Selection Of Thermocouple

01/08/2012 12:59 PM

i have purchased a dry temperature bath (150 to 1100degree C) for calibration of other thermocouples in laboratory. pl. guide which class 1 thermocouple is better for calibration

k type Accuracy 1.5%

R or s Type accuracy 1.0 %

I have used k type in field but pl guide calibration point of view

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.
Guru

Join Date: Mar 2007
Location: by the beach in Florida
Posts: 33392
Good Answers: 1817
#1

Re: selection of thermocouple

01/08/2012 1:20 PM

The R or S type seem to be more accurate, why are considering the K type?

__________________
All living things seek to control their own destiny....this is the purpose of life
Register to Reply
Guru

Join Date: Jul 2010
Posts: 669
Good Answers: 176
#2

Re: Selection Of Thermocouple

01/08/2012 2:37 PM

The concept in commercial dry block calibrators is that the displayed temperature value of the heated insertion dry block is calibrated at the factory (or subsequently by a calibration lab) to a stated accuracy by comparison to a reference thermometer with traceable certification.

The uncertainty expressed in the dry block's specification and cal cert is based up this calibration and traceability of the reference thermometer. A dry block calibrator is generally purchased with a calibration cert certifying the accuracy/unceratinty spec.

My question is why are you trying to compare to some other reference? Do you not trust the vendor's certification?

Register to Reply
Associate

Join Date: Apr 2011
Posts: 43
#4
In reply to #2

Re: Selection Of Thermocouple

01/08/2012 3:12 PM

on internet the sensitivity of k type is 41uV/degree C and of S type 10uV/degree C and mentioned to use s type for higher range. i can not understand this

the temperature calibrator manufacturer has provided thermocouple but did not provide its calibration certificate. he has provided calibration certificate of calibrator ( which he calibrated with his own thermocouple). in my lab the accuracy with that thermocouple is about 4 degree C.

i think if i get class 1 it might give good result.

moreover pl guide a thermocouple accuracy has any relation with dia and lenght of thermocouple. i should buy small or big dia thermocouple.

Register to Reply
Guru

Join Date: Mar 2007
Location: by the beach in Florida
Posts: 33392
Good Answers: 1817
#5
In reply to #4

Re: Selection Of Thermocouple

01/08/2012 3:24 PM

R and S are essentially the same, maybe this will help...

http://en.wikipedia.org/wiki/Thermocouple

__________________
All living things seek to control their own destiny....this is the purpose of life
Register to Reply
Guru

Join Date: Jun 2011
Location: Mossel Bay, SA
Posts: 777
Good Answers: 21
#7
In reply to #4

Re: Selection Of Thermocouple

01/09/2012 5:06 AM

1. No-one issues a calibration certificate for an individual type 'k' t/couple.

2. Length and diameter depends on the application only.

in my lab the accuracy with that thermocouple is about 4 degree C.

How do you know this ? Compared to what ? Type 'k' means a clearly defined and constant characteristic within a specified temperature range. The fault may possibly lie within the temperature indicator, but unlikely (not impossible) the t/couple itself.

The t/couple junction may be slightly damaged due to mechanical shock, or you may be immersing an unsealed t/couple in conductive liquids.

3. Refer to your operating manual. The temperature bath has a calibration certificate, and it seems from your query that you wish to use a reference t/couple each time you calibrate other t/couples ... WHY ?

Register to Reply
Associate

Join Date: Jan 2012
Location: Australia
Posts: 49
Good Answers: 1
#9
In reply to #4

Re: Selection Of Thermocouple

01/18/2012 7:22 PM

If you wish to use a reference thermocouple for a dry wall calibration unit, then an s type would be a logical choice. It has a temperature range that covers your calibrator and a fairly high accuracy accross that range. However, it's a secondary calibration and there really isn't any advantage in the 0.25% you'll maybe get by using the S type. If your calibrating K type thermocouples, use a K type reference. No use comparing apples with oranges?? As for the size thermocouple, it will depend on the size of the thermowell on your calibration unit. You want the most evenly distributed heat transfer between element and sensor. So a close tolerance is the key. There is also a thermal transfer compund that can be applied to the sensor/thermowell to ensure optimal heat transfer. Don't get carried away with infantecimal accuracies on field equipment and secondary calibrations. There has to be some amount of practicality applied in the field. Generally these instruments form part of a control loop which will have inherit errors anyway, which will nullify all your hard work in gaining perfect thermocouples. Hope this has been of some help.

__________________
An educated man is not necessarily knowledgable, but has the intellect to call upon his resources.
Register to Reply
Guru

Join Date: Oct 2008
Posts: 42355
Good Answers: 1693
#3

Re: Selection Of Thermocouple

01/08/2012 3:08 PM

I'm amazed by your very "interesting" questions regarding thermocouples and calibration.

I'd suggest that you do two things.

1. Ask Iris to help you identify the proper specifications and procedures for the instrumentation in your laboratory.

2. Purchase the specifications and procedures that Iris recommends, and then read them, over and over until you understand what they say and mean.

Good luck.

Register to Reply
Guru
India - Member - New Member Engineering Fields - Electromechanical Engineering - New Member

Join Date: Oct 2006
Location: India, 200 Km. North of Delhi.
Posts: 1393
Good Answers: 53
#6

Re: Selection Of Thermocouple

01/08/2012 3:37 PM

Idealy whole calibrating system (sensor +reader) should be10 times accurate than the accuracy to be gauranteed for the calibrated device.

Register to Reply
Guru
Hobbies - DIY Welding - New Member

Join Date: Sep 2007
Posts: 1601
Good Answers: 58
#8

Re: Selection Of Thermocouple

01/09/2012 9:25 AM

Before counting on one thermocouple to check the accuracy of another thermocouple, I would consider using a Platinum RTD as my reference standard. I would also check with NIST to determine what they recommend.

Register to Reply
Power-User

Join Date: Feb 2007
Location: Gdansk, Poland
Posts: 100
Good Answers: 2
#10

Re: Selection Of Thermocouple

01/19/2012 6:41 AM

My supplemet to what was said above..

1. At lower temperatures, up to 460'C [optionally..650'C] using PRT = Platinum Resistance Thermometer (or Standard Platinum Resistance Thermometer = SPRT) would give better accuracy than Thermocouple. There are SPRT up to 1000'C, but they are expensive.

2. Using the same type Reference Thermocouple as calibrated Thermocouple allows for comparison method. Using comparison (or replacement) You can reduce many sources of errors (milliVoltmeter scale error).

3. Gold - Platinum Au-Pt Thermocouples are the most accurate, but also the most expensive.

Re size: probably 0.4mm diameter would be near optimum, but it depends on diameter and length of the hole ( insertion depth) in Your block.

Register to Reply