Clamp-On Ultrasonic Flowmeter for Cement Lined Pipe

01/24/2013 2:20 AM

Dear All,

Can anyone please advise whether Clamp-on-ultrasonic flowmeters (transient time) may work on Cement lined pipe or not? One of manufacturers of clamp-on-ultrasonic flowmeters asked me that these flowmeters don't work on cement lined pipes. Please confirm if this is true & what is the reason ?

Also, in case if medium contains solid & air bubbles, flowmeter not performs well? Is this right? what is the reason?

Register to Reply
Pathfinder Tags: ultrasonic flowmeter
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#1

Re: Clamp-on-ultrasonic flowmeter for cement lined pipe

01/24/2013 3:32 AM

The more "lumpy" the fluid is, the better an ultrasonic meter will work; those lumps could be solids or bubbles, it doesn't matter.

There is some experience here of using them on ultrapure water (<1ppb impurities) in PVDF pipes.

<ducks the question on cement-lined pipe>

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
Guru
Engineering Fields - Instrumentation Engineering - EE from the the Wilds of Pa.

Join Date: Feb 2006
Location: middle of nowhere, Pennsylvania
Posts: 2603
Good Answers: 63
#4
In reply to #1

Re: Clamp-on-ultrasonic flowmeter for cement lined pipe

01/25/2013 8:23 AM

I'll say - the unbonded version with an ultrasonic detector is the basis for determining when an oil or gas well casing is not correctly cemented. The return signal on unbonded pipe is just the pipe ringing - you would get that same effect here. With the pipe bonded to cement you get what's behind the pipe in a well, and what's inside in this case. (I ran such an instrument for 11 years in the oil field.)

__________________
Remember when reading my post: (-1)^½ m (2)^½
Register to Reply
Guru

Join Date: Jul 2010
Posts: 669
Good Answers: 176
#2

Re: Clamp-On Ultrasonic Flowmeter for Cement Lined Pipe

01/24/2013 11:24 AM

The former Controlatron, now Siemens, clamp-on ultrasonic flow meters (FUS 1010 models) can work on lined pipe if the lining has not separated or unbonded from the metal pipe.

Siemens best fit is transit time ultrasonics that do NOT want suspended anything (particles or bubbles) in the fluid. The cleaner the fluid, the better for transmit time. Doppler ultrasonic technology requires suspended stuff, but transit time is just the opposite. Siemens does have doppler transducers, but with nowhere near the accuracy of their transit time siblings.

Register to Reply Score 1 for Good Answer
2
Associate

Join Date: Apr 2009
Location: Australia
Posts: 54
Good Answers: 5
#3

Re: Clamp-On Ultrasonic Flowmeter for Cement Lined Pipe

01/24/2013 11:03 PM

Sorry PWslack, what you wrote is wrong as discussed above. A transit time flow meter sends a pulse through the pipe wall into the liquid at about 45 degrees in a downstream direction. At that point, on the opposite wall, there is another sensor that picks up the signal and sends out another one in the upstream direction. The difference in transit time upsteam and downstream gives you an indication of the liquid velocity.

If the cement lining is thick or cracked or coming away from the wall, the signal won't transmit through it very well. Similarly, air bubbles and particles attenuate or disperse the signal too much and degrade performance.

So yes, they are correct in what they told you. (Don't worry , if they could have sold you one of their instruments they would have.)

__________________
Frankston
Register to Reply Good Answer (Score 2)
Register to Reply 4 comments

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

Frankston (1); Iris (1); Phys (1); PWSlack (1)

Previous in Forum: Labview   Next in Forum: LCD Dot Matrix Failure Modes

Advertisement