Oil Condition Testers

01/25/2018 2:54 PM

I have been looking at purchase of an oil conductivity test instrument (hand-held, portable, and only have one quote in hand thus far. If you have information on vendors for this, let me know (no I did not yet check CR4 techs and specs, but I intend to.)

Here is the question: One of the articles I was reading mentioned the use of a so-called "grid capacitor" test (not that kind of grid), it has a grid of wires over an insulating platen, and test lube oil is poured on the platen, then the top "plate" is put down over this, and tests for dielectric constant begin. One measures over the span of probably not more than about 10 minutes, and watches in the data how D.C. (dielectric constant) changes over time, always increasing, due to settling out of particulates suspended in oil, including water micro-droplets, carbon specs, dust, etc.

This latter test is a strong indicator of oil condition, providing a presumptive test on new oil, oil in service, and whether to use, continue using, or discard.

Is anyone out there familiar with this type of equipment and where to obtain it? Thus far, all my searches have struck out. I am not interested in taking this to a vendor for lab services, we already have that. We need something ASAP turnaround when we suspect oil contamination...and we always do.

__________________
If it ain't broke, don't fix it. Just build a better one.
Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#1

Re: Oil condition testers

01/25/2018 3:19 PM

<...oil conductivity test instrument...>

Oils are generally a mixture of organic molecules and do not have a conductivity, which is a property reserved for ionic species usually in aqueous solution, and oil and water do not mix.

<...dielectric constant...> is not <...conductivity...>.

<...water micro-droplets, carbon specs, dust, etc....> would suggest there is a significant materials segregation problem that is overdue for a remedy. Maybe the equipment that is being oiled is being pushed beyond its economic lifespan, or deep maintenance is being postponed inordinately, in which case catastrophic failure and revenue damage is imminent.

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
Guru
Safety - Hazmat - New Member United States - US - Statue of Liberty - New Member Engineering Fields - Chemical Engineering - Old Hand

Join Date: Mar 2011
Location: Lubbock, Texas
Posts: 14331
Good Answers: 162
#2
In reply to #1

Re: Oil condition testers

01/25/2018 3:46 PM

Class II lubricating oils (heavy hydro-treated base oils) without additives measure in around less than 10 pS/m. With anti-corrosion additives (formerly were zinc dithiophosphate), whatever they may now be as zinc ones were unsafe in the environment apparently, the conductivity value may rise to 400-600 pS/m.

I have found at least one meter available on the market for said ASTM D2624 protocol.

I am really more interested in the grid capacitor test. It has to do with particles settling out of the oil.

__________________
If it ain't broke, don't fix it. Just build a better one.
Register to Reply
Guru
Hobbies - DIY Welding - Don't Know What Made The Old Title Attractive... Popular Science - Weaponology - New Member United States - US - Statue of Liberty - 60 Year Member

Join Date: Apr 2009
Location: Yellowstone Valley, in Big Sky Country
Posts: 7425
Good Answers: 295
#3
In reply to #2

Re: Oil condition testers

01/25/2018 5:22 PM

ASTM Test Method D2624 is limited to fuels containing static dissipator additive.

Is this test method germane to Class II lube oils?

Link: Guide to ASTM Test Methods for the Analysis of Petroleum Products and Lubricants: 2nd Edition

__________________
Semper Ubi Sub Ubi
Register to Reply
Guru
Safety - Hazmat - New Member United States - US - Statue of Liberty - New Member Engineering Fields - Chemical Engineering - Old Hand

Join Date: Mar 2011
Location: Lubbock, Texas
Posts: 14331
Good Answers: 162
#5
In reply to #3

Re: Oil condition testers

01/26/2018 9:25 AM

Yes it is germane to testing lube oils.

http://www.machinerylubrication.com/Read/29407/oil-condition-monitoring

https://ac.els-cdn.com/S2212017314001625/1-s2.0-S2212017314001625-main.pdf?_tid=c5fabe1e-02a3-11e8-929c-00000aacb361&acdnat=1516976486_a3867dbfca8166395f18ea38df89be8b

The first link is about using EC (electrical conductivity) to test lube oils, the second includes a paragraph about using capacitive ( The dielectric constant (D.C.) is also known as relative permittivity) bridge testing to indirectly determine the D.C.

