Load Cell Cale Length

09/07/2010 9:30 AM

Load cells or weigh cells for tanks come with a 5 wire cable usually about 10 meters long. They have a very nicely prepared end for connection to terminal strips in a summation box. Many customers have told me that you must never cut these wires and expect me to bundle them up in a junction box if the run from the cells to the summation box is far less than 10 meters. Where does this idea come from? Is it because the summation box operates on a wheat stone bridge principle? Can a few meters of wire length difference really cause a problem?

Note - I have cut them and suffered no ill effects on the operation of the weighing system, but I did match the cable lengths. (i.e. - cut 2 meters from one, cut 2 meters from all)

__________________
Remember when reading my post: (-1)^½ m (2)^½
Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Guru
Engineering Fields - Instrumentation Engineering - New Member

Join Date: Dec 2009
Location: San Antonio, TX USA
Posts: 844
Good Answers: 29
#1

Re: Load Cell Cale Length

09/07/2010 11:43 AM

The reason cutting wires isn't recommended is that the signal fom the wheatstone bridge is measured in millivolts. This signal is amplified in your "summation box". If your tanks operate with any comparative logic, the difference in wire length can result in variances in the voltage drop in the individual cell circuits. This is why if you trim them, always make them the same length.

__________________
"Do not worry about your difficulties in Mathematics. I can assure you mine are still greater". - Albert Einstein
Register to Reply
Guru
Safety - Hazmat - New Member Safety - ESD - New Member Engineering Fields - Transportation Engineering - New Member Popular Science - Evolution - New Member Technical Fields - Procurement - New Member Hobbies - Target Shooting - New Member Popular Science - Cosmology - New Member Engineering Fields - Architectural Engineering - New Member Technical Fields - Marketing/Advertising - New Member Engineering Fields - Food Process Engineering - New Member

Join Date: Dec 2005
Location: Mariposa Ca
Posts: 5800
Good Answers: 114
#2

Re: Load Cell Cale Length

09/08/2010 2:22 AM

I used to work in a bakery, which used 3 cells per hopper & regularly cut the wires to length [one cell was less than 6" from the sum box], It didn't affect the results as a full calibration had to be carried out anytime a cell was changed... The toledo displays we used have sophisticated electronics, the analog output from the readout fed the plc controlling the flour system, we spent far more time repairing the mechanical components.

the acceptable deviation was required to be less than 1/2%

Register to Reply Score 1 for Good Answer
Guru
Engineering Fields - Instrumentation Engineering - EE from the the Wilds of Pa.

Join Date: Feb 2006
Location: middle of nowhere, Pennsylvania
Posts: 2603
Good Answers: 63
#4
In reply to #2

Re: Load Cell Cale Length

09/08/2010 8:02 AM

That's what I thought. Thank you.

__________________
Remember when reading my post: (-1)^½ m (2)^½
Register to Reply
Power-User

Join Date: Jun 2009
Location: New Zealand
Posts: 403
Good Answers: 5
#3

Re: Load Cell Cale Length

09/08/2010 4:40 AM

Yea I cut them all the time, but since you are using a load cell summing junction box it shouldn't be a problem. The problem only arises when a load cell summing junction box is not used, and even then if the system is designed correctly it is usually not a problem.

You only have a problem with systems that are using 4 load cells, think about it a a chair with four legs but the fourth leg is shorter than the rest, in this case it can cause some calibration problems and a summing box is necessary. Using 3 load cells is a bit like triangulation it is very to go wrong.

Cheers

Joe

Register to Reply Score 1 for Good Answer
Guru
Engineering Fields - Instrumentation Engineering - EE from the the Wilds of Pa.

Join Date: Feb 2006
Location: middle of nowhere, Pennsylvania
Posts: 2603
Good Answers: 63
#5
In reply to #3

Re: Load Cell Cale Length

09/08/2010 8:04 AM

That's also what I thought - two so far agree with my thoughts on this. Thank you.

You can't believe the mess we have generated following others requests to stuff all that cable in a junction box.

__________________
Remember when reading my post: (-1)^½ m (2)^½
Register to Reply
Register to Reply 5 comments

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!