Dear friends, I want to know what the slope is of an instrument. By changing the value of the factory setting slope, how will it affect instruments output?
Re: What is Slope of an Instrument
11/27/2011 7:59 AM
What is it you are trying to say?
Reply Score 1 for Off Topic
11/27/2011 11:13 AM
At a minimum it will change which end of the instrument is higher uphill. At worst it will make the instrument uncalibrated and any measurement taken by the instrument worthless. It all depends on what is the instrument in question?
If the instrument is an analog oscilloscope then the slope adjustment is there to remove any tilting of the trace caused by nearby stray electromagnetic fields. If the instrument's slope setting refers to the RS-422/423 serial communication, then this extends the maximum communication distance possible at the cost of the speed of that communication. If the instrument is a selectable filter network then the slope refers to how sharply the "out of band" frequencies will be removed from the signal path. There's many more possibilities depending on the instrument, but they all relate some how to the mathematical definition of the slope of a line. slope=Δy/Δx.
Good Answers: 53
11/27/2011 3:44 PM
If you are refering to process instrumentation slope is a term that is similar to gain.
11/27/2011 9:35 PM
For a typical instrumentation example, suppose that you have a 0-100°C range that you want to monitor with a 4-20ma current loop. The 4ma corresponds to 0°C, and the 20ma corresponds to 100°C. The slope is (20-4)/(100-0) = 0.16.
If the wrong slope value is used, the instrument will read accurately at one point at the most.
Fans of Old Computers - Commodore 64 - Grumpy Old Bear
11/28/2011 12:01 AM
For chemical/analytical instruments:
UV/VIS or other absorption/transmission equipment it would be the change in response of the detector per unit change of concentration,
ISE or instruments that respond to voltage change it would be change in mV/unit change of concentration (for pH theoretical is 59.157 mV/decade change of Hydrogen ion activity with the slope sometime being expressed as percent of theoretical)
In all cases the slope is set when the instrument is calibrated using two or more standards. Single point calibrations with manual entry of slope is usually very inaccurate.
"Any fool can know. The point is to understand." - Albert Einstein
Good Answers: 102
11/28/2011 2:12 AM
What is the instrument?
Good Answers: 21
11/28/2011 5:21 AM
Artwork just for you..the import of the document may have pulled a line a bit skew.
See the differences to the factory slope, and the effect ? Do you still need an explanation ? Naturally, this example is when the factory slope was linear in the first place....
11/29/2011 12:37 PM
thamks for ur reply
Good Answers: 27
11/28/2011 10:40 AM
In the A/D conversion circuit of any digital instrument, you will change its output accuracy when you alter or change the slope and linearity of the ramp signal used in the conversion process..
vsar
11/29/2011 1:53 AM
If the relationship between the input and output is linear, that can be represented by a line, every line will have a gradient , the gradient is called the slope.
11/29/2011 2:12 AM
In a musical synthesizer the slope would be how fast the volume changes when you press a key. Adjusting it would change what sort of instrument it sounds like.
Logic is the art of going wrong with confidence. -- Morris Kline
12/01/2011 7:02 PM
"Slope" is usually a change of gain.
Eg: In conductance measurements different liquids have different "slopes" in response to temperature. eg the conductance of a liquid might measure x uS at 50C and y uS at 70C. The plotted difference is the "slope" of the response to temperature. If the liquid being measured has a known "slope" the slope is entered into the instrument to correct the concentration measurement to a standard (usually 25C "International critical tables") The slope does not have be linear. Many instruments have polynomial correction formula's built in or look up tables.
Good Answers: 102
12/02/2011 3:02 AM
Slope is the ratio of 2 parameters. Slope can be constant (linear) or changing (non-linear)
Without knowing what the instrument is, what parameter the instrument is measuring, how it is detecting the parameter, what the other parameter is and so on then all these responses are just good examples and best guesses.