Supco Humidity Sensor Calibration

03/02/2014 3:33 AM

I have bought on Amazon a Supco air quality meter and now according to the instruction enclosed need to calibrate it for 2 humidity levels with 33 % and 75 % salt solutions. Can I prepare myself the solution with purified NaCl and distilled water, and more importantly - is not the sensor going to be damaged after having been in contact with salt for an hour. It looks so delicate !

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Guru
Hobbies - Fishing - New Member

Join Date: Jun 2008
Location: Raleigh, NC USA
Posts: 13529
Good Answers: 468
#1

Re: Supco humidity sensor calibration

03/02/2014 8:28 AM

You don't place the sensor in the salt solution, you put the probe in the salt solution. Just rinse it when you are done.

From one of their manuals:

11Plug the sensor probe into 33% salt bottle. Hold down and under normal mode to enter 33% calibration(Fig.17). "CAL" and calibrating value (32.7% if at are blinking on the LCD with current temperature at the left.Meter is now calibrating, and will finish in about 60 minutes when "CAL" and humidity stop blinking. (Fig.18) o25 C) The meter defaults to be calibrated the humidity with 33% and 75% salt solution. The ambient condition is recommended oto be at 25 C and stable humidity(better to be close to the calibrating value). To abort calibration, hold down for more than 1second at any time. RH CALIBRATION %%OOFig. 17%%OOFig. 18After 33% calibration, plug the sensor probe into 75% salt bottle, then press to enter 75% calibration (Fig.19)

__________________
Those who would give up essential Liberty, to purchase a little temporary Safety, deserve neither Liberty nor Safety. Ben Franklin
Register to Reply
Power-User

Join Date: Jul 2010
Posts: 136
Good Answers: 19
#2

Re: Supco Humidity Sensor Calibration

03/02/2014 7:28 PM

>Can I prepare myself the solution with purified NaCl and distilled water?

If you can figure out how to make a saturated solution at a specific % RH, you can, but you won't have a valid calibration certificate certifying the solution.

Register to Reply
4
Guru

Join Date: Jul 2006
Location: Eastern Kansas USA
Posts: 1503
Good Answers: 128
#3

Re: Supco Humidity Sensor Calibration

03/02/2014 11:15 PM

Friends,

An error appears to be in the answers here!!! The salt solutions in a sealed air chamber at a specified and stable temperature will create a constant and reproducible concentration of water in the air above the salt solution. This will be a known percent of the amount of water that can be held by the air at that temperature, and is called the relative humidity.

At no time should the sensor be placed in the salt solution. Equilibrium is essential to getting a good sensor calibration, and this may require many minutes or even an hour of time (depending on the desired accuracy) after placing the sensor into the air in the chamber.

For practical work, a solution made according to standard inorganic chemists' practices should be acceptable (reagent grade salt and deionized water). The best and easiest method to check the calibration of the instrument is with a measurement of dry and wet bulb temperatures in the air being sampled (using a sling psychrometer).

--John M.

Register to Reply Good Answer (Score 4)
Guru
Hobbies - Fishing - New Member

Join Date: Jun 2008
Location: Raleigh, NC USA
Posts: 13529
Good Answers: 468
#4
In reply to #3

Re: Supco Humidity Sensor Calibration

03/03/2014 4:38 AM

GA. I was confused by my own answer, when I pasted the info from the manual.

This explains the term, "Plug the probe into the bottle".

I'm a little surprised that the company doesn't include the test bottle with the sensor, along with adequate instructions for calibration.

__________________
Those who would give up essential Liberty, to purchase a little temporary Safety, deserve neither Liberty nor Safety. Ben Franklin
Register to Reply
Guru

Join Date: Jan 2006
Location: Tamworth, UK.
Posts: 1782
Good Answers: 45
#5

Re: Supco Humidity Sensor Calibration

03/03/2014 5:02 AM

You might find that 'salt' in this context is a general word. Unless it clearly says so, it probably isn't NaCl.

__________________
When arguing, remember mud-slinging = lost ground.
Register to Reply
Guru

Join Date: Jul 2006
Location: Eastern Kansas USA
Posts: 1503
Good Answers: 128
#7
In reply to #5

Re: Supco Humidity Sensor Calibration

03/03/2014 8:09 AM

Horace40,

I didn't want to pay for access to the academic journal article that discussed the various salt solutions, but the abstract that I viewed on-line did mention NaCl as one of the solutions that was used for certain %RH measurements. It also said that the article looked at both saturated solutions as well as non-saturated ones. You are correct that many other salts besides NaCl are used for achieving the various %RH values.

--John M.

Register to Reply
Guru

Join Date: Jan 2006
Location: Tamworth, UK.
Posts: 1782
Good Answers: 45
#6

Re: Supco Humidity Sensor Calibration

03/03/2014 5:59 AM

Damn time out.

Horace40 Post should end '....for 33%'

Tail end of sentence got clipped and lost in the copy and paste routine.

A known amount of 'salt is dissolved in a known amount of water in an enclosed container with air space.

The probe is exposed to the air in the space above the solution.

__________________
When arguing, remember mud-slinging = lost ground.
Register to Reply
Guru
Hobbies - DIY Welding - New Member

Join Date: Sep 2007
Posts: 1601
Good Answers: 58
#8

Re: Supco Humidity Sensor Calibration

03/03/2014 9:47 AM

As you have been advised, it is only the gas above the saturated solution that is used for humidity measurements. Also, the 75% solution is for NaCl only. To prepare a 33% solution, you will be advised to use MgCl.

Register to Reply
Power-User

Join Date: Jun 2009
Posts: 262
Good Answers: 1
#9

Re: Supco Humidity Sensor Calibration

03/04/2014 12:10 PM

Thank you !

Evidently, it turned out my doubt was not without reason !

Register to Reply
Guru
Hobbies - DIY Welding - New Member

Join Date: Sep 2007
Posts: 1601
Good Answers: 58
#10
In reply to #9

Re: Supco Humidity Sensor Calibration

03/04/2014 1:42 PM

The use of saturated salts to calibrate humidity sensors is very easy and straight forward. The problem is that we have a lot of good folks, each with good intentions, that are confusing you. Get a pile of sodium chloride and mix it with water to make a slurry. There must be enough sodium chloride to saturate the solution (more than will dissolve in the water). Put this mixture in a bowl. Place your sensor in the space between the solution and a sealed lid. Use a tight fitting plastic lid or saran wrap with an elastic band to create a seal. The air between the cover and the salt slurry, the RH will reach equilibrium(75%RH) in about an hour. No psychrometer is needed because moisture calibration using saturated salt is more accurate than a psychrometer. You could calibrate with a chilled mirror sensor, but it's not necessary. Repeat the same process with magnesium chloride for 33% RH. The air inside your chamber will contain the required water vapor. Just don't get any of the liquid salt solution in contact with your sensor.

Register to Reply Score 1 for Good Answer
Power-User

Join Date: Jun 2009
Posts: 262
Good Answers: 1
#11

Re: Supco Humidity Sensor Calibration

03/05/2014 5:10 AM

Thank you very much ! Have not encountered the subject of RH sensors calibrating till now.

Register to Reply
Register to Reply 11 comments

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

Bud (1); horace40 (2); jmueller (2); kramarat (2); welderman (2); Yuri B. (2)

Previous in Forum: Problem in Operating Enpac 2500 Data Collector   Next in Forum: When To Use NAMUR Standard In A Switch?

Advertisement