Masterflex Peristalic Pump Issues

06/17/2008 9:27 AM

We use masterflex pumps and heads for our liquid product fills. We are experiencing a large failure rate in milliliters - goal 531-534ml , currently seeing 521.2-538.5ml. The fill amounts drift up or down with no pattern after "recalibration" of the pump drive.

I have checked pump belt, brushes, RPMs, and heads. check ok

Tubing Is correct type but shows striations on interior and minor inside diameter variation on a Y tube after less than 100 hours of use. Running two heads at half fill rate of 256.0ml

Is anyone familiar with these units in what causes the drift. We currently do not have our calibrated balance for weighing fill vials connected to "automatically" adjust fill rate. But is possible.

Tubing Platinum silicone L/S 35 ID .31" rated 3.8 ml/rev per head

2.23 psi coming from product source tank as line pressure, calculated

Avg RPM tested 547.8 per cycle with a 1.5 second delay between cycles.

Any feedback appreciated.

Register to Reply
Pathfinder Tags: masterflex Pumps
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Power-User
United States - Member - New Member Engineering Fields - Chemical Engineering - New Member Hobbies - Musician - New Member

Join Date: Jun 2006
Location: Wilmington, NC
Posts: 167
Good Answers: 4
#1

Re: Masterflex peristalic pump issues

06/17/2008 11:18 AM

Hi maizeegirl4,

Any temp difference between fill point and measuring point? Could be density difference causing the discrepancy.

Just a thought,

Mike

__________________
"Blessed is the man who, having nothing to say, abstains from giving us wordy evidence of the fact." George Eliot [Mary Ann Evans] (1819-1880)
Register to Reply
Participant

Join Date: Jun 2008
Posts: 4
#6
In reply to #1

Re: Masterflex peristalic pump issues

06/23/2008 11:26 AM

Any thoughts on options for what people are using instead?

Register to Reply
Guru
Popular Science - Weaponology - New Member United Kingdom - Member - New Member

Join Date: May 2007
Location: Harlow England
Posts: 16512
Good Answers: 670
#9
In reply to #6

Re: Masterflex peristalic pump issues

06/23/2008 12:10 PM

You can get diaphragm pumps, you can accurately measure the dose per stroke..some of the more expensive ones have adjustable stroke.

Seko, Prominent, Grunfos...loads of people do 'em. More expensive than a peristaltic..but much longer lasting/less maintenance.(Noisier too )

Just do a search for diaphragm metering pump, or some of the above names.

Del

__________________
health warning: These posts may contain traces of nut.
Register to Reply Score 1 for Good Answer
Participant

Join Date: Jun 2008
Location: Maumee, Ohio
Posts: 2
#10
In reply to #6

Re: Masterflex peristalic pump issues

06/23/2008 12:48 PM

Depending on product viscosity and consistency, you could use a small gear pump or progressive cavity pump. Both are good as dosing pumps.

Register to Reply
Guru
Popular Science - Weaponology - New Member United Kingdom - Member - New Member

Join Date: May 2007
Location: Harlow England
Posts: 16512
Good Answers: 670
#2

Re: Masterflex peristalic pump issues

06/17/2008 11:51 AM

I think you are just asking for too much accuracy from a peristaltic.

There are so many variables, even a bit of tension on the tube can change the delivery (if the tubes extand out of the pump head and are not anchored).
Does each pump have it's own power supply and is the supply over-rated by a good margin so it doesn't sag.

Is the motor/gearbox geared down nicely eg plenty of revs on the motor for low revs on the output shaft.

What is the wall/bore size of the tube and what material?

You may be able to improve things a bit.

Del

Del

__________________
health warning: These posts may contain traces of nut.
Register to Reply
Participant

Join Date: Jun 2008
Posts: 4
#7
In reply to #2

Re: Masterflex peristalic pump issues

06/23/2008 11:27 AM

Any information about what people are using or doing instead for product fills?

Register to Reply
Power-User
United States - Member - American by birth,southern by the grace of GOD Popular Science - Weaponology - you can't carry that! Technical Fields - Technical Writing - why can't I use a pencil? Hobbies - Target Shooting - New Member

Join Date: Jan 2008
Posts: 203
Good Answers: 2
#3

Re: Masterflex Peristalic Pump Issues

06/18/2008 8:17 AM

Just to make sure ... you pump is connected to a controller correct? If so the pump motor(rpm) is has a feed back from an opto coupler on the shaft near the end that has the brushes. make sure the slotted ring the coupler is counting, the holes are not plugged . We had some solution get in to ours and the speed went crazy. The RPMs will show good but the pump will speed up and slow down. And be enough to throw it off a bit.

__________________
The more complex the mind, the greater the need for the simplicity of play... If you think you can or think you can't you'll always be right
Register to Reply
Participant

Join Date: Jun 2008
Location: Maumee, Ohio
Posts: 2
#4

Re: Masterflex Peristalic Pump Issues

06/20/2008 9:46 AM

I agre with the previos reply about inside diameter of the hose. Furthermore, I woulod reccomend using a Tygon hose. The longevity is decreased but because it is a softer compound, more accurate results are possible. Just my cents.

Register to Reply
Guru
Popular Science - Evolution - New Member

Join Date: Jul 2006
Location: India-Chennai.
Posts: 722
Good Answers: 30
#5
In reply to #4

Re: Masterflex Peristalic Pump Issues

06/22/2008 12:54 AM

This subject pump gives about 521ml instead of original 531ml, which amount to about 1.8%.

Like Del the cat feels, this is too much of accuracy from a peristaltic type.

__________________
A picture worth thousand words: needless to say if it is animated.
Register to Reply Score 1 for Good Answer
Participant

Join Date: Jun 2008
Posts: 4
#8
In reply to #5

Re: Masterflex Peristalic Pump Issues

06/23/2008 11:28 AM

Any ideas about what people are using instead?

Register to Reply
Register to Reply 10 comments

"Almost" Good Answers: