05/09/2007 1:08 AM
i wonder if any knows about the procedure of rosemount 1151 level transmitter first calibration procedure, electrical and control wiring, and installation??
Good Answers: 54
Re: smart transmitter calibration
05/09/2007 2:30 AM
I should just tell you to go to the Emerson-Rosemount website but I'm feeling generous. Here are the pdf manuals for the 1151. Typing what you asked for is going to take a long time.
Happy Reading!
1151 Pressure Transmitter
1151 Smart Pressure Transmitter
05/09/2007 3:43 AM
It is important to differentiate between the initial calibration of the transmitter, which will be done by the manufacturer and a certificate provided if requested at the time of purchase so as to link the instrument to national and international standards, and setting the range of the transmitter, which can be done at intervals if required by the user. The user needs to keep records of both.