Thermohygrometer Edgetech Dewmaster

12/07/2008 6:54 AM

Hello

I have an Edgetech Dewmaster with a an error of 1,3%RH at conditions of 20ºC / 90%RH.

Does anyone knows how to adjust it? Is it possible?

Thank you,

Filipe Rocha

info@activeng.pt

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.
Guru
Hobbies - DIY Welding - New Member

Join Date: Sep 2007
Posts: 1601
Good Answers: 58
#1

Re: Thermohygrometer Edgetech Dewmaster

12/07/2008 10:52 PM

I suggest you contact the manufacturer. The Edgetech is a chilled mirror hygrometer with advertised accuracy of +/- 0.2degC. These devices can give false readings if the gas flow rate through the sensor is not within specified limits. Under normal conditions it is a primary standard and should not require calibration.

Register to Reply
Guru
Hobbies - DIY Welding - New Member

Join Date: Sep 2007
Posts: 1601
Good Answers: 58
#2

Re: Thermohygrometer Edgetech Dewmaster

12/07/2008 11:08 PM

On second thought, +/- 0.2 deg C is about 2.2% RH at the conditions you specify(20 deg.C, 90%RH). Therefore, your Dewmaster may be giving readings within its specified accuracy.

Register to Reply
Guru
Popular Science - Weaponology - New Member Safety - ESD - New Member Hobbies - Fishing - New Member

Join Date: Sep 2006
Location: Near Frankfurt am Main, Germany. 50.390866N, 8.884827E
Posts: 17996
Good Answers: 200
#3

Re: Thermohygrometer Edgetech Dewmaster

12/08/2008 3:55 AM

I cannot believe that any process needs more accuracy than this!! Plus the fact that you appear to know the error, so you could easily add/subtract the known error from the measured value to get almost 100% correct!!!

Where are you coming from?????

__________________
"What others say about you reveals more about them, than it does you." Anon.
Register to Reply
Power-User

Join Date: Jan 2007
Location: Berlin (Germany)
Posts: 332
Good Answers: 1
#4

Re: Thermohygrometer Edgetech Dewmaster

12/08/2008 6:29 AM

You can mix your own calibration solutions as follows:

1. Take LiCl and mix a saturated solution in a closed glass. After some hours you get 13% Relative Humidity above the solution`s surface at 20°C which can be measured via a small sensor`s opening. Give the sensor some time to stabilize.

2. Take NaCl and you get 75% RH. For most processes this is accuracy enough to make a 2-point-calibration.

3. If not, you should preferably use a different type of measurement (which is used by the weather frogs). The best accuracy is achieved by measuring two temperatures - the dry thermometer temp and the wet bulb temp with an air velocity going over the wet bulb of about 2 m/s. This is called "aspiration psychrometry", with the dry temperature and the temp-difference gou go into aspiration tables or programs and read the % RH. This gives you the best accuracy because the errors are reduced on those of temp-measurements, which can be done very exactly.

Regards Uwe

__________________
The sum of intelligence on earth is a constant. And the population grows and grows and .....
Register to Reply
Guru
Hobbies - DIY Welding - New Member

Join Date: Sep 2007
Posts: 1601
Good Answers: 58
#5
In reply to #4

Re: Thermohygrometer Edgetech Dewmaster

12/08/2008 8:50 AM

Certainly saturated salt solutions are useful for generating known humidities. For a better description see:
http://www.omega.com/temperature/z/pdf/z103.pdf
It can be assumed from the statement of the problem that the error is already known. Aspiration psychrometry has many known sources of errors and are not primary standards. Dewpoint instruments are considered primary standards. The Dewmaster described, has an accuracy of 0.2 deg.C. Other similar instruments can be obtained with accuracies of 0.1deg.C.

Register to Reply
Participant

Join Date: Dec 2008
Posts: 2
#6

Re: Thermohygrometer Edgetech Dewmaster

12/08/2008 3:28 PM

Thank you all for your help.

However, since there is a deviatin on %rh, possibly it may be cused by a deviation on them temperature sensor. Is this sensor "fixed"? Can it be adjusted? Maybe, being adjustable, I can get better 5rh results.

Thank you for all you help.

Filipe Rocha