Electromagnetic Flow Meters

02/10/2014 9:12 AM

I have recently bought two new electromagnetic flow meters (MAG) for measuring water. there were installed in series because of the location and need for that setup. The challenge I am facing is there are giving me two different readings (ranging in the range of approximately 50cubic meters per day) and there is no leakage in between. I have already accounted the amount of water in the pipe between the two valves(6 cubic meters). Is it possible that electromagnetic valves were factory calibrated in correctly?

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Guru

Join Date: Oct 2008
Posts: 42355
Good Answers: 1693
#1

Re: electromagnetic flow meters

02/10/2014 9:18 AM

How many cubic meters flow through the pipe in one day?

What is the accuracy range of the meters?

Take the average of the two meters and don't worry.

Register to Reply
Guru

Join Date: Aug 2007
Location: Earth - I think.
Posts: 2143
Good Answers: 165
#2

Re: Electromagnetic Flow Meters

02/10/2014 9:37 AM

1. Check the specs and piping arraignment for the proper upstream and downstream parameters.

2. Check for proper grounding on both meters.

__________________
TANSTAAFL (If you don't know what that means, Google it - yourself)
Register to Reply
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#3

Re: Electromagnetic Flow Meters

02/10/2014 10:26 AM

Turn one of them off.

What did the flowmeter manufacturer say on the phone (apart from thanks for buying two instead of just one) about this?

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply Score 1 for Good Answer
2
Guru

Join Date: Jul 2010
Posts: 669
Good Answers: 176
#4

Re: Electromagnetic Flow Meters

02/10/2014 12:31 PM

Without numbers and values, model number, like maximum configured flow rate, sample readings showing error difference, distance between the meters, it's all speculation.

So I'll speculate. The answer to your question is no, the meters were not mis-calibrated at the factory. Prove to me with numbers that the meters are outside of the accuracy/uncertainty spec.

Register to Reply Good Answer (Score 2)
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#5
In reply to #4

Re: Electromagnetic Flow Meters

02/10/2014 1:40 PM

That's why one of them needs to be turned off!

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
Guru
Australia - Member - New Member

Join Date: Feb 2008
Location: Australia
Posts: 2181
Good Answers: 255
#6

Re: Electromagnetic Flow Meters

02/10/2014 4:04 PM

It is truly unlikely that quality meters were incorrectly calibrated.

What is more probable is that at least one of the meters has an instalation anomally.

It might be too close to a bend, or a valve or a reducer or some other flow obstruction that is causing turbulence and so disturbing the reading. It could even be the rubber seal is projecting inside the pipe.

Depending on configuration, there might be air inside the pipe etc.

We have some lines with two flowmeters. One at each end, to monitor that what we pump into the pipe arrives at the destination.

One other option (unlikely) is that in your software, the scaling parameter for mV to ML is incorrect. We spent some time chasing an 8ML "change" in one of our plants. It was a SCADA scale error, not the MAG. If that is the case, then the error you see will be an exact ratio to volume measured.

__________________
Just an Engineer from the land down under.
Register to Reply Score 1 for Good Answer
Power-User

Join Date: May 2013
Posts: 293
Good Answers: 3
#11
In reply to #6

Re: Electromagnetic Flow Meters

02/11/2014 9:58 AM

We have some lines with two flowmeters. One at each end, to monitor that what we pump into the pipe arrives at the destination.

Great idea!

Register to Reply
Guru
Australia - Member - New Member

Join Date: Feb 2008
Location: Australia
Posts: 2181
Good Answers: 255
#12
In reply to #11

Re: Electromagnetic Flow Meters

02/11/2014 3:38 PM

Yes,

Some of those lines are over 35km long across various landscapes and cover and since we are handling water, there is very little environmental "signal" of leaks, especially in wet times.

We also have some sewer lines (around 1km long) that go through environmentally sensitive locations where a "Spill" requires significant rectification and a lot of bother!! The use of flowmeters here provides som assurance that we are containing the material.

__________________
Just an Engineer from the land down under.
Register to Reply
Power-User

Join Date: Mar 2011
Location: oz
Posts: 252
Good Answers: 10
#7

Re: Electromagnetic Flow Meters

02/10/2014 4:11 PM

If you are using Krone mag flow meters, check that the GK numbers are the same on the sensor and on the controller.

Register to Reply
Guru

Join Date: Apr 2011
Posts: 8006
Good Answers: 286
#8

Re: Electromagnetic Flow Meters

02/10/2014 5:03 PM

"... there were installed in series because of the location and need for that setup...."

.

Following the word 'because' with unrelated info or circular reasoning doesn't really qualify as an explanation and isn't actually helpful. It does sound like the salesman was having a good day though. "why do you need two? we'll because it is going over there, of course....as well as the need to set it up that way"

.

.

"...two different readings (ranging in the range of approximately 50cubic meters per day) and there is no leakage in between...."

.

That may not be a problem at all. If the total daily is 10,000 m3 or greater, the meters could be accurate within ±0.25% and still show a 50 m3 discrepancy. What accuracy are the meters supposed to provide? What is the typical total daily flow?

.

"...I have already accounted the amount of water in the pipe between the two valves(6 cubic meters)..."

