We are testing chargers and inverters.

The procedure is to disconnect all PCB and fasten the ground lead to the ground and to touch AC input, DC output.

When the unit is tested for a first time DC output shows >550 M ohms.

After 2 hours of testing including 1 hour burn test on full load we must test the unit again. Normally we discharge capacitors, closing DC output breaker on the load and disconnecting the batteries.

The measurement of DC output to ground gives less than 300 M ohms.

Normally First Stage Filter capacitors are 6 in parallel.

What is wrong? Why the testing DC output to ground for a first time is showing >550, but finally testing is different?

Regards,

Ivan Araktchiev

Register to Reply
Pathfinder Tags: capacitors FLUKE ground
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Guru
Canada - Member - Our strength is our diversity

Join Date: Jan 2007
Location: Canada
Posts: 1024
Good Answers: 40
#1

Re: Using FLUKE1403

06/04/2008 3:31 PM

You are checking resistance with capacitors in the cct.

To test resistance, a small voltage is passed through the cct. This will charge or discharge the capacitors and give you false readings.

The resistance scale is designed to measure true resistance, (not capacitance or inductance)

__________________
Perfection is a subjective and abstract concept.
Register to Reply Score 1 for Good Answer
Power-User
Technical Fields - Education - New Member Technical Fields - Technical Writing - New Member

Join Date: Sep 2006
Posts: 367
Good Answers: 1
#2

Re: Using FLUKE1403

06/05/2008 12:30 PM

It could be "childhood insulation deterioration". And it is normal.

Register to Reply
Power-User
Technical Fields - Education - New Member Technical Fields - Technical Writing - New Member

Join Date: Sep 2006
Posts: 367
Good Answers: 1
#3

Re: Using FLUKE1403

06/05/2008 12:35 PM

Aha.. it is possible that Time Constant of the discharge is hight, over 2 hours. The effect well-known in the HV cables or transformers testing

Register to Reply
Guru
Popular Science - Weaponology - New Member Safety - ESD - New Member Hobbies - Fishing - New Member

Join Date: Sep 2006
Location: Near Frankfurt am Main, Germany. 50.390866N, 8.884827E
Posts: 17996
Good Answers: 200
#4

Re: Using FLUKE1403

06/05/2008 7:30 PM

Does the value measured change very slowly over say 10 minutes? If yes, you are slowly charging or discharging the caps via the battery in the meter.

To get an accurate reading, you need to leave the meter connected till the reading is completely steady. This can take a long period with really big caps.....

__________________
"What others say about you reveals more about them, than it does you." Anon.
Register to Reply Score 1 for Good Answer
Register to Reply 4 comments

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

Andy Germany (1); southern123 (2); techno (1)

Previous in Forum: stepper motor calculation   Next in Forum: Inline Refractometer Selections

Advertisement