Load Cell for Batching Plant

07/24/2010 9:51 AM

How do i test a load cell to know if its bad n is it possible to repair a load cell?

I have an Artech load cell for our batching plant(ODISA MOD.10) and the electronic signal am getting on my LCD screen for the weight measured is wrong.from my observation i think three of the load cells are bad n only one is functional.The readings am having are really strange, because the output weight shown on my screen is always lower than what i have feed in to be batched.The S beam gage load cell has the following performance specification:

Standard Capacities20, 30, 50, 75, 100, 150, 200, 250,300, 500, 750 - Kgf
1,1.5,2, 2.5,3, 5, 7.5, 10, 15, 20-Tf
Excitation Voltage10 VDC- Maximum 15 VDC
Nominal Output3.0 m V/ V
Non - Linearity< ±0.025% FSO
Hysteresis< ±0.02% FSO
Non-Repeatability< ±0.01% FSO
Creep (30 minutes)< ±0.03% FSO
Zero Balance±1.0% FSO
Input Resistance392 ± 15 Ohms
Output Resistance350 ± 3 Ohms
Insulation Resistance (50 VDC)> 1000 Mega Ohms
Cable Length (4 Core Screened)5 mtr. Minimum
Safe Overload of Rated Capacity150%
Ultimate Overload of Rated Capacity250%
Allowable Side Load50%
Temp. Compensated Range0-60

0C

Temperature Effect on Output< 0.0015% FSO/

0C

Temperature Effect on Zero< 0.0020% FSO/

0C

Deflection< 0.4 mm at FSO
Tightening TorqueSee Table Below
Finish & ConstructionElectroless Nickel Plated Tool Steel
Environmental Protection ClassIP67
Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Guru
United States - Member - New Member

Join Date: Mar 2010
Location: In the pool because it is too hot.
Posts: 3054
Good Answers: 141
#1

Re: Load Cell for Batching Plant

07/24/2010 11:31 AM

IMO your check for the load cell will stop at the end of the connection wires. Once you open the housing, you'll probably find a not repairable situation, unless you send it back to the factory. You might be lucky if the electronic board failed. Load cells have been manufactured according accurate production and technology standards and repair should be done accordingly. Did you contact the supplier for service at your place? Following his instructions would be your best bet.

__________________
Plenty of room here
Register to Reply
Guru

Join Date: Oct 2008
Posts: 42355
Good Answers: 1693
#2

Re: Load Cell for Batching Plant

07/24/2010 1:47 PM

You'll need to test it against a known reference cell.

I agree with the vagabond, there are no user serviceable components inside.

Register to Reply
Anonymous Poster
#3

Re: Load Cell for Batching Plant

07/25/2010 6:35 AM

IF you have access to a resistance substitution box or a precision potentiometer you can make a substitute for the load cell input to the input card.This will tell if the cell is bad or the input needs recalibrating.

Register to Reply
Commentator
Fans of Old Computers - Commodore 64 - Grumpy Old Bear

Join Date: Sep 2007
Posts: 67
Good Answers: 8
#4

Re: Load Cell for Batching Plant

07/26/2010 8:33 AM

We use a different make but all should be similar. Maintainers jumper around each cell in turn. If the readout stability improves or there is no change whether the cell is in circuit or not, it is deemed bad and is replaced.

Units are non-repairable.

__________________
"Any fool can know. The point is to understand." - Albert Einstein
Register to Reply
Power-User
Canada - Member - New Member

Join Date: Feb 2010
Location: The Great White North
Posts: 185
Good Answers: 13
#5

Re: Load Cell for Batching Plant

07/26/2010 3:12 PM

If you have four load cells doing the same job you should be able to tell which one is bad with this basic test.

Each load cell with zero load on it should produce more or less 0mV at Zero weight and 30 mV at capacity on the signal wires. Check the mV levels at no load, place a known load on the scale and check the values again. The problem load cell should become evident.

If the values on the damaged load cell are not too far off the correct readings you may be able to set and offset in the preamplifier if there's one present or if the signal goes into a plc analog input card you may be able to program an offset into the plc code.

If all of the readings are similar then the load cells are not likely the problem and you need to look elsewhere.

Good luck.

__________________
Knowledge makes what was impossible, seem easy.
Register to Reply Score 1 for Good Answer
Commentator

Join Date: Nov 2009
Posts: 89
#6
In reply to #5

Re: Load Cell for Batching Plant

07/26/2010 3:49 PM

Elroy thank you very much sir.Can u please expalin further what you mean by set and offset in the preamplifier and program an offset into the card if it goes into a plc analog input.my has an LCD display screen and there is a reset button inside the unit.

I have a new thread on how to config the load cell,can u please check it out and let me know what u think?Thank you.

Register to Reply
Power-User
Canada - Member - New Member

Join Date: Feb 2010
Location: The Great White North
Posts: 185
Good Answers: 13
#7
In reply to #6

Re: Load Cell for Batching Plant

07/29/2010 1:16 PM

I am not familiar with your specific piece of equipment and as such am offering information based on experience, with the equipment I have worked on.

Some systems are set up with a pre-amplifier between the load cell and the readout. If your system is set up this way you can adjust the problem load cell output at the pre-amplifier by following the instructions provided by the manufacturer. The manufacturer should be able to supply you with the correct way to calibrate your system. The procedures are generally specific to each manufacturer.

If a PLC (programmable Logic Controller) is used in your system then there is an opportunity to program an offset into the PLC code.

Don't overlook the simple things to check for, such as loose or damaged cables or dirty connections. Make sure your excitation voltage is correct.

Find a manual for your equipment, or call in someone who is familiar with your specific system.

Remember when you are done you need to be certain that the system is reading accurately, or the readings are meaningless.

Good Luck

__________________
Knowledge makes what was impossible, seem easy.
Register to Reply
Register to Reply 7 comments

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

Anonymous Poster (1); benotype (1); dvmdsc (1); Elroy (2); IBGrizzly (1); lyn (1)