Calibration Question

08/24/2013 8:41 AM

We have AO2000 analyzer of ABB to measure SOx, NOx, CO, CO2, O2.

For example,

In general to calibrate O2, we have to prepare O2 0% gas for zero calibration

and O2 21% for span calibraton. It is natural.

However, we have swaped zero and span gas each other due to some reason.

we calibrated zero by span gas, calibrated span by zero gas.

What happen ?
There is not any problem. Calibration is good and measurement is also good.

The problem is that we can not explan the reason.

I would apprciate it if who could explain technical background ?

Register to Reply
Pathfinder Tags: calibration
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Guru

Join Date: Oct 2008
Posts: 42355
Good Answers: 1693
#1

Re: Calibration Question

08/24/2013 9:38 AM

Why would you do that? It makes absolutely no sense.

Why do you not follow the calibration procedure of the manufacturer?

"swapped zero and span gas each other due to some reason" ?????????? You don't know why?

You obviously are not calibrating the equipment properly.

<Shakes head; palm slaps self; wonders how such nonsense can happen; unsubscribes in disgust>

Register to Reply
Guru
Engineering Fields - Electrical Engineering - Been there, done that. Engineering Fields - Control Engineering - New Member

Join Date: Dec 2008
Location: Long Island NY
Posts: 15600
Good Answers: 981
#2

Re: Calibration Question

08/24/2013 10:52 AM

This kind of a question cannot be accurately answered here. This question can only be answered by the manufacturer of the instrument and they may not tell you the answer. You see, unlike ABB, we have no idea what physical process their instrument uses to identify the gas quantities and ratios. We also do not know the quality of your procedural methods. We know that you believe that you did the testing correctly but we do not know if you actually did it correctly. For all we know this calibration routine might be needed only when very high resolution measurements are needed and you're not looking at high resolution results.

__________________
"Don't disturb my circles." translation of Archimedes last words
Register to Reply
Commentator

Join Date: Aug 2008
Posts: 82
#3

Re: Calibration Question

08/24/2013 11:50 AM

In this case ,

span is made as 0% ie no span for test sample and 21% is made as zero level ( minimum proportion ) .

Means that analyser demands min. 21% in test sample with 0% span range .

Register to Reply Score 1 for Good Answer
Guru
Safety - Hazmat - New Member Engineering Fields - Mining Engineering - New Member Engineering Fields - Construction Engineering - New Member Hobbies - Hunting - New Member

Join Date: Dec 2006
Location: Scottsdale, AZ
Posts: 1014
Good Answers: 12
#4

Re: Calibration Question

08/24/2013 10:53 PM

All of our calibration certifications are done by third party by out state's regulations. Agree best bet talk to manufacturer or a third party that calibrates and repairs that particular piece of equipment.

__________________
John J Baker
Register to Reply
3
Guru

Join Date: Oct 2006
Location: 44.56024"N 15.307971E
Posts: 8277
Good Answers: 270
#5

Re: Calibration Question

08/24/2013 11:16 PM

Normally, you would use Nitrogen for Zero reference,and atmosphere for 21% or Span reference for O2.

It also depends on which modules you are calibrating,and whether they use an internal calibration reference cell or not.These will override erroneous calibration efforts.

If there is not a Negative scale or sign, then the unit has a built in undocumented automatic " ID TEN TEE" correction factor.

You need to find "Some Reason" and instruct him on proper calibration procedure.

__________________
"A man never stands so tall as when he stoops to help a child." "Never argue with a stupid person.They will drag you down to their level and beat you with experience" "To create an apple pie from scratch, first you must create a universe"
Register to Reply Good Answer (Score 3)
2
Guru

Join Date: Sep 2009
Posts: 1056
Good Answers: 88
#6

Re: Calibration Question

08/25/2013 7:49 AM

The obvious reason is that the device firmware programmer had the provision to not allow easily detectable bogus data to get anywhere near the calibration data saving routine. You must recalibrate. S.M.

