Previous in Forum: Leakage Detection Through Piezoelectric Membrane   Next in Forum: Reason for Both Legs Connected in Wet Leg Level Measurement?
Close
Close
Close
13 comments
Rate Comments: Nested
Guru
Safety - ESD - New Member Popular Science - Cosmology - Amateur Astronomer Technical Fields - Technical Writing - Writer India - Member - Regular CR4 participant Engineering Fields - Optical Engineering - New Member

Join Date: Dec 2007
Location: 18 29 N 73 57E
Posts: 1390
Good Answers: 31

High Accuracy Temperature Indicator

09/18/2017 11:15 AM

I need high accuracy temperature indication for water at ambient temperature... range 0 - 50 deg C

The combined accuracy needed including sensor inaccuracy and indicator inaccuracy is better than 0.1 deg C and resolution 0.01 deg C

The sensor should be PT 100 (3 wire) , which can be integral or remote.

Can anyone suggest the supplier ? Fluke makes it which is very costly... almost $1000.

Register to Reply
Pathfinder Tags: accuracy PT 100 temperature
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Guru

Join Date: Oct 2008
Posts: 42355
Good Answers: 1693
#1

Re: High Accuracy Temperature Indicator

09/18/2017 11:21 AM

Google, "High Accuracy Temperature Indicator."

Register to Reply
Guru
Technical Fields - Technical Writing - New Member Engineering Fields - Piping Design Engineering - New Member

Join Date: May 2009
Location: Richland, WA, USA
Posts: 21017
Good Answers: 795
#2

Re: High Accuracy Temperature Indicator

09/18/2017 11:35 AM

You might a consider Pt 1000 sensor with appropriate instrumentation.

__________________
In vino veritas; in cervisia carmen; in aqua E. coli.
Register to Reply
2
Guru
Engineering Fields - Electrical Engineering - Been there, done that. Engineering Fields - Control Engineering - New Member

Join Date: Dec 2008
Location: Long Island NY
Posts: 15600
Good Answers: 981
#3

Re: High Accuracy Temperature Indicator

09/18/2017 12:01 PM

$1000 seems reasonable to me. Particularly if you want real credibility in the data you are taking.

Wait a minute, three wire measurements assume identical errors on the other leg. I guess you don't really want high accuracy then. You only want better accuracy than a two wire measurement.

__________________
"Don't disturb my circles." translation of Archimedes last words
Register to Reply Good Answer (Score 2)
2
Power-User

Join Date: Nov 2009
Posts: 335
Good Answers: 63
#4

Re: High Accuracy Temperature Indicator

09/18/2017 12:03 PM

1. Those kind of specs warrant a true 4 wire RTD.

2. 'high accuracy' and 'too costly' tend to be mutually exclusive.

Register to Reply Good Answer (Score 2)
2
Guru
Safety - Hazmat - New Member United States - US - Statue of Liberty - New Member Engineering Fields - Chemical Engineering - Old Hand

Join Date: Mar 2011
Location: Lubbock, Texas
Posts: 14331
Good Answers: 162
#5

Re: High Accuracy Temperature Indicator

09/18/2017 1:19 PM

For highest accuracy, you would need a good calibrator for the triple point of water, or solute-free freeze point, and solute-free boiling point at sea level.

You could go with a thermistor (cheaper), but you will sacrifice some accuracy.

If you cannot procure a good calibrator, you will need a thermometer that is certified to a defined standard. In the United States of America, I think we still use NIST traceable ones. At least I do when calibrating other temperature measurements against this secondary standard, such as wastewater measurement.

Pt1000 will offer you more measurement accuracy in the temperature range you are interested in, IMHO.

__________________
If it ain't broke, don't fix it. Just build a better one.
Register to Reply Good Answer (Score 2)
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#6

Re: High Accuracy Temperature Indicator

09/19/2017 2:57 AM

Why?

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#7

Re: High Accuracy Temperature Indicator

09/19/2017 7:49 AM

One should question whether the temperature in a liquid is anywhere the same within 0.01degC. Convection currents and evaporation render the measurement criterion dubious.

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
Guru
Safety - Hazmat - New Member United States - US - Statue of Liberty - New Member Engineering Fields - Chemical Engineering - Old Hand

Join Date: Mar 2011
Location: Lubbock, Texas
Posts: 14331
Good Answers: 162
#8
In reply to #7

Re: High Accuracy Temperature Indicator

09/19/2017 9:49 AM

What you are saying is true while the water is far from equilibrium (i.e.- being rapidly and locally heated, wherein convection, and conduction might not be keeping up with heat input, and evaporation is not able to restore a set temperature related to pressure.

__________________
If it ain't broke, don't fix it. Just build a better one.
Register to Reply
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#9
In reply to #8

Re: High Accuracy Temperature Indicator

09/19/2017 10:32 AM

Exactly. That is why the question stands as to the need for such a remarkable device in the first place. The original poster has withheld that detail from the forum.

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
Guru

Join Date: May 2016
Posts: 1746
Good Answers: 87
#10

Re: High Accuracy Temperature Indicator

09/19/2017 11:23 AM

A resolution of .01 and accuracy of .1 seems a bit pointless. Usually in measuring applications the desired accuracy of the measurement is 1/10th that of the desired measurement resolution.

Why would you want a reading of 23.04 if it could actually range from 22.99 to 23.09 and you don't know where in the range it actually is? In actual use, you would want a resolved reading of 23.04 that could range from 23.039 to 23.041.

Register to Reply
Guru
Safety - Hazmat - New Member United States - US - Statue of Liberty - New Member Engineering Fields - Chemical Engineering - Old Hand

Join Date: Mar 2011
Location: Lubbock, Texas
Posts: 14331
Good Answers: 162
#11
In reply to #10

Re: High Accuracy Temperature Indicator

09/19/2017 11:53 AM

Absolutely true, but sometimes we play trickery, and average a lot of numbers to arrive at central tendency? Yes? No?

__________________
If it ain't broke, don't fix it. Just build a better one.
Register to Reply
Guru

Join Date: May 2016
Posts: 1746
Good Answers: 87
#12
In reply to #11

Re: High Accuracy Temperature Indicator

09/19/2017 12:16 PM

Sounds like a reason to fudge temperature numbers...where have we seen that before?

Register to Reply
Guru
Safety - Hazmat - New Member United States - US - Statue of Liberty - New Member Engineering Fields - Chemical Engineering - Old Hand

Join Date: Mar 2011
Location: Lubbock, Texas
Posts: 14331
Good Answers: 162
#13
In reply to #12

Re: High Accuracy Temperature Indicator

09/19/2017 12:20 PM

Algore Industries?

__________________
If it ain't broke, don't fix it. Just build a better one.
Register to Reply Score 1 for Good Answer
Register to Reply 13 comments

Good Answers:

These comments received enough positive votes to make them "good answers".

"Almost" Good Answers:

Check out these comments that don't yet have enough votes to be "official" good answers and, if you agree with them, vote them!
Copy to Clipboard

Users who posted comments:

Carl_E (1); James Stewart (4); Jpfalt (2); lyn (1); PWSlack (3); redfred (1); Tornado (1)

Previous in Forum: Leakage Detection Through Piezoelectric Membrane   Next in Forum: Reason for Both Legs Connected in Wet Leg Level Measurement?

Advertisement