How Much Leakage Current Should be Allowed to Flow?

05/19/2011 5:55 AM

Hello,

I am performing a leakage current test for the epoxypolyester coated metallic (Galvanisesd steel) fixture with a total thickness of 2mm. Can anyone tell me what should be the leakage current that should be allowed to flow. Or what is the threshold value for the leakage current test to fail.

Thanks in advance

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#1

Re: How much Leakage current shoule be allowed to flow

05/19/2011 6:17 AM

A bit of context would help.

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
Active Contributor

Join Date: Sep 2009
Posts: 18
#3
In reply to #1

Re: How much Leakage current shoule be allowed to flow

05/19/2011 7:46 AM

tried that- thats why i am asking questions. I dont like posting questions for no good reason.

Register to Reply Off Topic (Score 4)
Power-User

Join Date: Nov 2008
Location: India
Posts: 333
Good Answers: 6
#2

Re: How much Leakage current shoule be allowed to flow

05/19/2011 6:23 AM

DEAR,

Any good quality insulation of electrical equipment should not allow more than 2-12 micro Amps. during HV test. ( not at rated voltage) There are not any guide lines for the leakege current apermitted while taking HV test, but various equipments I have tested, & found as

415 V motors - 2 mA

Cables - 33 KV / 11 kv / 22 KV - 2-4 Micro Amp

Distribution Transformers - 500 KVA - 11KV - 6 mA

- 1650 KVA - 33KV - 7 mA

But, general saying is ....for human safety, leakege current should be < 30 mA.

The leakage current starts affecting the body above 11mA.

So, in any case, for safety reasons, at rated voltage, leakage current should not be allowed more than 6 mA.

__________________
Exploring the Science of Electricity
Register to Reply Score 1 for Good Answer
Active Contributor

Join Date: Sep 2009
Posts: 18
#4
In reply to #2

Re: How much Leakage current shoule be allowed to flow

05/19/2011 7:49 AM

Thank You.

Register to Reply
Guru

Join Date: Jun 2010
Location: srilanka
Posts: 2725
Good Answers: 5
#5
In reply to #2

Re: How much Leakage current shoule be allowed to flow

05/20/2011 4:37 AM

Apart from HV test what are the permitted leakage currents for home/office/workshop equipment like computers,electric hand drills,hair driers,shavers,copiers,fax machines, electric cookers etc at rated voltage?. If you connect a device with more than 6mA leakage current in to a GFCI receptacle it will trip . Similarly if you connect 15 computers each with 2mA leakage current in a sub circuit protected by a 30mA RCD it too will trip.

From safety point of view what is the maximum leakage current allowed in any equipment?.

__________________
pnaban
Register to Reply
Power-User

Join Date: Jun 2010
Location: Kolkata, West Bengal, India
Posts: 172
Good Answers: 1
#7
In reply to #5

Re: How much Leakage current shoule be allowed to flow

05/24/2011 6:21 AM

Dear pnaban,

As per Indian Electricity Rules, 1956, clause no. 49 (2) (b), maximum leakage current allowable is one five- thousandth part of maximum current supplied to the consumer's installation. If the leakage current exceeds the above limit, supply of electrical energy must be discontinued for safety reason.

__________________
Manindra
Register to Reply
Guru

Join Date: Jun 2010
Location: srilanka
Posts: 2725
Good Answers: 5
#8
In reply to #7

Re: How much Leakage current shoule be allowed to flow

05/24/2011 8:30 PM

I need the maximum leakage current allowed in each device used by adults,babies etc as stupilated by manufacturer's or Govt standards/Regulations.

__________________
pnaban
Register to Reply
Participant

Join Date: May 2011
Posts: 1
#6

Re: How Much Leakage Current Should be Allowed to Flow?

05/20/2011 11:05 AM

I can tell you that for medical equipment and devices they are subject to the UL 60601-1 regulation or a maximum case leakage of 100 uA. That would be for an open ground, open neutral. You can Google that. It is to provide contact safety. However, this does not apply to non medical. It is stringent. An isolation transformer properly specified completely eliminates any problems and will usually keep it under 50 uA easy if not 20 uA.

Register to Reply
Active Contributor

Join Date: Sep 2009
Posts: 18
#9

Re: How Much Leakage Current Should be Allowed to Flow?

05/25/2011 1:47 AM

Thanks for providing the solution I have the correct values or the leakage current

Register to Reply
Register to Reply 9 comments

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!