02/28/2014 12:50 PM

Why blocking a port on a digital pressure (differential) manometer causes gradual changing of the value indicated ? As if pressure is increasing in the instrument's measuring chamber. I thought at first the reason was the current passing through the sensor heated that small volume of the air trapped inside and caused it to expand. Later I blocked the ports with a vessel connected to minimize the possible expansion, but the effect was the same. What causes that - the instrument sort of living its own life ? The pressure of the air inside should not change on its own, without a reason, should it?

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Guru
Engineering Fields - Electrical Engineering - Been there, done that. Engineering Fields - Control Engineering - New Member

Join Date: Dec 2008
Location: Long Island NY
Posts: 15602
Good Answers: 982
#1

Re: why closing manometer port causes pressure increas ?

02/28/2014 12:59 PM

Maybe you just found a tny leak.

__________________
"Don't disturb my circles." translation of Archimedes last words
Register to Reply
Power-User

Join Date: Jun 2009
Posts: 262
Good Answers: 1
#2
In reply to #1

Re: why closing manometer port causes pressure increas ?

02/28/2014 1:31 PM

No leaks. Try it yourself - close the "+" port tightly and you will see: 1.1, 1.2, 1.3 ..10...20 mB and so on steady increase.

Register to Reply
Guru
Engineering Fields - Electrical Engineering - Been there, done that. Engineering Fields - Control Engineering - New Member

Join Date: Dec 2008
Location: Long Island NY
Posts: 15602
Good Answers: 982
#4
In reply to #2

Re: why closing manometer port causes pressure increas ?

02/28/2014 1:56 PM

Sorry but unless this manifold was helium leak checked to UHV conditions I don't believe you that no leaks exist. At the same time the actual leak could easily be orders of magnitude less than this mysterious effect.

I think that I will take you up on the offer to try it myself. Who do I charge for my transportation and lodging expenses to get to your facility? Being a CR4 member I'll give you a big discount for my billable hours.

__________________
"Don't disturb my circles." translation of Archimedes last words
Register to Reply Score 1 for Good Answer
Power-User

Join Date: Mar 2011
Posts: 238
Good Answers: 14
#3

Re: why closing manometer port causes pressure increas ?

02/28/2014 1:47 PM

Maybe a drifty sensor? Are you plugging one port or both since it is a differential. If blocking only one port you may be seeing barometric pressure changing.

Register to Reply
Guru
Engineering Fields - Power Engineering - New Member

Join Date: May 2007
Location: NYC metropolitan area.
Posts: 3230
Good Answers: 444
#5

Re: why closing manometer port causes pressure increas ?

02/28/2014 2:57 PM

Closing a port on a differential manometer makes it an absolute manometer, so you're seeing the pressure change on the open port. With both ports blocked you're probably seeing "sensor drift", but it could be that one port is a little less airtight than the other.

__________________
“Tell me and I forget. Teach me and I remember. Involve me and I learn.” Ben Franklin.
Register to Reply Score 1 for Good Answer
3
Guru

Join Date: Sep 2007
Location: East Texas
Posts: 1439
Good Answers: 31
#6

Re: why closing manometer port causes pressure increas ?

02/28/2014 9:38 PM

I really shouldn't be playing so far out of my sand box; but although the crow I so often dine on always tastes the same, I always manage to learn something from the experience.

You didn't say whether the pressure you were measuring was of a gas or a liquid.

In either of these cases my first question would be - are you allowing the process to come to thermal equilibrium with the environment before taking your readings?

A quick review of PV=nRT would indicate an increase in pressure of a closed system if the sample entering the chamber is below the ambient temperature. Once the sample was isolated within the closed system wouldn't the pressure increase as a result of the sample temperature increasing towards equilibrium with ambient temperature?

If the pressure of the sample medium is much higher than the ambient pressure then wouldn't it have a tendency to cool below ambient as it entered the test chamber?

If the sample was soluble in the working fluid wouldn't the density of working fluid or sample change as it was brought into solution? In this case would this not cause an indicated change in pressure?

Are you taking measurements near the center of the instruments range?

If the instrument is using a piezoelectric sensor would it not be possible that the cooler sample would be cooling one side of the piezoelectric sensor relative to the other side? What is the coefficient of expansion of the piezoelectric sensor and how would this effect the flexing stress on the sensor?

__________________
"The greater danger for most of us lies not in setting our aim too high and falling short; but in setting our aim too low, and achieving our mark." -- Michelangelo
Register to Reply Good Answer (Score 3)
Active Contributor

Join Date: Feb 2014
Posts: 21
Good Answers: 2
#7

Re: why closing manometer port causes pressure increas ?

