Pressure Calibrator

06/28/2012 8:41 AM

Hi All,

During the Loop test we are using high range of pressure calibrator to do loop test, but the client do not allowed to use the high pressure calibrator.

inst range 0 to 20 barg

PRssure calibrator 0 to 100 barg

any advise please.

REgard

Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Guru
Technical Fields - Technical Writing - New Member Engineering Fields - Piping Design Engineering - New Member

Join Date: May 2009
Location: Richland, WA, USA
Posts: 21017
Good Answers: 795
#1

Re: Preessure calibrator

06/28/2012 9:18 AM

When reading only 20% of its full range, the calibrator might not be as accurate as desired. (It won't hurt anything, though.)

__________________
In vino veritas; in cervisia carmen; in aqua E. coli.
Reply
Active Contributor

Join Date: Feb 2012
Location: Manila, Philippines
Posts: 24
#2

Re: Pressure Calibrator

06/28/2012 4:53 PM

when calibrating a presure gauge you must consider the instrument range because instrument has their own min and max capacity for the accuracy and precision of the measurement.

__________________
Experience is something you don’t get until just after you need it.Good judgment comes from bad experience, and a lot of that comes from bad judgment.
Reply
Active Contributor

Join Date: Dec 2011
Location: Alberta Canada
Posts: 23
#3

Re: Pressure Calibrator

06/28/2012 8:56 PM

Are you doing a calibration or loop test. If you are just doing a loop check most HART compatabel instruments have a loop check function so you can make the instrument send a signal into the Control system.

Reply Score 1 for Off Topic
Anonymous Poster #1
#4
In reply to #3

Re: Pressure Calibrator

06/29/2012 6:37 AM

Hi all Actually were doing calibration but because I dont have pressure Calibrator close range to the test unit, therefore I use high range? Is it appropriate to use? Regard

Reply
Guru
Philippines - Member - New Member Engineering Fields - Instrumentation Engineering - New Member Engineering Fields - Control Engineering - Who am I?

Join Date: Oct 2006
Location: Northern Mindanao, Philippines
Posts: 2147
Good Answers: 53
#5

Re: Pressure Calibrator

07/11/2012 8:38 AM

I think your client is just afraid you'll damage his 20 barg pressure transmitter with your 100 barg calibrator.

This is usually not a problem unless you use an automated pressure generator. Say, you select 30 bars instead of 20 bars and the pump is just happy to oblige. We use a manually-operated pressure generator (hand-pump or screw-operated piston).

Our calibrator goes up to 100 barg as well and we use it to calibrate instruments down to 30 barg. We just make sure we inject no more than 30 bars.

regards,

Vulcan

__________________
Miscommunication: when what people heard you say differs from what you said. Make yourself understood.
Reply Score 1 for Good Answer
Reply to Forum Thread 5 comments

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

alpha (1); Anonymous Poster (1); decker1 (1); Tornado (1); Vulcan (1)

Previous in Forum: Passing Cylinder   Next in Forum: Hydro Test

Advertisement