__________________
Life is complex. It has a real part and an imaginary part.
Register to Reply Good Answer (Score 2)
Associate

Join Date: May 2012
Location: Ahmedabad, INDIA
Posts: 25
Good Answers: 1
#7

Re: Calibration Question

08/25/2013 12:20 PM

u have written due to some reason zero and span gas interchanged

it can also possible u haven't saved parameter at the end of calibration completion

o2 analyzer is still under working condition of previous calibration not this uproperly done calibration

I haven't seen a manual of ABB make o2 analyzer but if it is possible to check history parameter s of o2 analyzer which can show last calibration date and time

if inbuilt facility of history parameter check is available

[p| if u have provided 21% gas as zero and 0% gas as span

and u have set display range as 0 to 100%

if ur director is detecting 20% o2 u must get -95.23% at display a negative reading

Register to Reply
Guru

Join Date: Oct 2008
Posts: 42355
Good Answers: 1693
#8
In reply to #7

Re: Calibration Question

08/25/2013 12:41 PM

"u" is not a word.

It is "you".

This is an engineering forum, not a teen text forum.

Register to Reply
Associate

Join Date: May 2012
Location: Ahmedabad, INDIA
Posts: 25
Good Answers: 1
#9
In reply to #8

Re: Calibration Question

08/25/2013 1:31 PM

ok SIR I will kept in mind not to use "u" for you

infront of you I am like teen with not much knowledge compared to you

using small screen mobile keypad writing is so difficult thats why

Register to Reply
Guru

Join Date: Oct 2008
Posts: 42355
Good Answers: 1693
#10
In reply to #9

Re: Calibration Question

08/25/2013 1:38 PM

OK, that explains it.

Never mind.

Register to Reply
Guru

Join Date: Oct 2009
Posts: 1460
Good Answers: 30
#11

Re: Calibration Question

08/25/2013 2:43 PM

In the days in which I calibrated oxygen analysers for physiological experiments, we were taught why you do things as well as how you do them. The philosophy was something like this. You have an oxygen transducer with a linear output. Calibration at two points is the equivalent of establishing the coordinates of two points on the line and solving simultaneous equations. In theory you could put either point in first. However, on the machines that we used we had a knob which set the zero first. The second knob in effect adjusted the gain (slope of the equation) in order to set the second point. If you used the second knob first, and the zero was out of adjustment, you set the wrong slope, and you made matters worse by then setting the zero. If you use the wrong gas, which is what I think you are implying, you set a negative slope, which is impossible. Our analyser electronics were analogue, but I assume that modern equipment is digital and would simply refuse to recognise the calibration.

There is a further aspect to this, in that the accuracy of the calibration is improved by having the calibration points as far apart as possible. If your analysis involves concentrations above 21% and you want more accurate results then the correct second calibration gas is not air but 100% oxygen. In our lab, as I remember, we reckoned that air was not 21% but 20.93%. Medical oxygen is at least 99.5%. Industrial and research oxygens may be of even higher purity (up to 99.9999%).

Register to Reply
Guru
Safety - Hazmat - New Member United States - US - Statue of Liberty - New Member Engineering Fields - Chemical Engineering - Old Hand

Join Date: Mar 2011
Location: Lubbock, Texas
Posts: 14331
Good Answers: 162
#12

Re: Calibration Question

08/26/2013 9:14 AM

Don't be a cheapskate, spend the money on the calibration gases, and do over. Your instrument may be fooling you into thinking the response is correct when if fact, there is no change in response when the sample is changing. Make sure you follow the calibration procedure exactly as specified in the manuals in the future, and then test the calibration with a known sample less than the span.

__________________
If it ain't broke, don't fix it. Just build a better one.
Register to Reply
Power-User

Join Date: Mar 2011
Location: oz
Posts: 252
Good Answers: 10
#13

Re: Calibration Question

08/28/2013 1:56 AM

You can check your calibration by applying your span and Zero gases, to confirm what you have done is what you think.If zero gas reads 0 then you did not calibrate the instument the way you think.

Register to Reply
Register to Reply 13 comments

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

bakerjohn (1); eltech (1); HiTekRedNek (1); James Stewart (1); lyn (3); MUSUTH86 (2); phph001 (1); redfred (1); SimpleMind (1); sivaa (1)

Previous in Forum: Difference of PZT and PT Transmitters   Next in Forum: Question About Pressure Using Vacuum Truck

Advertisement