The grid capacitor test is not typically not or setup to be done online, but could be.

It requires the introduction of test oil between plates of a capacitor such that the bottom plate is actually a wire grid, and the test is under no flow condition, such that the particles in the oil settle to the grid, and affect the D.C. over the observation time of several minutes. This can give clues to water micro-droplet concentration, carbon particles from static-induced micro-sparks, etc. It is not typically viewed as a chip sensor like what would be on the oil system of a gas turbine.

__________________
If it ain't broke, don't fix it. Just build a better one.
Register to Reply
Guru

Join Date: Mar 2007
Location: by the beach in Florida
Posts: 33392
Good Answers: 1817
#4

Re: Oil condition testers

01/25/2018 9:37 PM

I don't think this is quite what you want....but we gotta start someplace....

https://www.indiamart.com/shree-abirami-enggineering/transformer-oil-test-lab.html

__________________
All living things seek to control their own destiny....this is the purpose of life
Register to Reply
Guru
Safety - Hazmat - New Member United States - US - Statue of Liberty - New Member Engineering Fields - Chemical Engineering - Old Hand

Join Date: Mar 2011
Location: Lubbock, Texas
Posts: 14331
Good Answers: 162
#6
In reply to #4

Re: Oil condition testers

01/26/2018 9:36 AM

That link is apparently for a full service laboratory offering an entire panel of available testing services, and not a bad idea at all for an organization to have contacts with.

We do already have oil analysis services, and our maintenance personnel are well-versed in grab sampling technique to ensure proper samples are recovered and sent periodically to the lab.

We want a "cheap" tester for on-site frequent use (if management will just approve it).

As to the grid capacitor I think I can make one, I do not see anything sophisticated about this at all, as long as all residues are easily cleaned away between tests.

I can see doing an inlay of copper wires in PE (polyethylene) or PTFE (polytetrafluoroethylene teflon) platen, machined to close tolerances with a lip or added gasket spacer of the same material to control spacing in the oil chamber. The upper platen can be the same thing, or a simple (very flat) piece of metal. Oil would be introduced in such a way that excess oil would be squeezed out when the top platen is closed. Capacitance bridge circuit then begins logging test results and continues for up to ten minute interval. Results would then be brought up and charted, or statistics such as the linear least squares fit or polynomial fit would be characterized. Dielectric constant changes also computed from the changes in this test capacitor. Test frequency could even be 60 Hz, as the limit for this type of test is a mere 104 Hz, since we want to be on the flat part of the curve. No high frequency is required as would be in solid dielectric testing.

__________________
If it ain't broke, don't fix it. Just build a better one.
Register to Reply
Guru

Join Date: Mar 2007
Location: by the beach in Florida
Posts: 33392
Good Answers: 1817
#15
In reply to #6

Re: Oil condition testers

01/26/2018 4:51 PM

Maybe something here you can use...

http://www.agarscientific.com/tem/grids-agar.html

__________________
All living things seek to control their own destiny....this is the purpose of life
Register to Reply
Guru

Join Date: May 2016
Posts: 1746
Good Answers: 87
#7

Re: Oil Condition Testers

01/26/2018 11:05 AM

I googled conductivity meters and got a lot of results for Hach, Rosemound, Extech, etc. I had started looking for a mhometer, but the industry calls them conductivity meters.

Any reason you couldn't make a parallel copper plate plunge type probe and put it on a megger?

