Previous in Forum: Sensor   Next in Forum: Instrument Earth Bar and Safety Earth Bar
Close
Close
Close
11 comments
Rating: Comments: Nested
Power-User
Australia - Member - New Member

Join Date: Jul 2006
Posts: 267
Good Answers: 22

Thermocouple Errors - What to Expect?

09/03/2011 11:35 PM

I'm doing some measurement of temperatures using K-type thermocouples. I'm not using probes, just have the wires twisted together and sitting in fluid. In general this works well, but I have an annoying issue where, dunked in the same jar of water, three thermocouple wires give three slightly different steady state readings (12, 13 and 14 degrees C).

I haven't used thermocouples before except when someone else has set them up... can anyone make a suggestion why three seemingly identical devices (all cut from the same strand of cable) give three slightly different results? When I boil the water they all give slightly different temps as well - similar magnitudes.

I've tried swapping around the plugs at the DAQ - but still they stubbornly give these results. Could it just be due to slightly different resistances in the twisted-together wires? Or is there some other reason someone out there has seen before?

DAQ: Pico TC-08 running Picolog logging software.

Register to Reply
Pathfinder Tags: thermocouple error
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
2
Guru

Join Date: Feb 2008
Location: East Tenneesse, USA
Posts: 685
Good Answers: 46
#1

Re: Thermocouple Errors - What to Expect?

09/04/2011 12:47 AM

Yup, your correct. The contact points in the twist, The liquid influence in the joint.

Even if you weld your own the size of the joint/bead size will affect its reading.

Also if you weld the two wires to a piece of metal that metal will affect the reading also.

A company called Omega has a great handbook on temperature and it free (to qualified people). They also have other sections flow/pressure, strain, pH, data acquisition, and a lot more.

There books have a lot of information mixed in among there products and a tutorial sections in the back that gives a lot of good theory in each of the different books.

Hope this helps.

__________________
Metal is the material, The forge is life, The anvil and hammer bring character and soul.
Register to Reply Good Answer (Score 2)
Power-User
Australia - Member - New Member

Join Date: Jul 2006
Posts: 267
Good Answers: 22
#3
In reply to #1

Re: Thermocouple Errors - What to Expect?

09/04/2011 1:37 AM

Thanks - confirms my suspicions (and that I'm not going crazy). I'll head to the Omega site - thanks for the link.

At boiling point I'm getting variations between thermocouples of around 2 degrees C... so 2% error I can live with... it's the 16% error at low temps that's annoying as I was trying to catch some small variations. Never mind - I've redesigned the experiment to run over longer periods and I'm pumping the water round and round... so I don't have to try to capture 1 or 2 degree rises any more.

Register to Reply
Guru

Join Date: Feb 2008
Location: Van Nuys, CA
Posts: 563
Good Answers: 33
#2

Re: Thermocouple Errors - What to Expect?

09/04/2011 12:52 AM

Just twisting the wires is not sufficient. A proper junction requires welding, but I know some people get reasonable results with soldering. However, the dissimilar metals used for thermocouples do not cooperate well with soldering. Google "making thermocouples" and you will find many suggestions.

FYI: Here's a good "how to" guide on using thermocouples, including error compensation: http://zone.ni.com/devzone/cda/tut/p/id/7108

Register to Reply
Power-User
Australia - Member - New Member

Join Date: Jul 2006
Posts: 267
Good Answers: 22
#4
In reply to #2

Re: Thermocouple Errors - What to Expect?

09/04/2011 1:40 AM

Thanks for the links. And over the past few days I have worked out by experiment that it's a real bugger to solder thermocouple wires together!! Can't do it with my equipment... hence the twists. I'll see how the links help me!

Thanks

Robert

Register to Reply
Participant

Join Date: Sep 2011
Posts: 2
#5
In reply to #4

Re: Thermocouple Errors - What to Expect?

09/04/2011 6:48 AM

Understanding the material composition of thermocouple wires will enable the user to understand that it cannot be soldered. Even if it is possible it would not be wise. The essence of the measuring capability of thermocouples lies in the fusion of the two wires. The junction created between the two wires should prefferably be free from carbon buildup due to too much oxygen in the fusion process if a gas flame is used. For the laymen like you and me specialised equipment for fusing the wires is not available.

I suggest you use a gas flame, something like oxy-accetaleen or similar. Twist the wires as tight as possible for approximately 4 turns and then melt together. You should have perfect results.

Believe this will solve your questions

Regards

LJMarais

Register to Reply Score 1 for Good Answer
Power-User
Australia - Member - New Member

Join Date: Jul 2006
Posts: 267
Good Answers: 22
#6
In reply to #5

Re: Thermocouple Errors - What to Expect?

