03/01/2014 12:17 PM
I am new to vibration measurement and using enpac 2500 data collector in offroute mode.
When connected the transducer(9000 rev B accelrometer) at CH1 and set the offroute mode with dispalcement unit pk pk their is no response in the meter. Seeking support and want to know the possible reasons with solutions.
Regards,
Tech-Freaky
Re: Problem in operating Enpac 2500 data collector
03/01/2014 1:18 PM
Are you new to the instrument? Instrumentation in general? Data collection?
Never used this instrument myself, but it seems to be straight forward and simple.
Without something to give a clue of the set-up, etc. all you'll get here are blank stares.
First, download the manual if you don't have one. READ IT if you do have one.
Enpac 2500 Data Collector User Manual - Rockwell Automation
03/01/2014 1:41 PM
Thanks for your response but i am looking for specific solution and discussion on queries to improvise my understanding. Please if you reply to my post next time, I would expect a specific response.
Awaiting for reply from someone who has used enpac 2500.
03/01/2014 1:51 PM
Well, pardon me for living!
You say, "Seeking support and want to know the possible reasons with solutions."
I say, contact technical support at Rockwell and pay somebody to read your mind.
The solution to your problem is to learn how to operate the equipment at hand.
A talent you do not currently posses.
I'll be more than happy to leave you now to wait for someone to solve your undefined problems for you.
<unsubscribe>
Good Answers: 102
Re: Problem in Operating Enpac 2500 Data Collector
03/02/2014 2:21 AM
So, you've read and understood the manuals for the 2500 (around page 183) and the 9000 rev B accelerometer?
03/03/2014 5:32 PM
The best source of support is the supplier of the equipment. Go on. Make the call.
03/06/2014 3:56 AM
Thanks guru for replying. You are right i have gone thru the manual and done the settings as mentioned in it. But still not able to get data. Need support badly :(
03/07/2014 3:18 AM
Hello friends,
Found the problem and resolved it, the sensitivity parameter was the troubling me. Changed the sensitivity to 50 mV instead of default 100mV/EU and found the readings OK.
Thanks all for your suggestions. Looking forward to get similar support in future.