Calibration Calculations

02/05/2009 9:38 AM

my question

one of close tank, tank hight 8840mm

cap fluid S.G 1.07

process fluid S.G 1

so how to calculate,what is the formula for calibration

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Anonymous Poster
#1

Re: Calibration Calculations

02/05/2009 10:04 AM

What are trying to calibrate? Fluid level? What kind of instrumentation are you using?

Register to Reply
Anonymous Poster
#2

Re: Calibration Calculations

02/05/2009 12:52 PM

Y=A+B/B-A+Q+R+T*96/Z

Register to Reply
Guru
Safety - ESD - New Member Popular Science - Cosmology - Amateur Astronomer Technical Fields - Technical Writing - Writer India - Member - Regular CR4 participant Engineering Fields - Optical Engineering - New Member

Join Date: Dec 2007
Location: 18 29 N 73 57E
Posts: 1390
Good Answers: 31
#3

Re: Calibration Calculations

02/05/2009 11:43 PM

There is never a formula for tank volume, as you can not make perfect geometry of a tank.

Practically you have to fill /empty the tank with known quantity of fluid and calibrate it.

Register to Reply
Active Contributor

Join Date: Jan 2009
Posts: 15
Good Answers: 1
#4

Re: Calibration Calculations

02/06/2009 3:29 AM

Depending on the size of your tank you would either measure the weight or the volume of a fluid of well known SG @ P and T.

That also depends on the calibration means you can access.

What is this tank for ?

You need to be more accurate in your statement of conditions if you want an accurate answer

thanks

Register to Reply
Associate

Join Date: Oct 2007
Posts: 29
Good Answers: 1
#5

Re: Calibration Calculations

02/10/2009 3:02 AM

Hello abmmm123,

Assuming that you are looking for Level Transmitter Range Values on a Closed Tank, with 8840 is distance between 0% & 100 % marks and LT is installed at the same height as that of 0% mark; LRV= -9458.8 mmwc & URV= -618.8 mmwc.

Regards

__________________
Best Regards, srzambre.
Register to Reply
Associate

Join Date: Oct 2007
Posts: 29
Good Answers: 1
#6

Re: Calibration Calculations

02/13/2009 11:59 PM

From the limited data supplied and assuming that Level Transmitter High side is at the same datum level with 0% mark,

LRV = -9458.8 mmwc

URV = -618.8 mmwc

Is this you are looking for?

__________________
Best Regards, srzambre.
Register to Reply Score 1 for Good Answer
Register to Reply 6 comments

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

Anonymous Poster (2); gsuhas (1); Reivax (1); srzambre (2)

Previous in Forum: Valve Benchsets   Next in Forum: Vacuum Gauge Calibration

Advertisement