09/04/2011 7:16 AM

Thank you LJ - good thinking - the oxy torch is definitely worth a try.

Robert

Register to Reply
2
Power-User

Join Date: Nov 2009
Posts: 335
Good Answers: 63
#7

Re: Thermocouple Errors - What to Expect?

09/04/2011 3:56 PM

1) If the thermocouples come from different sources, different lots, if they're from different spools of wire, you can expect errors as shown on the limit-of-error table below, which states the standard and special limits of error for thermocouples. This table is published by all T/C manufacturers.

Thermocouple elements are metal alloys, which vary in consistency. Hence, the specification for commercial limits of error.

2) If you have a multichannel DAQ, there can be differences in cold junction compensation from channel to channel and only very sophisticated DAQ's have per-channel CJ compensation calibration.

3) If you have a single channel DAQ you could see variations in CJ just in the disconnection/connection process.

4) Even a jar of liquid has thermal gradients; varying temperatures at varying levels. That's why temperature controlled liquid baths used for calibration are agitated.

5) The twisted junction is the least likely source of error and that would be due to surface corrosion on the individual elements, quite unlikely on a K type whose elements are nickel alloys, unless you're using a polluted, heat damaged K wire with green rot.

Tens of thousands of lab experiments are done annually with twisted junctions T/C's, thousands of heat treat furnaces are surveyed to NADCAP and AMS2750 specs annually with twisted wire junction T/C's.

The reason for welding a junction is for reliability - holding up in whatever service it is in. Day in, day out industrial control or monitoring applications require the reliability of a welded junction.

Register to Reply Good Answer (Score 2)
Power-User

Join Date: Nov 2009
Posts: 335
Good Answers: 63
#8

Re: Thermocouple Errors - What to Expect?

09/04/2011 4:25 PM

>Also if you weld the two wires to a piece of metal that metal will affect the reading also.

This is true to the extent that thermal mass at the measurement point is changed. The transfer of heat is affected by the mass being measured and the mass of the measuring element. It is not true that the 3rd metal affects the thermocouple output because the chemistry of the junction is altered.

>A company called Omega has a great handbook on temperature

Omega does indeed have good material.

Their "Thermocouples" fundamentals app note (document #021-032), available here:
http://www.omega.com/temperature/z/pdf/z021-032.pdf
clarifies the 3rd metal issue.

Parts of page 3 of that document are shown below; those pertaining to the Thermocouple Law of

Intermediate Metals:


Check it out.

Register to Reply
Power-User

Join Date: Jun 2010
Posts: 344
Good Answers: 17
#9

Re: Thermocouple Errors - What to Expect?

09/04/2011 10:29 PM

How many decimal points in the readout? If you are displaying degrees, no decimals, the reading can be +/- one unit (one degree) out also.

Register to Reply
Guru

Join Date: Dec 2010
Location: North West England
Posts: 1170
Good Answers: 153
#10

Re: Thermocouple Errors - What to Expect?

09/05/2011 4:05 AM

For school electronics projects, I provided crimp ferrules normally used to put stranded wire into terminals. They come in various sizes but the ones suitable for 0.25mm - 0.5mm wire would be best for you. Cut back the insulation on both wires leaving bare wire longer than the ferrule. Insert both (untwisted) wires fully and crimp the ferrule. (There are specialist tools to do the crimp but squeezing in a vice, trying to apply the same pressure each time, will do) Cut back any surplus wire protruding from the end. It is not perfect but as all the ferrules are the same length the joints are both durable and consistent. Typical results are +/- 0.5°C For better accuracy, make more than you need and compare them against a purchased/borrowed (known standard) thermocouple. Use the ones that are nearest to the standard.

Register to Reply
Guru
Hobbies - DIY Welding - New Member

Join Date: Sep 2007
Posts: 1601
Good Answers: 58
#11

Re: Thermocouple Errors - What to Expect?

09/05/2011 8:55 AM

Twisting wires together is not the best idea because you don't know where along the twisted length the actual electrical contact is made, or not. This gives spacial uncertainty. It is better to make a weld at the end of the wires so you know the exact source location of your signal. You can buy or build a small arc welder at very low cost. Many other type of welding (gas, plasma, laser,etc.) work well also.

Register to Reply Score 1 for Good Answer
Register to Reply 11 comments

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

Carl_E (2); jhhassociates (1); LJMarais (1); metalSmiths (1); pantaz (1); Phaddy (1); RobertOz (3); welderman (1)

Previous in Forum: Sensor   Next in Forum: Instrument Earth Bar and Safety Earth Bar