.

Okay, now you have my attention. How exactly are you accounting for the water in between? Can you describe the process by which the pipe volume between changes the flow past the two points? There isn't a point when the pipe it less than full, is there?

__________________
Eternal vigilance is the price of knowledge. - George Santayana
Register to Reply Score 1 for Good Answer
Guru

Join Date: Dec 2010
Location: North West England
Posts: 1170
Good Answers: 153
#9

Re: Electromagnetic Flow Meters

02/11/2014 4:39 AM

All liquid flow meters should ideally be installed in vertical pipe with an upwards flowstream. This arrangement helps to guarentee a solid column of water so the meter is not seeing partially filled pipes and registering it as full bore flow. If one or both are mounted in horizontal pipe you are likely to see differences in the output.

Register to Reply
Active Contributor

Join Date: Apr 2012
Posts: 22
#13
In reply to #9

Re: Electromagnetic Flow Meters

02/12/2014 6:33 AM

thanks jhhassociates, have realised that its one area which was overlooked in terms of installation. One of the meter is installed in a vertical pipe while the other is installed in a horizontal pipe. Apparently the one which is installed in a vertical pipe is giving us the smaller reading. Under this scenario is it that which is supposed to be expected.

Register to Reply
Guru

Join Date: Nov 2007
Location: Eastern Arizona mountains on Route 666 about a mile from God's country
Posts: 1676
Good Answers: 122
#10

Re: Electromagnetic Flow Meters

02/11/2014 9:00 AM

What is the distance between the flow sensors (flowtubes)?

You must have uninterrupted flow (no disruptions) for a minimum of 5 x pipe ID upstream of the first sensor as well as in between the meters with 2 x pipe ID on the downstream side of the last sensor. (no bends, no fittings, no penetrations, no dents, no pipe flexibility, no connections,etc.)

Any/all cavitation and/or disruption in the flow will cause offset and/or error(s).

If the distances are not acheivable, my advice is to use an four-hole orifice plate meter with temperature compenasation. ER, EH, ABB, & Siemens all make this type of meter.

__________________
They said; "Brain size?" I heard; "Train size?" so I said: "I'll take a small one, thank you."
Register to Reply
Active Contributor

Join Date: Apr 2012
Posts: 22
#14

Re: Electromagnetic Flow Meters

02/13/2014 6:29 AM

i need clarification on this, whats the effect of air in the system, how does this affect the overall flowrate considering that the medium being measured is water and how does turbulance affect this. Want to reiterate that the one connected in the vertical column is giving me the lesser reading.

Register to Reply
Guru

Join Date: Dec 2010
Location: North West England
Posts: 1170
Good Answers: 153
#15
In reply to #14

Re: Electromagnetic Flow Meters

02/13/2014 12:47 PM

To avoid turbulence affecting (lowering) the reading you need straight pipe with a minimum of 10 pipe diameters upstream of the meter (20 is better if you can achieve it) and 5 pipe diameters downstream of the meter. Air in the system will give you a false reading. The meter is reading the magnetic field strength and velocity of the liquid. As more air is entrained the field strength goes down but the velocity goes up. With a small amount of air the results are unpredictable, but with a large amount of air (10% +) the reading will always go down because field strength is the dominant factor.

Register to Reply
Power-User

Join Date: Mar 2011
Location: oz
Posts: 252
Good Answers: 10
#16
In reply to #14

Re: Electromagnetic Flow Meters

02/13/2014 3:17 PM

Flow=area of pipe X velocity. Mag flows measures velocity and always assumes it is a full pipe, so if the pipe is not full , your reading will be wrong.Some magflows works with a level meter installed besides it.

Register to Reply
Guru
Philippines - Member - New Member Engineering Fields - Instrumentation Engineering - New Member Engineering Fields - Control Engineering - Who am I?

Join Date: Oct 2006
Location: Northern Mindanao, Philippines
Posts: 2147
Good Answers: 53
#17

Re: Electromagnetic Flow Meters

03/21/2014 9:16 AM

Using double instruments to measure the same thing is asking for a headache. Unless they are measuring exactly the same point (e.g. two temperature sensors within an inch of each other) you're most certainly going to get a difference.

In our process, we have two flow meters that are essentially connected in series though there are vessels in between them. A difference usually signals a leak somewhere. This is the only reason I can think of to have this kind of setup.

Are your pipes completely filled? No bubbles? No debris? Is there enough straight run upstream of the flow tubes? If you're using remote heads, are the excitation cables of the same length and size? Do you have the same voltage for their supplies?

If your answers to the above are all "yes" maybe you can try connecting the flow tubes in series within a few feet of each other and comparing their readings. It's possible there is a slight difference in calibration. Or you can send them to a certified calibration facility.

regards,

Vulcan (",)

__________________
Miscommunication: when what people heard you say differs from what you said. Make yourself understood.
Register to Reply
Register to Reply 17 comments

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

aDIYguy (1); eltech (2); Iris (1); jhhassociates (2); Just an Engineer (2); Kilowatt0 (1); lyn (1); PWSlack (2); SHOCKHISCAN (1); tawanda (2); truth is not a compromise (1); Vulcan (1)