| Dear Sir, |
| This is Ravi .M (m_ravi1110@yahoo.com) |
| We have a list of test & measurement equipments for our testing activity, |
| Presently we are operating this instrument manualy and I need to convert to this as a automition system |
| Examper : I need to verifiy RF signal generator it's having Frequency rang from |
| 10 MHz to 26.5 GHz and Amplitude rang -110dB to +13 dB |
| We will verifiy this RF signal generator using Spectrum analyzer with same rang with following below Procedure |
| 1 | Connect the equipment as shown in Figure |
 |
| 2 | On the Signal Generator , Press
PRESET. Set the controls as follows: |
|
CW frequency . . . . . . . . . . . . . . . . . . . 1.5 GHz |
|
Power level . . . . . . . . . . . . . . . . . .. . −10 dBm |
| 3 | On the spectrum analyzer, press PRESET. Set the controls as follows: |
|
Center frequency. . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 1.5 GHz |
|
Span . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . . 1 MHz |
|

|
| 4 | On the spectrum analyzer, press
RECALL, MORE 1 OF 2,FACTORY PRSEL PK. |
| 5 | On the spectrum analyzer, press
PEAK SEARCH. |
| 6 | Record the MKR frequency in Table 1-1 as the marker reading. |
| 7 | Repeat steps 2 and 6 for all frequency listed in table 1-1 |
| Frequency Accuracy Test |
| Input Frequency | Measured Frequency |
| 100MHz |
| 1GHz |
| 10GHz |
| 13GHz |
| 18GHz |
| 22GHz |
| 26GHz |
| Table 1.1 |
|
|