03/01/2014 5:50 AM

I'm pretty much a total amateur in this area but given what you've said my guess is that the results you are getting definitely sound like a leak. - You must remember that once closed the instrument is a very small system so the effects of any leak will be massively amplified. Connect it back to the main system and the leak effectively disappears. Such a small leak should not interfere significantly with performance of the overall system unless the system is very small.

From the figures you've given I'm guessing that your system is intended for medium vacuum.

To answer in more detail my first question would be- exactly how does the particular type of manometer actually work? My guess is that being differential and with what you've said that the instrument either works by measuring the flow rate through a calibrated channel or uses a common bellows between two sealed chambers. (However it could be using some other method - like two isolated calibrated sensors with an electronic comparator.)

You seem to imply that the instrument is heated, is that to give a calibrated fixed temperature or is it part of the mechanism that measures the pressure? like a Pirani gauge. (just looked that up )- https://en.wikipedia.org/wiki/Pressure_measurement#Thermal_conductivity

Further note if it is a Pirani type gauge then I would guess you should be getting a pressure rise on a closed chamber without a leak. A gas when heated expands and in a closed space that means a rise in pressure. However it should also produce an opposing effect producing an apparent pressure drop due to heating reducing the temperature differential between the gas and the heating element. Afraid you probably need a real expert to answer that one...

Register to Reply Score 1 for Good Answer
Guru

Join Date: Oct 2006
Location: 44.56024"N 15.307971E
Posts: 8277
Good Answers: 270
#8

Re: why closing manometer port causes pressure increas ?

03/01/2014 11:42 AM

Try hooking the port to an inclined manometer(Such as those made by Dwyer or others),which can read in very low ranges of pressure, to the opened port under question.If the pressure is really increasing, you will know by observing the liquid manometer.If not, you have a problem internal to your manometer.Make sure your port is open.The liquid will act as a plug,same as valving it closed.The advantage is you get to see what is Really happening inside your instrument in question.

__________________
"A man never stands so tall as when he stoops to help a child." "Never argue with a stupid person.They will drag you down to their level and beat you with experience" "To create an apple pie from scratch, first you must create a universe"
Register to Reply
Power-User

Join Date: Jun 2009
Posts: 262
Good Answers: 1
#9

Re: why closing manometer port causes pressure increas ?

03/01/2014 1:30 PM

Probably leak and/or drift, indeed, thank you !

I was hoping for an effect of an absolute pressure manometer, but have proven to myself the idea was hopeless.

Register to Reply
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#10

Re: Why Closing Manometer Port Causes Pressure Increase?

03/03/2014 5:35 PM

What is the weather doing in the vicinity of this blocked manometer? And what does <...gradual changing...> mean? +/-50mBar is the differnce between a pleasant sunny day and the most atrocious storm.

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
Power-User

Join Date: Jun 2009
Posts: 262
Good Answers: 1
#11
In reply to #10

Re: Why Closing Manometer Port Causes Pressure Increase?

03/04/2014 12:17 PM

No no, in 2-3 minutes pressure rises to 20-30 mB. The more or less the same at two manometers : Testo 510 and a more cheaper Pyle.

Register to Reply
Guru
Popular Science - Weaponology - New Member Hobbies - Fishing - New Member

Join Date: May 2007
Location: Aloha or
Posts: 659
Good Answers: 19
#12
In reply to #11

Re: Why Closing Manometer Port Causes Pressure Increase?

03/04/2014 2:31 PM

you have about a 10mB a minute leak in that volume you are measuring. IF we knew the volume we could tell you the size of the leak.

__________________
Closed biased minds are utterly impervious to any factual evidence which contradicts their beliefs
Register to Reply
Active Contributor

Join Date: Feb 2014
Posts: 21
Good Answers: 2
#13

Re: Why Closing Manometer Port Causes Pressure Increase?

03/04/2014 4:37 PM

I don't think you need to worry too much Yuri B. Like I said before the volume inside the instruments will be tiny and so the leak(s) must (should) be very small. It actually occurs to me that the most likely place for the leaks is in the valves that close off the chamber themselves.- In fact its quite probable that most such valves leak a little bit and that they are probably still within specifications. A look in the instruments manuals might be of help.

Register to Reply
Register to Reply 13 comments

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

Gavilan (1); harley (1); HiTekRedNek (1); PWSlack (1); RAMConsult (1); redfred (2); Robert Lucien (2); silvCrow (1); Yuri B. (3)

Previous in Forum: Pressure Calibration Uncertainty Constraints   Next in Forum: Problem in Operating Enpac 2500 Data Collector

Advertisement