Register to Reply
Guru
Safety - Hazmat - New Member United States - US - Statue of Liberty - New Member Engineering Fields - Chemical Engineering - Old Hand

Join Date: Mar 2011
Location: Lubbock, Texas
Posts: 14331
Good Answers: 162
#8
In reply to #7

Re: Oil Condition Testers

01/26/2018 11:13 AM

That might work, since we already (obviously) in a power company have the megger.

There is about 8 order of magnitude more sensitivity (i.e.-less current) for a pS/m type conductivity instrument for oil, than a μS/cm type instrument for water conductivity.

Thanks for the idea. The grid capacitor thing, I can take care of with some Arduino wizardry, I suspect.

How are things in the far north (Oregon)?

__________________
If it ain't broke, don't fix it. Just build a better one.
Register to Reply
Guru

Join Date: May 2016
Posts: 1746
Good Answers: 87
#9
In reply to #8

Re: Oil Condition Testers

01/26/2018 11:20 AM

Wet.

Register to Reply
Guru
Safety - Hazmat - New Member United States - US - Statue of Liberty - New Member Engineering Fields - Chemical Engineering - Old Hand

Join Date: Mar 2011
Location: Lubbock, Texas
Posts: 14331
Good Answers: 162
#10
In reply to #9

Re: Oil Condition Testers

01/26/2018 12:04 PM

Wet megger test, is that what you are referring to? Yeah, this can be a problem is really wet oil is tried.

__________________
If it ain't broke, don't fix it. Just build a better one.
Register to Reply
Guru

Join Date: May 2016
Posts: 1746
Good Answers: 87
#11
In reply to #10

Re: Oil Condition Testers

01/26/2018 12:09 PM

Nope. Wet. It's raining outside.

Register to Reply
Guru
Safety - Hazmat - New Member United States - US - Statue of Liberty - New Member Engineering Fields - Chemical Engineering - Old Hand

Join Date: Mar 2011
Location: Lubbock, Texas
Posts: 14331
Good Answers: 162
#12
In reply to #11

Re: Oil Condition Testers

01/26/2018 12:14 PM

Oh, we are "wet" here also, since today was the first in about 100-120 days I have been able feel, smell the moisture in the air, and we had enough precipitation to dampen the sidewalks in a couple of places from dew fall. That is how things are here, either feast or famine.

Since the Christ child (El Nino) is over, I would not expect much this year, and also at solar minimum. I can be wrong also.

__________________
If it ain't broke, don't fix it. Just build a better one.
Register to Reply
Guru

Join Date: Oct 2014
Location: Hemet, Land of milk and honey.
Posts: 2365
Good Answers: 36
#13
In reply to #12

Re: Oil Condition Testers

01/26/2018 2:39 PM

I don't know anything about this stuff. Is what you want the kind of stuff provided by Parker Hannifin. ?

Register to Reply
Guru
Safety - Hazmat - New Member United States - US - Statue of Liberty - New Member Engineering Fields - Chemical Engineering - Old Hand

Join Date: Mar 2011
Location: Lubbock, Texas
Posts: 14331
Good Answers: 162
#14
In reply to #13

Re: Oil Condition Testers

01/26/2018 3:40 PM

I looked at some of their stuff, and did not find oil conductivity tester, but other kit.

__________________
If it ain't broke, don't fix it. Just build a better one.
Register to Reply
Guru
Engineering Fields - Power Engineering - New Member

Join Date: May 2007
Location: NYC metropolitan area.
Posts: 3230
Good Answers: 444
#16

Re: Oil Condition Testers

01/29/2018 10:49 PM

Not quite sure which paper you were reading, but there are many out there that describe the use of capacitor based sensors. Similar to DGA (dissolved Gas Analysis) of transformer oil, the "gotcha" is that you need a series of baseline data points to compare with the data as the oil ages.

You might find some more information on capacitor-based sensors in some of the papers discussing oil health condition monitoring.

__________________
“Tell me and I forget. Teach me and I remember. Involve me and I learn.” Ben Franklin.
Register to Reply