High Potential Testing

05/14/2009 9:02 AM

what is the acceptable limit of leakage current for high potential test of electrical equipment??

Register to Reply
Interested in this topic? By joining CR4 you can "subscribe" to
this discussion and receive notification when new comments are added.
Guru
United Kingdom - Member - Indeterminate Engineering Fields - Control Engineering - New Member

Join Date: Jan 2007
Location: In the bothy, 7 chains down the line from Dodman's Lane level crossing, in the nation formerly known as Great Britain. Kettle's on.
Posts: 32175
Good Answers: 839
#1

Re: High Potential Testing

05/14/2009 11:41 AM

What sort of equipment?

__________________
"Did you get my e-mail?" - "The biggest problem in communication is the illusion that it has taken place" - George Bernard Shaw, 1856
Register to Reply
Active Contributor

Join Date: Jan 2008
Location: Bedford Massachussets
Posts: 11
#2
In reply to #1

Re: High Potential Testing

05/15/2009 9:54 AM

It depends on what kind of device. Medical devices are more stringent then others. Even for a medical device it depends on the kind and what leakage test you are performing. Per EN 60601-1 the limit for No Fault in a type B medical device is 0.5 mA with a Single Fault it is 1.0 mA for Earth Leakage current. For Enclosure Leakage current it becomes 0.1 No Fault and 0.5 Single Fault.

__________________
Gerwoman
Register to Reply
Participant

Join Date: Jun 2008
Posts: 2
#3
In reply to #2

Re: High Potential Testing

05/16/2009 8:27 AM

For example, i want to perform high potential test for any kind of high voltage circuit breaker. Or its better for me to refer the instruction manual for the acceptable limit of leakage current if there is...

Register to Reply Off Topic (Score 5)